I have just started my free trial Libre sensor. I also started new dexcom, and intentionally delayed Libre startup to make better comparison. Libre is on left arm, dexcom on right.

I inserted Libre Sat 2 pm, not realizing the startup time was 12 hours, so wished I had done it sooner. Started new dexcom 2:30 pm.

So started getting Libre readings at 2:10 am, but did some logs, half asleep.
F = flat
U = 45 arrow up
D = 45 arrow down
SU or SD if straight up or down, but haven’t gotten this yet.

Majority on both have been flat.

Time Libre Dexcom. Meter Bg
2:10a. 107 F, 125 F . 133
2:15a. 109 F, 130 F. 128
8:53a 100 F, 141 U, 125
9:03a. 108 F, 149 U,
11:32a. 86 F, 88 F,
12:18p. 85 F, 98 F, 94

Now 24 hrs since start Libre and dex. They are much closer.
2:07p. 127 F, 125 F. 135
3:51p, 154 U, 149 U,
4:34p, 105 chm
4:44p, 88 chm
5:05p, 65 chm
5:13p, 69 chm, 93 on Neo.
5:21p, 81 F,
5:50p, 103 F,
6:05p, 114 F, 110 F
6:17p, 115 F, 107 F, 122 BG verio
7:00p, 84 F, 85 F.
7:10p, 82 F, 83 F. 91 on Neo BG chk
8:07p, 48 F, 59 F. 87 on Neo BG chk

(Does not show arrow when there is the check symbol, noted above as chm)

The 3-430 ish time frame shows libre a bit confused. I was “snacking” at costco, walking, and gave bolus so was quite a bit going on BG wise. When I scanned, it kept giving me the indicator to check BG with meter, but I chose not to until I got home. At the same time dexcom was rising, to about 160, then fell quickly.

5 Likes

It seems like at first there was quite a bit of difference betweenness your Libre and Dexcom numbers, then Libre got closer throughout the day? How’s it going today?

I finally got the replacement reader/sensor thing straightened out with the pharmacy. I inserted a new sensor Friday morning, then activated it Saturday night. This time it hasn’t been hurting like the first sensor did at first, so that may be part of why the first sensor was so off. I started getting data yesterday morning, and so far it seems a bit more accurate. There was a slight period where it was telling me I was going low when I wasn’t (2:10, while I was in church), not sure why. It’s been really nice to be able to use this for testing “new” foods (baked potato last night), instead of using a ton of strips.

8:35am - 82 Libre, 72 ReliOn
10:15am - 106 Libre, 115 ReliOn
12:20pm - 92 Libre, 97 ReliOn
2:10pm - 66 Libre, 87 ReliOn
5:50pm - 69 Libre, 72 ReliOn
6:50pm - 145 Libre, 127 ReliOn
7:10pm - 157 Libre, 139 ReliOn
7:45pm - 146 Libre, 135 ReliOn
10:30pm - 82 Libre, 87 ReliOn

I normally would’ve corrected last night, but I wanted to see what would happen, plus I’ve been having pretty unpredictable drops because of some renewed beta cell activity (my endo is excited…I have mixed feelings, as the unpredictability is not fun).

2 Likes

Yes, the numbers on both got better, as expected.
So far today:

Time. Libre. Dexcom Meter BG
9:27 83 F, 84 F
9:43 70 F, 90 F. Neo 81, Next 86
11:16 131 F, 126 F
11:46. 126 F, 126 F

I like the ability to get reading on demand from Libre, so can see rapid changes, minute by minute. Dexcom only updates every 5 minutes and averages, so it is slower to show the rise or fall.

4 Likes

@MM2, these are great numbers! I am shocked that it converged so qickly.

2 Likes

My observation is that Libre shows slant arrow at 1 pts per min, whereas dexcom is 2. So while dexcom stays flat, Libre seems to oscillate more between flat and slant arrow when I am rising or dropping slowly. Dexcom stays flat arrow, but the numbers are slowly increasing/decreasing.

Time. Libre. Dexcom. Meter BG
9:14p. 92 D. 98 F. Next 110, Neo 99
9:20p. 87 D. 96 F
9:25p 82 D. 93 F
9:30p. 80 F. 90 F
9:40p. 74 D. 87 F
9:45p. 77 F. 87 F. Next 94, Neo 86.
9:55p. 71 F, 82 F
10:05p 73 F, 81 F.

11:10p 80 F, 86 F.

9:45 Bounce from 74 to 77, was that due to Neo BG check ?? What if not using FS Libre Neo strips, and use another meter ? I have not found option to log/enter BG from another meter. Does Libre try to correlate/adjust based on Meter BG ?
But dexcom also showed 5 pt drop. So maybe coincidence.
I did not enter BG to dexcom.

Since I am doing many more readings than listed, the battery is wearing down fast. I charged last night, and now about 1/3 left. May have to charge every day !
Update: after 2nd recharge, Libre is now lasting 2 days, and still showing full charge.

Summary for 50 minutes
Libre 19 point drop, several with slant arrow down. (92 - 73)
Dexcom. 17 point drop, always flat. (98 - 81)

Next… experiment with super fast up or down to get dexcom double arrows !!!

EDIT: According to manual, the Neo meter BG readings do not ‘calibrate’ the Libre. The sensor is calibrated ‘at the factory’.

5 Likes

These numbers are still very good!

I wonder if the Libre skews slightly low in order to make sure patients don’t end up too low. Emily’s numbers also skewed left.

It’s not clear to me if Libre takes meter input from Neo and tries to converge sensor BG readings with it, similar to when calibrations are put into dexcom. Using Neo strips is not required to use Libre.

