# Problem with Jelly Pro, Dexcom G5 and xDrip+

**URL:** <https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642>\
**Category:** C/GM\
**Tags:** xdrip, cgm, small-phone\
**Created:** [November 15, 2017, 12:27am UTC](https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642 "2017-11-15T00:27:16Z")\
**Posts on this page:** 20\
**Page:** 2

<div class="post-metadata">

**Author:** ![Laura](https://avatars.discourse-cdn.com/v4/letter/l/a4c791/32.png) [@Laura](https://forum.fudiabetes.org/u/Laura)\
**Post date:** [November 17, 2017, 1:46pm UTC](https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642/21 "2017-11-17T13:46:33Z")

</div>

I adjusted the high accuracy. (took off GPS) and the battery is at 66% after 5 hours.  
Yesterday it had reached 66% after 4 hours.

---

<div class="post-metadata">

**Author:** ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)\
**Post date:** [November 17, 2017, 3:14pm UTC](https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642/22 "2017-11-17T15:14:58Z")

</div>

@Laura That’s great! Sounds like you’re making some progress.

---

<div class="post-metadata">

**Author:** ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)\
**Post date:** [November 17, 2017, 4:28pm UTC](https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642/23 "2017-11-17T16:28:04Z")

</div>

> [@Laura](#):
>
> the battery is at 66% after 5 hours.

Great!

Well, not great, but better 🙂 That’s good news. I agree with @docslotnick btw on high accuracy. I also turn mine off to increase battery range.

---

<div class="post-metadata">

**Author:** ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)\
**Post date:** [November 17, 2017, 4:29pm UTC](https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642/24 "2017-11-17T16:29:48Z")

</div>

Also, @Thomas’s suggestion is a good one. I have several of the ones he links to, as well as larger models that still fit in a pocket or a small slingbag. I actually bought one for my T1D boy, in case his iphone battery went out (he lost it a fortnigth later…).

---

<div class="post-metadata">

**Author:** ![Thomas](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/thomas/32/1719_2.png) [@Thomas](https://forum.fudiabetes.org/u/Thomas)\
**Post date:** [November 17, 2017, 4:40pm UTC](https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642/25 "2017-11-17T16:40:55Z")

</div>

> [@Michel](#):
>
> he lost it a fortnigth later

Next time get either the pink one or spray paint hunter orange on it before you hand it over.

😄

---

<div class="post-metadata">

**Author:** ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)\
**Post date:** [November 17, 2017, 4:52pm UTC](https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642/26 "2017-11-17T16:52:23Z")

</div>

> [@Thomas](#):
>
> the pink one

Maybe he would lose the pink one ever faster 🙂 You are right though, I tend to get him super bright items for everything important.

---

<div class="post-metadata">

**Author:** ![Laura](https://avatars.discourse-cdn.com/v4/letter/l/a4c791/32.png) [@Laura](https://forum.fudiabetes.org/u/Laura)\
**Post date:** [November 18, 2017, 7:21pm UTC](https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642/27 "2017-11-18T19:21:41Z")

</div>

We got at least 17 hours out of the battery yesterday. There was 7% left. I believe it would have lasted at least another hour or two. That’s good enough for us as it easily surpasses the time she is up and about before bed time. Thank you all for your help. I must save the xDrip+ and Jelly phone settings as backup.

---

<div class="post-metadata">

**Author:** ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)\
**Post date:** [November 18, 2017, 9:10pm UTC](https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642/28 "2017-11-18T21:10:40Z")

</div>

@Laura Actually from looking around at Jelly reviews, 17 hours is really quite good.

Really glad you got this problem solved! It’s hard enough to deal with a young child with D without having to struggle with technology. You are to be congratulated!

