Transient Left–Right Fingerstick Glucose Differences and Dexcom G7 Discrepancies — A Call for Scientific Investigation.

An Unexpected Observation: Transient Left–Right Fingerstick Glucose Differences and Dexcom G7 Discrepancies — A Call for Scientific Investigation

Hello everyone,

I have Type 2 diabetes, use insulin, and monitor my glucose using a Dexcom G7 continuous glucose monitor (CGM).

I’m sharing an observation that I believe deserves scientific investigation.

During a series of glucose measurements, I noticed substantial differences between fingerstick readings from my left and right hands. These differences subsequently diminished. At the same time, my Dexcom G7 readings differed from the capillary blood glucose measurements.

I’m particularly interested in whether these observations can be explained entirely by measurement variability and known CGM limitations, or whether regional physiological factors, potentially involving insulin activity and peripheral circulation, deserve investigation.

My objective is to attract researchers, clinicians, physiologists, biomedical engineers, and glucose-monitoring specialists to examine these questions.

  1. WHY THIS MATTERS

Like many people using CGMs, I’ve become accustomed to relying heavily on continuously displayed glucose readings.

The convenience of CGMs means that fingerstick testing can become less frequent.

However, when glucose levels are changing rapidly, discrepancies between CGM readings and actual capillary blood glucose measurements could have important consequences.

If CGM measurements differ significantly from fingerstick measurements at critical moments, understanding the causes becomes particularly important for people using insulin.

I am not suggesting that CGMs are inherently unreliable. I’m asking whether there are physiological or measurement-related effects that deserve more systematic investigation.

  1. EQUIPMENT AND METHODOLOGY

Measurement date: October 16, 2026

CGM: Dexcom G7
Display: Dexcom application on iPhone
Sensor location: Right waist/hip

Fingerstick meter: ReliOn Premier
Test strips: ReliOn Premier blood glucose test strips
Test-strip lot/batch: ZJ2688TRA
Test-strip use-by date: 2028-02-25

Fingerstick measurements were taken sequentially from the left and right hands using the same glucose meter.

The fingers were cleaned with alcohol before testing.

Earlier bilateral comparisons used the ring fingers, with the left hand tested first.

Important methodological limitation: The Dexcom G7 was positioned near my right waist/hip rather than at a standard FDA-approved insertion site in the United States. The possible effects of sensor location must therefore be considered.

The test strips were within their labeled shelf life.

  1. BILATERAL FINGERSTICK OBSERVATIONS

Five sequential pairs of fingerstick measurements produced the following results.

All values are in mg/dL.

Test 1:
Left: 152
Right: 186
Difference: 34

Test 2:
Left: 143
Right: 176
Difference: 33

Test 3:
Left: 129
Right: 140
Difference: 11

Test 4:
Left: 133
Right: 134
Difference: 1

Test 5:
Left: 132
Right: 139
Difference: 7

The first two measurements showed differences exceeding 30 mg/dL between hands.

Subsequent measurements converged substantially.

Because these were sequential measurements rather than simultaneous blood samples analyzed under laboratory conditions, changing glucose levels and measurement variability remain possible explanations.

Nevertheless, the pattern raises questions about the relative contributions of measurement uncertainty and physiological processes.

  1. DEXCOM G7 OBSERVATIONS

During the broader observation period, the CGM displayed readings including:

201, 233, 237, 220, 182, and 175 mg/dL.

These values were not all precisely synchronized with individual fingerstick measurements.

Later, the Dexcom G7 displayed a reading of 119 mg/dL.

At approximately that time, I obtained additional fingerstick measurements:

Dexcom G7: 119 mg/dL
Right pinky finger: 100 mg/dL
Left-hand finger: 103 mg/dL

The two fingerstick readings differed by only 3 mg/dL.

The CGM reading was approximately 16–19 mg/dL higher than the fingerstick readings.

This observation is interesting because the bilateral fingerstick values had become much closer than during the initial measurements.

The precise elapsed time between all measurements, and the rate of glucose change, were not independently established.

  1. THE SCIENTIFIC HYPOTHESIS

I wonder whether transient differences in regional physiology could contribute to some glucose measurement discrepancies.

