# Tips on resetting transmittor on g6 via android phone

**URL:** <https://forum.fudiabetes.org/t/tips-on-resetting-transmittor-on-g6-via-android-phone/8160>\
**Category:** Tips, Tech & Gear\
**Created:** [August 22, 2019, 8:01am UTC](https://forum.fudiabetes.org/t/tips-on-resetting-transmittor-on-g6-via-android-phone/8160 "2019-08-22T08:01:09Z")\
**Posts on this page:** 12\
**Page:** 1

<div class="post-metadata">

**Author:** ![shaye](https://avatars.discourse-cdn.com/v4/letter/s/b782af/32.png) [@shaye](https://forum.fudiabetes.org/u/shaye)\
**Post date:** [August 22, 2019, 8:01am UTC](https://forum.fudiabetes.org/t/tips-on-resetting-transmittor-on-g6-via-android-phone/8160/1 "2019-08-22T08:01:09Z")

</div>

can u plz help me to reset my g6 transmittor as i only have an android LG V20 phone! thank you

---

<div class="post-metadata">

**Author:** ![MM2](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mm2/32/3032_2.png) [@MM2](https://forum.fudiabetes.org/u/MM2)\
**Post date:** [August 22, 2019, 6:34pm UTC](https://forum.fudiabetes.org/t/tips-on-resetting-transmittor-on-g6-via-android-phone/8160/2 "2019-08-22T18:34:35Z")

</div>

Many use xDrip on android phone. See details here.

> [@New to xdrip, how do I install it?](https://forum.fudiabetes.org/t/new-to-xdrip-how-do-i-install-it/5242):
>
> I recently read about the xdrip app for the dexcom 6 cgm. Everything I read is for using ann apple or android watch with it to view results. Can I somehow use it just on my Samsung galaxy 8+? Can someone please help me and tell me steps to download it and get it working on my phone? Also how I can use the sensors more than 10 days? Thanks to anyone who responds

This essentially replaces your Dexcom app, and provides more features.

With xDrip, you can reset trans, but don’t need to because xDrip does not stop reading based on number of days, like the dexcom does.

Another post on reset with xDrip.

> [@G6 with xDrip+ on Android](https://forum.fudiabetes.org/t/g6-with-xdrip-on-android/5402):
>
> I’m running a G6 straight to a Pixel and then to a Garmin Forerunner (no receiver or Dexcom app). This has been running great for months, but then today when I tried to start a new sensor I’m having connection issues; after the 2-hour warm-up I get this error: “G6 State isn’t currently known. Next connection will update this” I’ve tried going into System Status and doing the Restart Collector and restarting the phone, but that hasn’t worked. I thought I should try doing Stop Sensor, but the …

---

<div class="post-metadata">

**Author:** ![MM2](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mm2/32/3032_2.png) [@MM2](https://forum.fudiabetes.org/u/MM2)\
**Post date:** [August 24, 2019, 2:16am UTC](https://forum.fudiabetes.org/t/tips-on-resetting-transmittor-on-g6-via-android-phone/8160/3 "2019-08-24T02:16:28Z")

</div>

It appears the newer G6 transmitters may be prevented from reset. What is first 3 digits of your G6 transmitter?

> **[Restarting Dexcom G6 sensors – Updated](https://seemycgm.com/2019/08/01/restarting-dexcom-g6-sensors-updated/)**
>
> Awhile ago, I posted a page about how to restart your Dexcom sensors to last beyond 10 days. There’s need for some updated discussion on the topic. For clarity- let’s take one quick asi…

---

<div class="post-metadata">

**Author:** ![jbowler](https://avatars.discourse-cdn.com/v4/letter/j/e274bd/32.png) [@jbowler](https://forum.fudiabetes.org/u/jbowler)\
**Post date:** [August 24, 2019, 2:34am UTC](https://forum.fudiabetes.org/t/tips-on-resetting-transmittor-on-g6-via-android-phone/8160/4 "2019-08-24T02:34:53Z")

</div>

> [@MM2](#):
>
> With xDrip, you can reset trans, but don’t need to because xDrip does not stop reading based on number of days, like the dexcom does.

I have never tried to reset it (I use xDrip+ exclusively), but I do replace before 120 days; I seem to get increasing connect failures over time and this _seems_ to correspond to the drop in the transmitter battery ‘load’ voltage “Voltage B” in xDrip+.

As a result at present my supply problem is the transmitters; I have sensors stacked to the eaves since they just keep on working, but I don’t have a reserve transmitter at this point. Roll on G7; it makes much more sense to replace the battery with the sensor and the electronics are micron scale so that just fits.

John Bowler

---

<div class="post-metadata">

**Author:** ![MM2](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mm2/32/3032_2.png) [@MM2](https://forum.fudiabetes.org/u/MM2)\
**Post date:** [August 24, 2019, 3:00am UTC](https://forum.fudiabetes.org/t/tips-on-resetting-transmittor-on-g6-via-android-phone/8160/5 "2019-08-24T03:00:02Z")

</div>

Dexcom has been making changes, and now has G6 transmitter id starting with 80, 81 and 8G, that behave differently, which may impact those that extend (restart) sensors, or extend (reset) transmitter.

Details in the 8/1/2019 document in link.

