# Xdrip version corrupted?

**URL:** https://forum.fudiabetes.org/t/xdrip-version-corrupted/8892
**Category:** Apps
**Created:** [December 15, 2019, 1:41pm UTC](https://forum.fudiabetes.org/t/xdrip-version-corrupted/8892 "2019-12-15T13:41:11Z")
**Posts on this page:** 14
**Page:** 1

<div class="post-metadata">

### Author: ![Osoblu](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/osoblu/32/2258_2.png) [@Osoblu](https://forum.fudiabetes.org/u/Osoblu)
#### Post date: [December 15, 2019, 1:41pm UTC](https://forum.fudiabetes.org/t/xdrip-version-corrupted/8892/1 "2019-12-15T13:41:11Z")

</div>

I recently downloaded an update a few days ago. Yesterday I started a new sensor. I now notice error on the status page.

 ![Screenshot_20191215-082647_xDrip](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/e/e6ad62a53af981b04a9c8d460fc116d70bd8ba3b.jpeg) ![Screenshot_20191215-082643_xDrip](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/a/a4f95063828e67f8a44bd3d5eb72f93f12314ebe.jpeg)  
Connection Status:FW: null  
Version : Information corrupted

---

<div class="post-metadata">

### Author: ![Osoblu](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/osoblu/32/2258_2.png) [@Osoblu](https://forum.fudiabetes.org/u/Osoblu)
#### Post date: [December 15, 2019, 1:48pm UTC](https://forum.fudiabetes.org/t/xdrip-version-corrupted/8892/2 "2019-12-15T13:48:14Z")

</div>

Glucose readings seem ok. Any ideas on how to resolve this?  
Thanks.

---

<div class="post-metadata">

### Author: ![Trying](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/trying/32/4078_2.png) [@Trying](https://forum.fudiabetes.org/u/Trying)
#### Post date: [December 15, 2019, 1:53pm UTC](https://forum.fudiabetes.org/t/xdrip-version-corrupted/8892/3 "2019-12-15T13:53:55Z")

</div>

Are you using G6? Did you change your transmitter, too? The G6 ID usually starts with an 8. The FW message could be referring to not understanding the transmitter id.

---

<div class="post-metadata">

### Author: ![Osoblu](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/osoblu/32/2258_2.png) [@Osoblu](https://forum.fudiabetes.org/u/Osoblu)
#### Post date: [December 15, 2019, 2:11pm UTC](https://forum.fudiabetes.org/t/xdrip-version-corrupted/8892/4 "2019-12-15T14:11:51Z")

</div>

It’s a G5 and is 60 days old. I downloaded the update “thinking” that I had Stable- tried and tested version selected. It was actually set to Beta. Oops. Xdrip has so many settings that I prefer to keep things simple. I wish there was a good tutorial to help me totally understand it.

---

<div class="post-metadata">

### Author: ![Trying](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/trying/32/4078_2.png) [@Trying](https://forum.fudiabetes.org/u/Trying)
#### Post date: [December 15, 2019, 2:32pm UTC](https://forum.fudiabetes.org/t/xdrip-version-corrupted/8892/5 "2019-12-15T14:32:03Z")

</div>

So glad you resolved it!!

---

<div class="post-metadata">

### Author: ![Osoblu](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/osoblu/32/2258_2.png) [@Osoblu](https://forum.fudiabetes.org/u/Osoblu)
#### Post date: [December 15, 2019, 2:41pm UTC](https://forum.fudiabetes.org/t/xdrip-version-corrupted/8892/6 "2019-12-15T14:41:28Z")

</div>

No. It’s not resolved. Same issues. I think reverting back to my last software version will solve it but I don’t know how. 🤔

---

<div class="post-metadata">

### Author: ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)
#### Post date: [December 15, 2019, 3:49pm UTC](https://forum.fudiabetes.org/t/xdrip-version-corrupted/8892/7 "2019-12-15T15:49:17Z")

</div>

Have you tried restarting the collector? Hit the button three times. That solves a myriad of connection problems , and some info from the transmitter is screwed up.

---

<div class="post-metadata">

### Author: ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)
#### Post date: [December 15, 2019, 3:51pm UTC](https://forum.fudiabetes.org/t/xdrip-version-corrupted/8892/8 "2019-12-15T15:51:15Z")

</div>

> [@Osoblu](#):
>
> I think reverting back to my last software version will solve it but I don’t know how. 🤔

You can revert simply by downloading and installing the desired version. But I don’t think xDrip+ is the problem.