Possible mechanisms worth investigating include:

  • Differences in peripheral blood perfusion.
  • Regional differences in glucose utilization.
  • Effects of changing insulin activity on glucose dynamics.
  • Differences between capillary blood glucose and interstitial glucose during rapid changes.
  • Temporal differences between sequential measurements.

Alternative explanations include:

  • Normal meter and test-strip variability.
  • Fingerstick sampling conditions.
  • Residual alcohol or other sample contamination.
  • Differences in sample volume.
  • CGM interstitial glucose lag.
  • Effects of nonstandard CGM sensor placement.
  • Ordinary changes in blood glucose during sequential testing.

These explanations are not mutually exclusive.

My observations do not establish that regional insulin action or physiological differences caused the discrepancies.

The purpose of publishing them is to invite controlled investigation.

  1. WHY I BELIEVE THIS QUESTION IS IMPORTANT

CGMs have transformed diabetes monitoring.

However, the more convenient continuous monitoring becomes, the easier it may be to overlook differences between sensor readings and independently measured capillary glucose.

For insulin-treated individuals, accurate interpretation of glucose readings can be consequential.

An especially important question is whether glucose-monitoring technologies and their accuracy evaluations adequately characterize discrepancies during periods of rapid glucose change.

A related question concerns whether apparent differences between anatomical measurement sites sometimes reflect physiological processes rather than measurement noise.

My observations cannot resolve either question, but they may provide a useful starting point for discussion.

  1. QUESTIONS FOR RESEARCHERS

I would welcome informed responses to the following questions:

  1. Are transient left-versus-right capillary glucose differences of 30 mg/dL or more documented in controlled scientific studies?

  2. Could differences in peripheral circulation, regional glucose uptake, or insulin-mediated glucose metabolism plausibly contribute to transient bilateral discrepancies?

  3. How much variability should be expected when sequential glucose measurements are performed using the same meter and test-strip brand?

  4. What role might interstitial glucose lag play during rapidly changing glucose concentrations?

  5. How might Dexcom G7 sensor placement influence agreement with fingerstick measurements?

  6. Are there published studies that systematically compare simultaneous bilateral capillary glucose measurements with CGM readings?

  7. What experimental protocol would best distinguish measurement artifacts from physiological effects?

A useful controlled experiment might involve simultaneous bilateral capillary sampling, repeated measurements, an independent laboratory glucose reference, and synchronized CGM recordings.

  1. INVITATION TO SCIENTIFIC INVESTIGATION

This is a preliminary observation from one individual.

I recognize its limitations, including sequential sampling, the absence of laboratory reference measurements, incomplete timing information, and the nonstandard CGM insertion site.

I am not claiming to have discovered a new physiological mechanism or demonstrated that insulin causes differences between glucose measurements from opposite hands.

Rather, I am presenting the observations transparently and asking whether they raise questions worth investigating.

My hope is that endocrinologists, researchers, physiologists, medical-device engineers, and experienced CGM users can contribute relevant scientific evidence, published research, and suggestions for controlled experiments.

The ultimate objective is to improve our understanding of glucose measurement reliability, especially for people whose daily decisions depend heavily on continuous glucose monitoring.

Thank you for your scientific input and discussion.

I didn’t read all of your post, but readings within 20% are considered acceptable, so although your finger stick readings aren’t identical, they’re close enough for a home monitoring kit.

1 Like

@rajalur blood glucose levels vary throughout the body and with time. CGMs **do not ** measure blood glucose levels but interstitial glucose levels. There is a lag around 15 minutes between BG and interstitial levels. This is most observable when BG is changing rapidly.

As @allison said plus or minus 20% is considered acceptable. The important thing is overall trends.

These devices are not like volt meters. There are many more factors that they have to deal with electrical, chemical and biochemical.

It is miraculous that meters and CGMs work so well.

5 Likes

Your differences are all entirely believable for correctly functioning equipment. I just did a similar test; my left hand gave 64mg/dL, my right hand 74mg/dL. My G7 was displaying an instantaneous reading of 75mg/dL and a smoothed reading of 79mg/dL. A couple of minutes later the CGM reading is 78mg/dL and the smoothed result is now 78mg/dL.