---

<div class="post-metadata">

**Author:** ![shaye](https://avatars.discourse-cdn.com/v4/letter/s/b782af/32.png) [@shaye](https://forum.fudiabetes.org/u/shaye)\
**Post date:** [August 24, 2019, 2:00pm UTC](https://forum.fudiabetes.org/t/tips-on-resetting-transmittor-on-g6-via-android-phone/8160/6 "2019-08-24T14:00:18Z")

</div>

Hi & thank you for your reply to my emmail!! The first 3 digits of my transmitter are 80B! What purpose does that serve?

thx again😉

Ms. Shaye

Shaye-B-Promotions & Entertainment

---

<div class="post-metadata">

**Author:** ![MM2](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mm2/32/3032_2.png) [@MM2](https://forum.fudiabetes.org/u/MM2)\
**Post date:** [August 24, 2019, 4:47pm UTC](https://forum.fudiabetes.org/t/tips-on-resetting-transmittor-on-g6-via-android-phone/8160/7 "2019-08-24T16:47:21Z")

</div>

> [@shaye](#):
>
> The first 3 digits of my transmitter are 80B!

This means you have an older transmitter, which can be reset. Newer transmitters have changes that prevent this.

---

<div class="post-metadata">

**Author:** ![jbowler](https://avatars.discourse-cdn.com/v4/letter/j/e274bd/32.png) [@jbowler](https://forum.fudiabetes.org/u/jbowler)\
**Post date:** [August 25, 2019, 3:20am UTC](https://forum.fudiabetes.org/t/tips-on-resetting-transmittor-on-g6-via-android-phone/8160/8 "2019-08-25T03:20:52Z")

</div>

In my case my previous transmitter, from EdgePark, was one of the 80’s, manufactured 2018-09-04, my current transmitter from Walmart is an 81, manufactured 2019-05-14. Both transmitters have an expiry date that is exactly 244 days after the manufacture date; with a day of 8 months. My guess is that they don’t trust the installed battery for more than a year.

---

<div class="post-metadata">

**Author:** ![jbowler](https://avatars.discourse-cdn.com/v4/letter/j/e274bd/32.png) [@jbowler](https://forum.fudiabetes.org/u/jbowler)\
**Post date:** [September 5, 2019, 12:03am UTC](https://forum.fudiabetes.org/t/tips-on-resetting-transmittor-on-g6-via-android-phone/8160/9 "2019-09-05T00:03:18Z")

</div>

I just got an 8G transmitter with my refill (90 days after the 81 transmitter), manufacture date 2019-08-25. I just hope xDrip+ supports this before I have to use it, the idea of going back to the Dexcom receiver definitely does not appeal to me.

---

<div class="post-metadata">

**Author:** ![Thomas](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/thomas/32/1719_2.png) [@Thomas](https://forum.fudiabetes.org/u/Thomas)\
**Post date:** [September 5, 2019, 1:21am UTC](https://forum.fudiabetes.org/t/tips-on-resetting-transmittor-on-g6-via-android-phone/8160/10 "2019-09-05T01:21:44Z")

</div>

> If your Dexcom G6 transmitter’s serial no. is starting with 8G… try nightly build 2019/07/28 or later

Sounds like the xDrip coders are on it. Although the “try” does seem to imply the 8Gxxxx transmitters may still be a work in progress in terms of xDrip support

[https://androidaps.readthedocs.io/en/latest/EN/Configuration/xdrip.html](https://androidaps.readthedocs.io/en/latest/EN/Configuration/xdrip.html)

---

<div class="post-metadata">

**Author:** ![Pbm58](https://avatars.discourse-cdn.com/v4/letter/p/3ec8ea/32.png) [@Pbm58](https://forum.fudiabetes.org/u/Pbm58)\
**Post date:** [April 11, 2020, 2:47am UTC](https://forum.fudiabetes.org/t/tips-on-resetting-transmittor-on-g6-via-android-phone/8160/11 "2020-04-11T02:47:19Z")

</div>

My G6 transmitter expired a few days ago. After reading many forums I installed xDrip+ on my Samsung Galaxy tab, connected to the transmitter and reset. I checked the dates and it showed 108 days remaining. Unpaired fromTab. Now on the iPhone (that I usually use with Dexcom app) I unpaired and deleted connection, rebooted. Tried to set up a new connection to the transmitter with the transmitter ID. Dexcom app did not connect. It appeared my Dexcom account had the info about that transmitter and ‘knew’ that transmitter expired after 90 days. Did anybody face this problem? Any workaround? Thanks.

---

<div class="post-metadata">

**Author:** ![MM2](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mm2/32/3032_2.png) [@MM2](https://forum.fudiabetes.org/u/MM2)\
**Post date:** [April 11, 2020, 5:14am UTC](https://forum.fudiabetes.org/t/tips-on-resetting-transmittor-on-g6-via-android-phone/8160/12 "2020-04-11T05:14:25Z")

</div>

> [@Pbm58](#):
>
> Dexcom app did not connect. It appeared my Dexcom account had the info about that transmitter and ‘knew’ that transmitter expired after 90 days.

You could try updating transmitter id to a different number, then retry. Expect it to fail. The start again as though you had a new transmitter.

Or If you have G6 Receiver, try similar steps, with phone app off.