Thanks Scott for the chance to review a ReliOn Premier Voice meter! I appreciate the opportunity.
First, a few observations:
-
The meter setup is very easy. Everything is voice assisted.
-
Testing is easy, and the results are given both via voice and on meter display.
-
Many of the functions on the meter are voice assisted.
-
Occasionally I got an E4 - “the sample was not enough or too thick” Okay, but the thing beeped after I put the blood on. Mr. Meter, why would you beep and start the countdown if there was not enough blood? Most every other meter I have ever used will wait until there is enough blood before it starts its countdown. If the problem is that my blood is too thick, that’s a problem because it’s the only blood I have. (My blood is occasionally thick from dehydration after a workout, but I have never had that error on any other meter).
-
I like the desiccant container that is built into the vial cap. I have not used any meters with that feature. Smart.
-
The meter allows data transfer to your computer but I could not test this because I don’t have the cable.
-
You can review 1, 7, 14, 30, and 90 day BG averages. That is very nice. Not all meters allow this much granularity in displaying averages.
-
You can set hypo alert from 20-60. Nice that you can pick the range, but some people might want to be able to set their hypo alert higher, like 70.
(Side-note: I tried many times to get a test that would give me a hypo alert. But it never happened! Despite testing a number of times when I was absolutely low and knew I was low. This was maddening!
A little more info on meter accuracy problems is given below.)
-
You can set time alarms and post-meal alarms, but the post-meal alarm is only available for 2 hours. Allowing some adjustment of the post-meal alarm time increment would be nice.
-
You can test yourself without looking, the voice does walk you through it. But you can’t really tell if the strip is inserted right-side up or not. Perhaps having one side of the test strip rough and the other side smooth would make this possible to do without sight?
-
No light on the meter
Seems like a lot of meters do not have this, which is unfortunate. A light is a very useful thing on a meter. It’s why I take my FreeStyle when I go to the movie theater… -
This is a big one - no voice review of past tests?!?!?!? Nicky mentioned this too, and it is extremely puzzling to me. Why make a voice meter and not allow people to hear their previous test results? It makes no sense. This really is a feature that should be added. If a person is able to see the past results, why would they be using a voice meter in the first place?!? Voice review of past tests should be added to this meter!
Now for the test result comparisons, which were a little bit disappointing.
I compared the ReliOn Premier Voice meter with two very accurate meters I have been using for a long time - the Contour Next One and the FreeStyle Lite.
Most of the tests I did were a direct one-to-one comparisons with these meters, with the tests done only seconds apart. 34 of the 50 tests I did were direct comparisons. I did these tests at very random times and with a wide range of BG’s.
-
72% of the time, the BG shown on the ReliOn Premier Voice was higher than what was shown on the Contour Next One.
-
100% of the time, the BG shown on the ReliOn Premier Voice was higher than what was shown on the FreeStyle Lite.
-
The Contour had about a 7% average difference from the Relion.
-
The FreeStyle had a 34% average difference. 3 tests were off by a very large percentage (greater than 30%). If I throw those 3 results out, the avg difference for the FreeStyle was still 19%.
-
I had a number of times when I tested and was significantly low but the ReliOn Premier Voice did not pick it up! For example, being 42 and the ReliOn Premier Voice showing a 64. That was a serious problem to me. Happened several times. That was kind of a deal-breaker for me…
For average deviation tests, the ReliOn Premier Voice was very good when I did the tests just by itself (with one exception).
The average deviation was 0.875, which by itself looks great.
But one time when not testing average deviation, I saw a number on the ReliOn Premier Voice that was waaaaaay off. I re-did it and saw a completely different number. I saw 179 and 73 (the Contour was showing 109 at the time, which was likely closer to correct at the time).
So that 179 is what really bothered me - a lot!
Obviously that 179 was wrong, but I found it troubling to see. I don’t see one-offs like that on any meter I trust. So that might be something that I would be troubled by, until I tested over and over with the meter and after not seeing an error like that after a few hundred tests with it I might be comfortable with it again.
Bottom line - this meter may work okay for you, and the voice feature may be useful.
But the failure to flag the lows, and the meter almost always being higher than what I felt was my true BG value, make this a meter I would not entirely trust.