Omnipod issues massive pod recall today

Omnipod issued a massive pod recall today around 5:30PM EDT.

If you need to replace more than 20 affected Omnipod 5 pods, or more than 10 Omnipod Dash pods, you have to call Product Support directly. Otherwise, you can just use their online reporting form for those smaller pod amounts.

I would recommend checking your lot numbers now and calling ASAP. They anticipate running out of pod stock during this recall.

10 Likes

Did a bit of digging through ChatGPT ands found the below.

Insulet/Omnipod posted an Urgent: New Medical Device Correction for Omnipod Pods today, May 26, 2026. It affects specific lots of:

Omnipod 5 Pods
Omnipod DASH Pods
Omnipod Eros / Omnipod Insulin Management System Pods

This is separate from the March 12, 2026 Omnipod 5 recall. The March issue involved a possible internal tubing leak inside the Pod. This new May 26 issue involves a possible small tear in the cannula/tubing just above the skin, meaning insulin may leak outside the infusion site instead of fully going into the body. (Omnipod)

The risk is under-delivery of insulin, which can cause unexplained highs, highs that do not respond like expected to correction insulin, and in serious cases DKA. Insulet says there have been 24 serious adverse event reports globally, including hospitalization and DKA, with no deaths reported. (Insulet Corporation)

What concerns me most is that Insulet says this may happen without an alert, so families should not rely only on Pod alarms. You might notice wetness on the skin or adhesive, or smell insulin, but it can also be hard to detect. (Omnipod)

What you should do now:

Check the lot number on the Pod tray lid, the 5-pack Pod box, or the flat side/bottom of the Pod. Insulet says all Pods in the same box are from the same lot, so if one Pod box lot is affected, do not use any Pods from that box. (Omnipod)

Use Insulet’s official Check Pods & Request Replacements page, or call Product Support at 1-800-641-2049, available 24/7. The official checker page says you enter the Pod lot number and, if needed, request replacements. (Omnipod)

For the United States, the affected lots listed by Insulet include:

Omnipod DASH:
PD1U01082621, PD1U02152521, PD1U02192521, PD1U04242522, PD1U09062421, PD1U09092421, PD1U09102421, PD1U10242421, PD1U10252421, PD1U10282421, PD1U10282522, PD1U10292421, PD1U10292521, PD1U10302521, PD1U10312421, PD1U10312521, PD1U11012421, PD1U11022421, PD1U11032521, PD1U11042421, PD1U11042521, PD1U11052421, PD1U11192421. (Omnipod)

Omnipod 5:
The U.S. Omnipod 5 list is very long, so I would not manually rely on eyeballing it. Use the official lot checker. The list includes many PH1U and PR1U lots, and Insulet’s page has the full U.S. list. (Omnipod)

7 Likes

There are eros pods and other dash pods too. I’ll paste it in here from the email I received just for completeness. [thanks to Beacher for pointing out that this list is for a person in the USA, and the lot numbers differ elsewhere]

Appendix B: Affected Pod lots
UbicaciĂłn del nĂşmero de lote del Pod en el propio Pod (lado plano):

Omnipod (Eros) Part # 19191
L72514
Omnipod DASH Part # 18320
PD1U09062421, PD1U09092421, PD1U09102421, PD1U10242421, PD1U10252421, PD1U10282421, PD1U10292421, PD1U10312421, PD1U11042421, PD1U11192421
Omnipod DASH Part # 18325
PD1U01082621, PD1U02152521, PD1U02192521, PD1U04242522, PD1U10282522, PD1U10292521, PD1U10302521, PD1U10312521, PD1U11012421, PD1U11022421, PD1U11032521, PD1U11042521, PD1U11052421
Omnipod 5 Part # PT-001662
PH1U01032521, PH1U01102521, PH1U01132521, PH1U01172521, PH1U01222521, PH1U01232522, PH1U02132511, PH1U02212521, PH1U02222521, PH1U03202521, PH1U03262521, PH1U03282521, PH1U03312521, PH1U04032521, PH1U04042521, PH1U04302524, PH1U05012521, PH1U05022521, PH1U05052521, PH1U06282422, PH1U09112421, PH1U09122421, PH1U09122422, PH1U09162421, PH1U09212421, PH1U09262411, PH1U09262421, PH1U10042421, PH1U10102421, PH1U10112422, PH1U10122421, PH1U10172522, PH1U10182521, PH1U10202521, PH1U10212521, PH1U10222521, PH1U10232521, PH1U10242521, PH1U10252521, PH1U11152521, PH1U11172511, PH1U11172521, PH1U11182521, PH1U11192511, PH1U11192521, PH1U12022421, PH1U12092421, PH1U12102412, PH1U12102421, PH1U12122421, PH1U12132422, PH1U12172421, PH1U12182421, PR1U01132621, PR1U01152621, PR1U01162621, PR1U01172621, PR1U01202621, PR1U01212621, PR1U01222621, PR1U01232621, PR1U01272621, PR1U01302641, PR1U01312631, PR1U02022641, PR1U02032631, PR1U02032641, PR1U02042641, PR1U02052631, PR1U02052641, PR1U02062631, PR1U02112641, PR1U02132611, PR1U02132641, PR1U11182531, PR1U11182541, PR1U11192531, PR1U11192532, PR1U11192541, PR1U11192542, PR1U11202521, PR1U11202532, PR1U11202533, PR1U11202541, PR1U11202542, PR1U11212521, PR1U11222521, PR1U11222531, PR1U11222541, PR1U11232541, PR1U11242521, PR1U11242531, PR1U11252512, PR1U11252521, PR1U11252531, PR1U11252541, PR1U11262521, PR1U11262541, PR1U12012521, PR1U12022521, PR1U12032521, PR1U12102522, PR1U12112521, PR1U12122521, PR1U12132511, PR1U12132521, PR1U12152521, PR1U12162521, PH1U09172421, PR1U12172521
Omnipod 5 Part # PT-001444
PH1U01092621, PH1U02202521, PH1U04222521, PH1U04232511, PH1U04232521, PH1U10182422, PH1U11142521, PH1U12042521, PH1U12052521
Omnipod 5 Part # PT-01977
PH1U01312521, PH1U02042521, PH1U04092521, PH1U04112521, PH1U10142421, PH1U10152421, PH1U10162421, PH1U11212421, PH1U11222421, PH1U11252421, PH1U11262421, PH1U01302522, PH1U01312511, PH1U02032521, PH1U04102521, PH1U09182421, PH1U09192421, PH1U09202421, PH1U10172421
6 Likes