---

<div class="post-metadata">

### Author: ![Osoblu](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/osoblu/32/2258_2.png) [@Osoblu](https://forum.fudiabetes.org/u/Osoblu)
#### Post date: [December 15, 2019, 4:11pm UTC](https://forum.fudiabetes.org/t/xdrip-version-corrupted/8892/9 "2019-12-15T16:11:13Z")

</div>

Thanks very much.  
I did revert back to the last version 5-18-2019. Everything looks fine now. I wish I had tried your 3X button press first. I only pressed that button once. ☹  
How did you learn that? Is there a good tutorial out there somewhere? I’d be more inclined to get the latest builds.

---

<div class="post-metadata">

### Author: ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)
#### Post date: [December 15, 2019, 4:19pm UTC](https://forum.fudiabetes.org/t/xdrip-version-corrupted/8892/10 "2019-12-15T16:19:31Z")

</div>

I’ve been lazy about this. I’m about half way through, but no ETA. I was able to figure out the three times by trial and error. Now I use the Tasker plugin for restarting the collector every hour, so my collection rate is 99-100%.

I think, since it’s so easy to revert, that updating to the nightlies involves very minimal risk.

---

<div class="post-metadata">

### Author: ![jbowler](https://avatars.discourse-cdn.com/v4/letter/j/e274bd/32.png) [@jbowler](https://forum.fudiabetes.org/u/jbowler)
#### Post date: [December 16, 2019, 12:04am UTC](https://forum.fudiabetes.org/t/xdrip-version-corrupted/8892/11 "2019-12-16T00:04:41Z")

</div>

> [@Osoblu](#):
>
> I did revert back to the last version 5-18-2019.

That was the beta channel version until a few days ago, then 20191027 dropped. The “System Status”/“G5/G6 Status” page is markedly different between G5 and G6, here’s a piccie of what shows with G6 on the same page with the same (October) xDrip+ version:

 ![xDrip 20191027 G6 status display screen](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/8/87bfb8ec52ecfcf00939dff88d57694e8b25f42c.jpeg)

So far as I can see so far the 20191027 version contains certain changes to fix the new messages emitted by the G6 ‘8G’ (and, maybe, later) transmitters and it contains a patch to disable preemptive restarts on those transmitters: “Preemptive restarts: Not capable” 😉

I suspect there is nothing new for G5 users, but that’s a hypothesis and therefore subject to disproval. The May (20190518) version was the previous version in the beta channel; I don’t know what is in stable.

---

<div class="post-metadata">

### Author: ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)
#### Post date: [December 16, 2019, 1:46am UTC](https://forum.fudiabetes.org/t/xdrip-version-corrupted/8892/12 "2019-12-16T01:46:07Z")

</div>

There is significant new info on the G5/G6 System Status page for the G5.

 ![Screenshot_20191215-193804_xDrip](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/4/4a315116348f5e1881afa5b38f373d89da31a132.jpeg)

We got 2 firmware versions, ASIC, and transmitter shelf life. I have no idea what firmware is being referred to, nor do I know what ASIC is.

---

<div class="post-metadata">

### Author: ![jbowler](https://avatars.discourse-cdn.com/v4/letter/j/e274bd/32.png) [@jbowler](https://forum.fudiabetes.org/u/jbowler)
#### Post date: [December 16, 2019, 1:56am UTC](https://forum.fudiabetes.org/t/xdrip-version-corrupted/8892/13 "2019-12-16T01:56:51Z")

</div>

> [@docslotnick](#):
>
> I have no idea what firmware is being referred to

In the transmitter; there is no other firmware. Interesting that the G5 transmitter reports a shelf life yet the G6 does not (so far as the authors of xDrip+ know).

FWIW the version of _xDrip+_ is on the “classic status page”; right drag off the above pages, and the relevant information is the second “Code” line, which is an ISO format date/time of the build of the software. (The bit above seems to be a unique number tagging the source code of the software - a “git tag” in software gobbledigook).

---

<div class="post-metadata">

### Author: ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)
#### Post date: [December 16, 2019, 2:45am UTC](https://forum.fudiabetes.org/t/xdrip-version-corrupted/8892/14 "2019-12-16T02:45:00Z")

</div>

Thanks @jbowler. I knew the Classic System Status Page had the version number and “git-tag”, (from searching for various versions), but I don’t know what ASIC refers to. I certainly can’t correlate the number it gives to an Application-Specific Integrated Circuit. And JamOrHam seems to get scarce when somebody brings it up.