New sensor won't accept any calibration - "Unknown code: 12"

I started a new G6 sensor on Xdrip+ yesterday and it’s been a PITA. First off it gives me the error “G5 sensor start failed” (remember, it’s a G6 sensor). But it still does the warm-up period and starts working. However, it rejects every single calibration, even maddeningly after it asks for one. The error is always the same:

Calibration Rejected:
Unknown code: 12

The first two calibrations were off by quite a bit - 16% and 32%, so I thought it might be out of too far off to believe. But the third fingerstick was exactly the same as the CGM so that isn’t it. Double calibration and repeated calibrations all rejected. Restarting the sensor gives the exact same results. Anyone have any ideas about what’s going on with this thing and how I might be able to fix it?

1 Like

@bwschulz The G6 shouldn’t be asking for calibrations unless you didn’t put the four digit sensor code in. Maybe you didn’t check “I am using G6” in settings>G5/G6 Debug Settings"?

1 Like

It might have asked for a calibration because I entered a double calibration in the attempt to get it to accept something (which may have bumped it out of G6 and into G5 mode), but that G6 tickbox is checked. I’ve been using this setup (G6 with Xdrip+) for a long time now and have never had anything like this. I’ve had some sensors that had issues on startup but they eventually settled down or restarted fine. This one is just goofy. I restarted it again this morning and it says “Please wait while sensor warms up!” while at the same time still merrily displaying BG readings. I’ve never had that happen either.

1 Like

Which transmitter are you using? The new 8g and 8H transmitters don’t allow third party apps to connect to them

1 Like

The 8Hxxxx sees the light of day…

lol

1 Like

I got a two pack last month

I don’t know the specific model, but this transmitter has been working fine for many other sensors.

1 Like

The first two characters of the Transmitter serial number.

With the various transmitters out there it helps to know which one is being discussed.

Sometimes it is relevant and sometimes not.

1 Like

Thanks- it’s an 81XXXX.
What are Xdrip+ users going to do/doing when they get those new 8g or 8H transmitters that don’t work? Is that forcing them back to the crappy Dexcom app, or is there any workaround?

The various transmitters were described in a posting here, but that seems inaccurate given, with regard to the 8G and xDrip+:

Anyway I will know in the coming month because my 81 transmitter has 88.6 days on the clock and the only replacement I have is an 8G (unless Walmart supplies me with an earlier transmitter in the next few days, this is entirely possible.)