I bet this is why my pump shipment is delayed!

1 Like

Note that these US lot numbers won’t be the same in other countries. I got the notice in Canada yesterday and there’s only one lot number for O5 pods (PH1U01222522). Numbers for DASH and Eros pods are also different.

ETA: To my knowledge, there was no recall in Canada last month. The May 26 notice is the first one I’ve received.

7 Likes

Thanks for posting about this, @T1Allison and others with links/lot #s!

2 Likes

Doe anyone know if there is a date associated with these pods?

Nevermind, I see the lot number actually reflects the dates shown on the pod boxes.

So that might be an easier way to do it, if you use your stuff by order of date, and have it stored that way.

For example, the 013124 number in the middle of the lot number means it was manufactured Jan 31, 2024.

I’m just going to turn all the lot numbers given above into dates. That will be easier for me to search!

5 Likes

Just finished going through @Liam-M’s supply. Here are the impacted lots / PODs from his supply.

9 Likes

Same. It killed two of my three month shipments.

5 Likes

I only did the Dash dates. Here they are, if this is useful for anyone.
(For me it’s just easier than looking at a long string of numbers.)

September 6, 2024
September 9, 2024
September 10, 2024
October 24, 2024
October 25, 2024
October 28, 2024
October 29, 2024
October 31, 2024
November 1, 2024
November 2, 2024
November 4, 2024
November 5, 2024
November 19, 2024
February 15, 2025
February 19, 2025
April 24, 2025
October 28, 2025
October 29, 2025
October 30, 2025
October 31, 2025
November 3, 2025
November 4, 2025
January 8, 2026

2 Likes

Have any of us who’ve been using these recalled pods actually experienced any issues?

Alright, I just got off of the phone with customer support…here’s what I can tell you based on my experience.

For those who Loop, they STILL WILL NEED TO KNOW PDM SN. IF it’s already associated / linked to your record, you are fine…in our case, Liam’s PDM wasn’t registered on his account so I had to email a picture of the PDM SN (I still had the PDM in the back of the closet).

They are going to work on getting the PDM registered to his account and then they’re going to call me back.

Once they call me back and confirm the PDM registration onto his account, they’re going to be shipping us SIXTY (60) PODS ONLY…they are only doing 60 at a time and they are expecting the other PODs to be returned. They will provide return label and I’ll be expected to return 60 of the impacted PODs and they once they receive those, they’ll ship out the remaining 35.

That’s the experience I had! Hope it helps you all if you’re doing the same!

Have that PDM in case you need it!

7 Likes

Yep, I have. I used 3 pods with this issue not being aware of the recall. All 3 pods caused massive increases in insulin (around 50% increase) especially post dinner. Despite frequent corrections, I still went very high for me. I didn’tnotice leakage per se on the pod though, but I really didn’t look, not being aware of the recall. I called Insulet this morning and they will be replacing 8 boxes plus 2 pods (42 pods) for my most recent shipment. Thankfully I have a supply of older Dash pods with no issues. I replaced the recalled one with the older pod and my insulin dosage is back to normal.

4 Likes

Hmm, they did not ask for my PDM SN. I just told them my issue and the lot number for the 9 boxes. They said I should receive them next week along with a prepaid envelope to send the recalled pods back. They did ask me how long I’ve used OmniPod and I said for years.

4 Likes

Maybe they had PDM already linked to your account. That was their issue with me tonight. They said it would have been faster if that linkage was already established.

3 Likes

WOW, @ClaudnDaye! Is there anything left?! Quite the haul there. But if we order as needed and include the 8 hour grace, they do add up! I read elsewhere that some had in excess of 40 pods! Also, there’s a reported separate channel folks have referred to if over 30 or 35 pods. The need for a PDM SN seems to depend on who you get when you call! With millions of pods impacted, I assume they’re quite busy!

3 Likes

That is about 1/6 of my stock. :smiley:

1 Like

I used the chat function on the Insulet site yesterday and it was a very smooth process. They didn’t ask for PDM info and are shipping 60 of my 65 replacements. They’ll send the other 5 when they get the recalled pods.

2 Likes

Yes, maybe but I haven’t used a PDM in about 10 years and that was with the Eros pods. They did route my call to a different dept because they didn’t have enough info on me. They had my email though!

1 Like