There is a known difference of several minutes between a fingerstick and the corresponding CGM reading. The CGM smoothing adds to this. For the Dexcom CGMs, where readings only happen every 5 minutes, the smoothing adds at least 2.5 minutes to the delay (the addition depends on the smoothing algorithm).

I hardly ever use my right hand for a fingerstick because I’m right handed; I had to poke my right fingers about 5 times before I got enough blood, partly because of my dexterity, partly because the skin on my right hand is thicker due to greater use and partly because both hands were cold.

The standard advice here and on other fora is that comparing CGM readings with fingersticks is only valid when flatlining; clearly I wasn’t in this case because of the low BG. Fingerstick readings should be within 15% of the corresponding veinous BG but that means that two fingerstick readings can be over 32% (of the veinous BG) apart.

If we assume your “Test 1” veinous BG was anywhere in the range 179mg/dL to 161mg/dL both fingersticks are within the measurement accuracy (152 is 15% less than 179 and 186 is 15% more than 161). Indeed if we assume a veinous BG of 168mg/dL both fingersticks are off by 10%.

Similarly the CGMs have instantaneous errors which don’t have any particular rules so far as I can see. I’ve seen what I consider massive, and dangerous, CGM errors from the Dexcom G6 and G7; false lows and massive errors with low BG.

The errors in the measurement devices (all of them) mean drawing any conclusions is pretty much impossible. The relevant tests that are done by manufacturers and, in the US, the FDA are to measure both fingersticks and CGMs against veinous BG but then that requires someone who can be hooked up to a continous hospital grade monitoring system and that is not a real world scenario for any diabetic!

More precisely: Over-reliance or trust in the absolute readings can be harmful. CGMs are a guide, a clue and an enormous help in managing diabetes. In particular there are reports of a lack of awareness of BG but that can be fixed by using CGMs and, when in doubt, fingersticks to achieve better Time-In-Range and do develop BG awareness.

Most non-Ds seem not to be aware of the effects of high BG and, often, low BG but with CGMs we can correlate the significant psychological and physical symptoms of both using the CGM readings. I.e. we can learn a high level of BG awareness. The awareness is from our venous BG; I know I’m going low and, often going high well before my CGM notices!

4 Likes

I had a similar observation. My finger stick bg on the right side is always higher than my left side.

Always found it interesting, but never investigated.

1 Like

Thanks, @jbowler, for the thoughtful technical analysis—and for actually running your own experiment! And @jim26, your observation that your right hand consistently reads higher is particularly intriguing.

I agree that my measurements could fall within expected instrument uncertainty. But here’s the distinction I’m interested in:

Demonstrating that a discrepancy falls within an instrument’s error tolerance is not the same as demonstrating that the discrepancy is caused by random measurement error.

Interestingly, researchers have investigated systematic left–right CGM differences:

A 2019 study, “Differences in Glucose Levels Between Left and Right Arm,” reported differences reaching 41 mg/dL.

A subsequent, larger study involving 46 participants found statistically significant differences between right- and left-arm CGM readings.

Neither study proves a physiological difference in capillary blood glucose, but both suggest the question deserves investigation.

@jim26, does your right-higher-than-left pattern persist across repeated measurements when sampling order and technique are controlled?

@jbowler, your observations about cold hands, skin thickness, sampling difficulty, and measurement uncertainty suggest exactly the variables a controlled experiment should account for.

Perhaps repeated bilateral fingersticks, randomized testing order, standardized sampling conditions, and ideally a laboratory reference could help distinguish random measurement noise from reproducible directional differences.

The null hypothesis may well be correct: it’s all measurement noise. But shouldn’t we test that hypothesis before declaring the mystery solved? :wink:

After all, specifications tell us how much uncertainty we tolerate. Science asks what’s producing it.

I’d welcome further observations and suggestions for investigating this systematically.

My observation is the same as before: “A man with one watch knows what time it is; a man with two watches is never sure!” Perhaps a third measurement? Say from a toe? (I’m smiling…in actuality, it’s a serious issue, perhaps without an answer other than we can only do what we can do.) We should continue to push for accuracy from CGMs!

3 Likes

…and BG meters!

I believe that’s on a hyperbolic curve, ever closer.