2 Likes

I posted on a Libre Facebook group about the overnight false lows (not compression lows), and several responded that they have similar experiences. That being said, as of last night, I’m no longer reading low all night - it looked perfectly normal - so it may just be the sensor sometimes has trouble reading correctly at night for whatever reason. The change for me could just be due to other factors, though: slight diet changes, no longer nursing baby overnight, etc.

A similar trend I’ve noticed that you will see below is that, generally speaking, if I’m on the low end of my range, the Libre will read lower than my meter. If I’m on the higher end of my range, it reads higher than my meter. It’s still very close, so not really an issue to me; just interesting.

Life’s been crazy, but here are the numbers from the past few days - I’m depending on the Libre more and more (as evidenced by the fewer finger prick tests), since it seems right on with my meter almost all the time and I’m so preoccupied with things it’s easier just to scan and go. I love it. I’ve been making some diet changes, and this first day I don’t think I was eating enough carbs, which is why my numbers were so great. :slightly_smiling_face:

1/22
7:30am 67 Libre, 69 ReliOn
9:45 105 Libre, 90 ReliOn
12:15 64 Libre, 76 ReliOn
1:35 74 Libre, 80 ReliOn
6:40 73 Libre, 72 ReliOn
8:00 79 Libre, 81 ReliOn
10:00 86 Libre, 94 ReliOn
11:00 72 Libre, 75 ReliOn

1/23
5:30am 65 Libre, 71 ReliOn
9:40 112 Libre, 133 ReliOn
6:10 92 Libre, 92 ReliOn
8:00 110 Libre, 113 ReliOn
10:15 104 Libre, 97 ReliOn

1/24
11:20am 91 Libre, 81 ReliOn
7:00pm 139 Libre, 125 ReliOn
8:45 142 Libre, 136 ReliOn
10:00 113 Libre, 112 ReliOn

4 Likes

The difference between how it’s been reading at night (left) versus last night (right).

3 Likes

I also continue to get results that are pretty close.

Just now
Libre sensor 84 F
Dexcom 79 F

Verio BG. 88
FS Neo BG 92

4 Likes

Two things, as I read through this thread:

First, laying on the sensor (Libre or Dexcom) can cause false lows. The area around the sensor gets compressed and the amount of interstitial fluid decreases, so the sensor readings drop. Lay on the other side, or get up in the morning and the interstitial fluid equalizes, returning readings to normal. I have this issue if I wear my Dexcom sensor on my left arm because I sleep on my left side.

Second, keep in mind a “flat arrow” only means the bg “may be flat” or that it is increasing/decreasing only slightly. Dexcom claims it is only increasing/decreasing less than 1mg/dl per minute. That means you could rise or fall 30 points in 30 minutes and still maintain a flat arrow.

4 Likes

I’ve been wearing Dex and Libre sensors on the same arm for about 9 days and have noticed that Libre will show a 3 am low that is not noticed by Dex. Will keep using both for a few more weeks to see whats causing it.

So far I find the Libre to be usually pretty close to Contour Next One meter BG.

3 Likes

Last night was rough, still a problem this morning - first time eating out with the Libre, which went really well until I got distracted picking up our kids from the sitter and getting them in bed so stopped checking my BG. By the time I caught the delayed spike (I think a combination of the Afrezza wearing off and I had eaten a very fatty meal), I couldn’t get it back down - only managed to keep it from going higher with my corrections.

5:14 118R 109L
5:25 Ate dinner
5:35 109L dosed 8u Afrezza
6:00 Ate dessert
6:05 115L dosed 4u Afrezza
7:10 128R 127L (flat)
8:05 187R 165L dosed 4u Afrezza correction
8:20 167R 190L
Stayed at 190L
9:05 189R 194L dosed 4u Afrezza correction
Stayed at 194L until 10
10:30 178R 178L
10:38 164L
10:55 131L
11:15 135L
11:20 143L
12:15 120L
12:45 92R 92chm

I didn’t correct again at 10 because I was ready for bed and knew my pancreas would kick in at some point and didn’t want to risk a bad low in my sleep. Once things seemed to level out around 11, I went to sleep, only to be woken by baby at midnight, which was fortunate because that’s when I started dropping more rapidly. I ate a spoonful of peanut butter and stayed in range around 100-120 until morning (fasting 115 - much higher than my normal).

I opted for a lower carb breakfast - sausage and spinach, cottage cheese with a couple strawberries and almonds, and my Crio Bru w/a little sweetened creamer - and even though I dosed a bit extra since I was already at 115, I’m now at 164 and rising according to the Libre (checked meter 15 mins ago and was at 144, dosed a 4u Afrezza correction). Today will be interesting.

Overall I was impressed with how well the Libre kept up with the spikes/drops and all the Afrezza doses. I’m glad I had it, because I might not have thought to check after that 7:10 check with my meter until bedtime (depending on how I felt), and by then who knows how high my BG would have gotten.

4 Likes

I’ve been trying to get my own rx for a libre filled and thus far it’s been a total donkey show. I can’t help but think that they are nowhere near ready for prime time. Quickly losing interest myself.

1 Like

Try Walgreen’s, they seem to be able to sort it out.

FYI so far the Libre has withstood all challenges I have thrown at it.

1 Like

Nearest walgreens is about 100 miles away-- I wonder if they’d mail it to me from there?

2 Likes

Got me there, I’m still learning the ins and outs of pharmacies. My local pharmacy “could not get the Libre from the supply house” and faxed my prescrip to a nearby Walgreen’s. For what it’s worth, the pharmacist there said they’ve been flying off the shelf.

2 Likes

My Walgreens can’t get them either. If you can find one that can get it, they will mail it.

My local Walgreens did not have it in their system, but CVS was able to get it in one day.

2 Likes