# The last two G6 sensors have stopped working after 10 days exactly using xDrip+, but didn't used to

**URL:** <https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763>\
**Category:** C/GM\
**Tags:** xdrip\
**Created:** [November 26, 2019, 5:14pm UTC](https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763 "2019-11-26T17:14:48Z")\
**Posts on this page:** 20\
**Page:** 1

<div class="post-metadata">

**Author:** ![Headlands](https://avatars.discourse-cdn.com/v4/letter/h/edb3f5/32.png) [@Headlands](https://forum.fudiabetes.org/u/Headlands)\
**Post date:** [November 26, 2019, 5:14pm UTC](https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763/1 "2019-11-26T17:14:48Z")

</div>

In the past xDrip+ would always keep on going after 10 days, for up to 30 days until the sensor actually died, now it seems like Dexcom has made the sensor die after 10 days so that xDrip says “sensor stopped”. I tried to re-start the sensor from scratch and it failed with the last sensor – am trying again right now with the next one. Has anyone else been having this experience? Is there a way to get around this so that xDrip+ keeps on going like it used to?

---

<div class="post-metadata">

**Author:** ![Chris](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/chris/32/382_2.png) [@Chris](https://forum.fudiabetes.org/u/Chris)\
**Post date:** [November 26, 2019, 5:21pm UTC](https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763/2 "2019-11-26T17:21:08Z")

</div>

The sensor isn’t the issue, the transmitter software is. I am not sure, but it might helps someone troubleshoot if you can post the transmitter firmware number.

---

<div class="post-metadata">

**Author:** ![Headlands](https://avatars.discourse-cdn.com/v4/letter/h/edb3f5/32.png) [@Headlands](https://forum.fudiabetes.org/u/Headlands)\
**Post date:** [November 26, 2019, 5:28pm UTC](https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763/3 "2019-11-26T17:28:11Z")

</div>

Ah, interesting. I only have the transmitter code, 8GJGRH, threw away the box. Is there a way to find the firmware version using xDrip?

---

<div class="post-metadata">

**Author:** ![Headlands](https://avatars.discourse-cdn.com/v4/letter/h/edb3f5/32.png) [@Headlands](https://forum.fudiabetes.org/u/Headlands)\
**Post date:** [November 26, 2019, 5:35pm UTC](https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763/4 "2019-11-26T17:35:15Z")

</div>

Would it work if I re-started/re-entered the transmitter code when re-starting the sensor in Drip?

---

<div class="post-metadata">

**Author:** ![MM2](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mm2/32/3032_2.png) [@MM2](https://forum.fudiabetes.org/u/MM2)\
**Post date:** [November 26, 2019, 5:36pm UTC](https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763/5 "2019-11-26T17:36:18Z")

</div>

> [@Headlands](#):
>
> the firmware version using xDrip?

It is on xDrip status page, and on G6 receiver.

Mine is 1.6.5.25, and can restart. (Trans 81xx)

Some have reported 1.6.5.27 as being harder to restart.

8Gxx transmitters have been reported as harder to restart sensors.

---

<div class="post-metadata">

**Author:** ![Chris](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/chris/32/382_2.png) [@Chris](https://forum.fudiabetes.org/u/Chris)\
**Post date:** [November 26, 2019, 5:36pm UTC](https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763/6 "2019-11-26T17:36:45Z")

</div>

That I am unsure of. At this point, it might be worth trying to restart. I am not an Xdrip+ user, but have followed the threads that have talk about it in the event my son wants to start using it. Any other ideas @docslotnick?

---

<div class="post-metadata">

**Author:** ![Headlands](https://avatars.discourse-cdn.com/v4/letter/h/edb3f5/32.png) [@Headlands](https://forum.fudiabetes.org/u/Headlands)\
**Post date:** [November 26, 2019, 5:40pm UTC](https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763/7 "2019-11-26T17:40:00Z")

</div>

Thanks for the help Chris. Mine is 2.18.2.88.

If anyone has an idea on how to get it going again I would greatly appreciate help. I’ll wait for a few hours before I re-start with a new sensor in case someone knows what I can try.

---

<div class="post-metadata">

**Author:** ![MM2](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mm2/32/3032_2.png) [@MM2](https://forum.fudiabetes.org/u/MM2)\
**Post date:** [November 26, 2019, 5:42pm UTC](https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763/8 "2019-11-26T17:42:46Z")

</div>

Check this post.

> **[G6 restart is possible!](https://forum.tudiabetes.org/t/g6-restart-is-possible/80615)**
>
> This video has the instructions on how to restart G6 sensors that say “No restarts” - I tried it today and it worked! 🙂 Thank you to Type1Tech TipsnTricks! I’d typed up the steps since the video is long: stop sensor on phone & receiver (or...