Just an observation no controlled technique other than I change my lancet twice a year. I will say my right hand is primarily used for finger sticks and if I feel it might not be accurate, I go to the other hand. So over time I thought the difference was interesting.

3 Likes

One of the hypotheses I offered (the one you didn’t quote) is handed-ness; my reference to dexterity. The paper you quote first is here:

Executive summary (read the TL;DR link):

The study took 1920 samples from two right handed non-diabetic males using the Abbott Freestyle Libre Pro CGM system with sensors on both arms readings every fifteen minutes. The readings were compared with each compared reading being within 5 minutes of the other. [More details in the link.]

Using appropriate statistical methods (I have limited experience but they used the right words) the experimental subjects consistently (96% of the time) showed higher readings on the right arm. There’s no information about where on the right arm but I assume they used somewhere in Abbott’s recommended location.

The authors did not produce an estimate of the differences between the two arms (that’s in the 46 patient study, which reports 3.7mg/dL difference.) They did give arithmetic means of each pair of arms but that’s not statistically relevant; IRC Student’s t-test is the gospel here and they did not report a statistical difference in the average measurement. (“Student” is the adopted name of the guy who invented the test.)

The authors did try the same experiment (768 data points, 8 days at 15 minute intervals, rather than the 960 they used on the male right paws) on a leftie:

Interestingly, when the sensor was worn by a left-handed person for 8 days (768 data pairs), the right arm was higher only 21.5% of the time.

Hum. Sorry, that suggests a certain amount of embarrassment on behalf of the authors to me, particularly in the use of the double negative. Nevertheless their methods are good, so far as I can see, and their results are clear. I’m going to quote their entire summary verbatim here because, well, this is what they said and it seems a very reasonable statement to me:

These findings need to be validated in larger studies with a diverse patient population to eliminate the presence of any sensor related bias. In addition, a study enrolling patients that are left-arm-dominant versus right-arm-dominant to delineate the mechanism behind any real differences due to sensor site selection would be valuable. Since interstitial fluid behaves as a different compartment relative to blood, it is possible that muscle/fat mass differences between the two arms explains the glucose level discordance between the arms.5 It may not be unreasonable to study the placement of two sensors on each arm to assess inter- and intra-arm differences. Bleeding from the skin can also interfere with the sensor suggesting the need for assessing the degree of bleeding during sensor placement and removal.6

The links are to the relevant entry in the references section of the paper (I verified this.)

The “46”. Here’s the paper:

Once again it’s a CGM, so interstitial BG as @CarlosLuis pointed out. The CGM is, again, the Abbott FreeStyle Libre Pro and they took readings every 15 minutes; i.e. identical to the first paper. 83% of the test subjects are right-paws.

They consider the “dominance” issue, report observing it in a prior study, and conclude that it is not relevant in the sense that their study showed no correlation. I.e. the 3.7mg/dL difference between the right arm and the left arm cannot be solely explained by the previous result’s suggestion of arm dominance as an issue. I worded that very carefully but I probably still got it wrong.

If I accept that (it’s only one study) I come back to my second hypothesis that the result is not arm dominance but rather the difference between blood flow to the left and right sides of our bodies.

This is certainly interesting.

2 Likes

Coincidentally…this morning I tested finger-stick glucose using Contour Next and finding right hand was 4 mg/dl higher than left. Nominate me for Patient #47!!

A number of years ago I was a patient in a clinical study for a new CGM. As a part of it, we sat around the clinic while blood was drawn numerous times and BG was tested on those blood draws. (They did not use both arms for these blood draws…should they have?)

It was a blind study so I never saw the correlation with the studied CGM. However I was wearing a Dexcom G6 the entire time and don’t recall any wild inaccuracy compared to the blood draws. My recollection is that the “10 minute lag” we attribute to CGM compared to fingerstick was not consistently observed for Dexcom G6 vs blood draws.

2 Likes

I’ve mentioned before on here that I find persistent differences in how my G7s read between my left and right arms, which I chalked up to differences in muscle/fat composition. It’s never off enough to make a huge difference in treatment (both arms are within the “acceptable” error bounds for CGMs). It would be good to have more studies, but it’s interesting that this shows up outside my n-of-1 observations. I wouldn’t be surprised if small biophysical differences account for variation in how BGMs read as well.

1 Like