So this has developed over the last 4-6 months to the point of being chronic and it’s driving me nuts.
Synopsis: I switched to T-Slim approaching 4 years ago from my old Medtronic Paradigm dumb-pump (hated their AID systems). I was using Fiasp, which I was very happy with but is off-label for T-slim. Read some complaints about Fiasp causing occlusions but mostly people said it was ok, including my trainer, who was a Fiasp user herself. For three-odd years I’ve had no problems other than the usual bad insertion or whatnot. I could often get 4 days on an infusion set/reservoir with no difficulties whatsoever.
But over the last months that has ceased to be the case. I can’t even get the statutory 72 hrs. Everything’s fine day one and most of day two but after 48 hrs the insulin potency noticeably drops off and then I start getting No Delivery alerts. Can’t dial up a bolus–just stops after a unit or not even anything. No choice but to swap everything out and start a new session.
Inset condition: usually fine but sometimes the last millimeter of the cannula is slightly bent.
The latter led to the suggestion to try switching to 6mm cannulas vs the 9mm I’ve always used. I’ve lost about 25 lbs over the last year and Clinical Support suggested that could be an issue. Tried two but nope, same problem. Day 3 occlusion nightmare.
I’ve done the standard occlusion test: disconnect from “pigtail” and deliver 5 units; if that checks out, reconnect pigtail and do the same at the inset end. Must have done this half a dozen times or more now. On one occasion I got insulin from the pigtail end but NOT from the inset clip. 5 units dialed up, nothing came out. But most of the time both check out.
The above result, along with the “insulin fade” issue after48 hrs seemed to indicate maybe Fiasp clumping was an issue, though why now after 3+ years is a mystery.
Scar tissue? I rotate religiously between abdomen out to the love handles and upper thighs. Been at this long enough that yeah, there’s a lot of marks everywhere. But the problem is constant no matter the location and I’ve tried some newer areas farther up on my abdomen with no luck.
I read up on Lyumjev and the info says it’s as fast or a little faster acting than Fiasp, yay, plus it IS on-label for T-slim (uses a different molecule for the rapidity effect). So I got my endo to prescribe it and I started on it yesterday. So far it’s been great. Definitely faster (I got into a little trouble yesterday when it came on sooner than I was ready for), and it seems like the efficacy is all I could want. I’ve been fasting today other than a.m. coffee (splenda and light cream) and I’ve been running right at 100 ± 4 all day.
BUT
This is only day one (about 27 hrs from the change as I write this). So it’s a little premature–I won’t know for at least another 24 hrs if this was the fix. I’ll report back. But meanwhile would love to hear from anyone else with experience around this because it has been baffling, frustrating and not a little depressing. Stuff I’d love a response on:
If it does turn out Fiasp was the problem, why is it only cropping up now after all this time?
If god forbid the Lyumjev craps out sometime around tomorrow night (per the rule that says persistent occlusion alarms must always occur between midnight and 4am), what the heck do I try next. Please don’t say Truesteel because I tried them a long while back and… shudder. But if I gotta I gotta.
It doesn’t seem like the pressure sensor or something in the pump itself is the issue but should I rule that out? It’s actually a replacement for my original one that bricked sometime last spring. Not sure if that means it’s a reconditioned one or brand new. There’s a lag between when I got the replacement pump and when I started to notice this issue but it was occasional at first before it became constant, so I might not have noticed if it’s some kind of slow-developing fault in the machine. Anyone experienced a pressure sensor going off and does any of this fit the pattern?
And sorry to have been absent from these precincts for a while. Just had other focuses and matters T1 have been pretty stable before this cropped up. Always do enjoy my time here.
I never got good results from Fiasp so maybe I didn’t spend long enough on it to see issues. I did get great results from Lyumjev. However this is a sample of 1. I use AAPS (an AIDS, like Tandem) and I’ve been getting maybe worse results from Lyumjev recently,
Next year I’m going to go to Merilog, because of insurance changes; that’s Fiasp without the skin irritant
We’re injecting stuff under our skin that was never meant to be there in that quantity. We have a hyper-active immune system. Is it any surprise that sooner or later our immune system notices and sets up a credible defense?
So while everyone on this forum consistently decries our insurance companies changing our insulin I think it is a good thing. I only wish I could still do Pig and Cow insulin; we’re just stuck with these ridiculous analogs which only have a one-base change. I look at it as me against my own immune system - I dodge, I weave, I certainly duck.
So I’m saying don’t worry about it until it fails, then try another insulin. Just make sure it is really different.
Thanks, a thoughtful and nuanced response. I do differ on the question of analogs vs the older stuff. I was on the old R/N regimen for twenty solid years, starting out on the animal insulins before the recombinant stuff came along but either way that regimen was SO restrictive. I had to beg to get my PCP (didn’t have an endo; HMO didn’t do specialist referrals without a gun to their heads) to let me try Lantus/Novolog and it took changing from university to private insurance before I finally got referred to Joslin and they put me on it, taught me the whole carb-counting regimen schtick. It was like being let out of prison. I literally wept. So we differ there.
But I think you’re right about changing insulins not being necessarily a bad thing (see my update below).
Update as promised. It is now officially the Dread Day Three: my pump just beeped a Site Change Reminder.
And oh god it didn’t crap out once!
No occlusions last night, and a.m. fasting BG was within tolerance–135. Higher than yesterday’s 103, but I did have a couple slices of pizza last night and I’ve always had a pretty severe DP. With great temerity I bolused for coffee with a slight correction added and sat there waiting for the hideous BEEP BEEP BEEP Non-Delivery message… which never came. Here I am a couple hours later at 105. I still have 70 units in the reservoir and I think I’m gonna just let this set run like I would have before this problem started biting my butt.
And as for Lyumjev itself I’m really liking it. It’s noticeably faster–I think I mentioned I had a low on Day 1 bc it came on sooner than expected. And it’s so nice to enter a correction and actually have it DO something. Hat tip @jbowler – I think you’re right about changing things up periodically.
Congrats on hitting day 3, and hope you have a working solution. I’m not familiar ay all with the insulin types you mentioned, am on Lispro/humalog ( I have been T1D so long remember the 1970s U-40 beef and pork Lilly bottles).
I wish I could get 3 days on a T-Slim! My issue is how much I seem to need, and even with getting closer to 80% time in range I still experience peaks and valleys, only rarely do I see pump alerts for occlusions. Heck, i actually do a cartridge refill and push a longer run, likely not a great strategy.
A big help is now having the Tru-Steel infusion sets available again, it always seemed the plastic tube inserts always gave me issues, I prefer the real steel
I was afraid that was going to be my last option if Lyumjev switch didn’t work out, but fortunately it seems to have done the trick. And it is noticeably faster and more effective at knocking down highs. If I’d known that I might have made the change when I got the pump. At the time I didn’t know it was in the same category as Fiasp–thought that was the only ultra-rapid.