---

<div class="post-metadata">

**Author:** ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)\
**Post date:** [November 26, 2019, 6:01pm UTC](https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763/9 "2019-11-26T18:01:10Z")

</div>

8Gxxxxx sensors have a hard switch off routine built in. Get around it by snapping out the transmitter (use of two guitar picks seems to work) from the sensor, leave them disconnected for about a half hour, then snap the transmitter back in and start your session.

PIA to be sure. JamOrHam is working on it, but the 8Hxxx transmitters seem to be able to be normally reset, so Dexcom may have relented.

---

<div class="post-metadata">

**Author:** ![Thomas](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/thomas/32/1719_2.png) [@Thomas](https://forum.fudiabetes.org/u/Thomas)\
**Post date:** [November 26, 2019, 6:53pm UTC](https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763/10 "2019-11-26T18:53:56Z")

</div>

> [@docslotnick](#):
>
> (use of two guitar picks seems to work)

Or two test strips.

---

<div class="post-metadata">

**Author:** ![MM2](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mm2/32/3032_2.png) [@MM2](https://forum.fudiabetes.org/u/MM2)\
**Post date:** [November 26, 2019, 9:17pm UTC](https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763/11 "2019-11-26T21:17:09Z")

</div>

> [@Headlands](#):
>
> In the past xDrip+ would always keep on going after 10 days, for up to 30 days until the sensor actually died, now it seems like Dexcom has made the sensor die after 10 days so that xDrip says “sensor stopped”.

Were you using same transmitter for all these ?  
We know there are changes in transmitter, but could also be something dexcom is changing in the sensors.

---

<div class="post-metadata">

**Author:** ![Headlands](https://avatars.discourse-cdn.com/v4/letter/h/edb3f5/32.png) [@Headlands](https://forum.fudiabetes.org/u/Headlands)\
**Post date:** [November 26, 2019, 9:39pm UTC](https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763/12 "2019-11-26T21:39:48Z")

</div>

No, I would use the transmitter for as long as it would last, which would span a few sensors.

I tried removing the transmitter, waiting 30 min, then putting it back in and re-starting the sensor, which worked perfectly. Thank you both for the advice/link.

---

<div class="post-metadata">

**Author:** ![jbowler](https://avatars.discourse-cdn.com/v4/letter/j/e274bd/32.png) [@jbowler](https://forum.fudiabetes.org/u/jbowler)\
**Post date:** [November 27, 2019, 3:31am UTC](https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763/13 "2019-11-27T03:31:00Z")

</div>

> [@docslotnick](#):
>
> 8Hxxx transmitters seem to be able to be normally reset

I received one 8G transmitter, which I haven’t used yet but will at the next sensor because my 81 transmitter is at 97 days and is flaking out (i.e. dropping bluetooth regularly). I just received my next transmitter, which is an 8H.

The logical interpretation is that 8G transmitters have a serious bug and this is fixed in 8H, otherwise the early rev doesn’t make any sense.

---

<div class="post-metadata">

**Author:** ![Thomas](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/thomas/32/1719_2.png) [@Thomas](https://forum.fudiabetes.org/u/Thomas)\
**Post date:** [November 27, 2019, 3:54am UTC](https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763/14 "2019-11-27T03:54:28Z")

</div>

> [@jbowler](#):
>
> The logical interpretation is that 8G transmitters have a serious bug and this is fixed in 8H, otherwise the early rev doesn’t make any sense.

Agreed

---

<div class="post-metadata">

**Author:** ![jbowler](https://avatars.discourse-cdn.com/v4/letter/j/e274bd/32.png) [@jbowler](https://forum.fudiabetes.org/u/jbowler)\
**Post date:** [November 30, 2019, 1:29am UTC](https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763/15 "2019-11-30T01:29:46Z")

</div>

> [@Thomas](#):
>
> > The logical interpretation is that 8G transmitters have a serious bug and this is fixed in 8H, otherwise the early rev doesn’t make any sense.
> 
> Agreed

I had no problems _starting_ my new 8G transmitter however one bug showed up immediately, though I think it has to be a bug in xDrip+.

With the 81 transmitter I would get dropouts and have to turn off bluetooth before xDrip+ would re-sync to the transmitter. xDrip+ then requests what jamorham calls “backsies” from the transmitter; readings from the transmitter’s 3 hour rotating buffer of past blood glucose measures. These “backsies” appear on the display with a faint halo around the yellow dot.

With the 8G transmitter the same thing happens however all the readings, even the ones from the regular successful 5 minute communication, appear with the halo.

I’ve examined the event log and those readings are apparently read fine; I see log messages similar to the following (I’ve reversed the order from the event log, which lists latest first, so the following is in the order the messages were posted):

```
17:12 Status: Got G6 glucose 17:12 
17:12 Checking 3 for backfill requirement
17:12 Got usable glucose data from G5!![sic]

```

There was no backsie request in the 17:12 communication, nevertheless the 17:10 blood glucose reading, along with all the others, has the backsie halo. I think this can only be an xDrip+ bug.

There are also some protocol changes:

```
17:22 Received indication bytes: [* * *]
17:22 Got unknown packet rx: D0002715
17:22 Received indication bytes: D0002715

```

In addition I always see:

`17:22 Speaking slowly`

This seems to be coming from the Ob1G5CollectionService thread, though I’m assuming here that the event log is completely reversed form what most people would think of as the normal order (i.e. the new lines always appear at the top, not the bottom). I’m pretty sure I haven’t seen that on a regular basis before, though I’m also pretty sure I have seen it before.

Dropouts seem to be occurring more often, though that’s just a feeling; I can’t prove or disprove it.

---

<div class="post-metadata">

**Author:** ![Thomas](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/thomas/32/1719_2.png) [@Thomas](https://forum.fudiabetes.org/u/Thomas)\
**Post date:** [November 30, 2019, 1:52am UTC](https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763/16 "2019-11-30T01:52:57Z")

</div>

I wonder if any changes in the 8Gxxxx will cause comm problems with the X2.

We were planning to use an 8Gxxxx for our next transmitter (currently on the 81xxxx) but I am wondering if it might be better to just skip over that one and go right to the 8Hxxxx.

We have come to rely on solid communication between the G6 and the X2.

---

<div class="post-metadata">

**Author:** ![jbowler](https://avatars.discourse-cdn.com/v4/letter/j/e274bd/32.png) [@jbowler](https://forum.fudiabetes.org/u/jbowler)\
**Post date:** [November 30, 2019, 2:47am UTC](https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763/17 "2019-11-30T02:47:53Z")

</div>

> [@Thomas](#):
>
> I am wondering if it might be better to just skip over that one and go right to the 8Hxxxx

If you have the option then I suggest trying the 8G and if it doesn’t work reliably immediately swapping to the 8H. If I had a hoard I would, at this point, just swap to the 8H on the next sensor, but I don’t; I’ve only got one transmitter (the 8H) as backup. Since you are using a different setup and since I suspect the problems are in xDrip+ I would certainly want to try it to see what happens.

I keep coming back to this other fud thread:

> [@Future of Diabetic Engineering](https://forum.fudiabetes.org/t/future-of-diabetic-engineering/8747):
>
> Dexcom and Xdrip+ are great tools for managing Diabetes. In the interview given by Dexcom CEO Kevin Sayer: “You’ve got to understand, medical device companies and hacking, those two concepts don’t go together well. So we were very upset and we were pretty aggressive speaking out that this is not a good thing. “But in all reality, the hackers are Dexcom patients. They purchase product from us, they’re so dedicated to it that they use it and try to make their lives better by developing tools th…

It seems to me that Sayer was reacting to a situation where products not under Dexcom’s control reacted to quite reasonable changes in the G6 communications protocol by doing bad things. Of course I’m just guessing, but I have experience of working for one manufacturer and implementing a piece of perfectly correct software which produced a perfectly valid GIF that terminated another manufacturer’s buggy GIF implementation. After they refused to fix the bug I had no real choice but to produce sub-optimal GIFs for their benefit. It was annoying.

---

<div class="post-metadata">

**Author:** ![emp00](https://avatars.discourse-cdn.com/v4/letter/e/9dc877/32.png) [@emp00](https://forum.fudiabetes.org/u/emp00)\
**Post date:** [November 30, 2019, 4:44pm UTC](https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763/18 "2019-11-30T16:44:17Z")

</div>

> [@docslotnick](#):
>
> JamOrHam is working on it, but the 8Hxxx transmitters seem to be able to be normally reset, so Dexcom may have relented.

@docslotnick: That’s interesting and new to me! Do you have a link/reference with more background information what has changed on 8H transmitters compared to the 8G’s? I’m reading JamOrHam’s gitter forum and must have missed this piece of information. Where have you seen/read/heard this? Thanks for feedback.

---

<div class="post-metadata">

**Author:** ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)\
**Post date:** [November 30, 2019, 6:16pm UTC](https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763/19 "2019-11-30T18:16:25Z")

</div>

@emp00 There was a mention of this on the Gitter boards about a week and a half ago. Haven’t read anything else about it.

---

<div class="post-metadata">

**Author:** ![glitzabetes](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/glitzabetes/32/4287_2.png) [@glitzabetes](https://forum.fudiabetes.org/u/glitzabetes)\
**Post date:** [December 6, 2019, 3:48pm UTC](https://forum.fudiabetes.org/t/the-last-two-g6-sensors-have-stopped-working-after-10-days-exactly-using-xdrip-but-didnt-used-to/8763/20 "2019-12-06T15:48:51Z")

</div>

I haven’t been on FUD much lately but when I was finding my sensor restart trick not working I knew where to come for a work-around, thanks 😁

For what it’s worth, my current transmitter is an 8H and it’s been catching me on my typical restart trick (new sensor no code, 15 minutes, stop sensor, start sensor w/ old code)