---

<div class="post-metadata">

**Author:** ![Laura](https://avatars.discourse-cdn.com/v4/letter/l/a4c791/32.png) [@Laura](https://forum.fudiabetes.org/u/Laura)\
**Post date:** [November 18, 2017, 9:36pm UTC](https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642/29 "2017-11-18T21:36:00Z")

</div>

We are and she is delighted with the phone. I was in contact with a couple of more people on Facebook and they did give good advice in relation to the battery. He said not to drain it completely too often as this is meant to affect the battery performance in the long term. We only need to get from 8am to 8pm so what we are not in that territory.  
Ava has Congenital Hyperinsulinism. Its kind of the opposite of diabetes. She had a subtotal pancreatectomy . However she still has faulty cells pumping out too much insulin causing hypos. They are starting to wear out now bit by bit and her levels are starting to get higher and higher. So with these faulty cells gradually wearing themselves out and with not having any pancreas she will eventually go to the opposite extreme and become diabetic. we are experiencing a lot of high lately. We only started using a CGM last December. We became aware of it kind of by accident. We did question specialists in GOSH in London if we should use a CGM but they were not interested in them as they believed they were not accurate. Looking back our believe is that they were actually not up to speed with the technology and knew nothing about it so they dismissed it. When we showed the new consultant in GOSH he was amazed by it wants every child to have one now. This technology changed our lives. IT allowed her to come off an overnight continuous feed into her tummy which kept her levels from dropping. Before we could not let her out of our sight because we were constantly checking her sugars. We did not sleep a proper nights sleep until we got the CGM.  
Thanks for all your help.

By the way below are the concluded settings for both the xDrip+ App and the Jelly Pro phone:  
Jelly Pro Settings:

1. I made sure the phone software had the latest update installed.
2. I agreed to having Location ON (it is left on continuously)
3. Went into Location settings and changed the setting “High Accuracy” to “Battery Saving” (Use WiFi, Bluetooth or mobile networks to determine location).  
Note: Do not confuse this with Intelligent Battery saving. This is Battery Saving in Location settings only.
4. Agreed to having xDrip request of NOT applying battery optimization
5. Whitelisted xDrip+ in Background task clear
6. In Battery Saving Mode - Battery Profile - My Mode - I have the following on:  
–Mobile networks  
–Bluetooth  
–GPS  
–Vibrate on tap
7. Battery saver set as “Off/Never turn on automatically”
8. Intelligent power saving standby turned off
9. xDrip+ selected in the Duraspeed list of background activities of apps switched on in the list. The only pps running here are the essential apps such as Google Services, phone, messaging etc. I have the screen brightness turned down to the minimum and in the display settings the screen is set to timeout after 15 seconds.
10. In Network Access Control I have everything set as on/enabled.
11. I turned off as many notifications as possible in the settings.
12. I turned off default settings for phone to vibrate for example receiving call etc.
13. I made sure xDrip+ is the only app installed. I had a panic alarm app installed but was wary of this requiring location to be on or it requesting it to be on.

xDrip+ Settings:

1. I downloaded the latest release of xDrip+ from [https://github.com/NightscoutFoundation/xDrip/releases:](https://github.com/NightscoutFoundation/xDrip/releases:) Nightly build 13th Nov 2017.  
Note: I think this may have helped as they said there were adjustments made in the software which should help with connectivity. This seemed to make a big difference in ensuring consistent collection.
2. OB1 ticked in G5 debug settings
3. Authenticate G5 before each read attempt is ticked
4. Ensure no other phones are also acting as collectors. I found out the hard way that the other phone was also trying to connect with the G5 transmitter and not allowing my intended collector Jelly Pro phone to connect with the G5. All other phone that we have present are set as follower phones using either the xDrip+ App (Follower mode) or the Nightwatch App.

---

<div class="post-metadata">

**Author:** ![Thomas](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/thomas/32/1719_2.png) [@Thomas](https://forum.fudiabetes.org/u/Thomas)\
**Post date:** [November 18, 2017, 9:43pm UTC](https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642/30 "2017-11-18T21:43:51Z")

</div>

Could you get the USB battery charger and just plug it into the phone and leave it like that during the day? That would ensure the battery charge becomes a non-issue. Based on reported mAh, the USB charger has about 3-1/2 times the amount of juice that the Jelly has.

---

<div class="post-metadata">

**Author:** ![Laura](https://avatars.discourse-cdn.com/v4/letter/l/a4c791/32.png) [@Laura](https://forum.fudiabetes.org/u/Laura)\
**Post date:** [November 18, 2017, 9:57pm UTC](https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642/31 "2017-11-18T21:57:01Z")

</div>

That would ensure the battery is kept going longer. However the phone’s battery is lasting more than enough time before she goes to bed, when it is charged again(Even the bit longer at weekend)  
Having the battery backup is additional weight again and would kind of defeat the entire purpose of choosing a Jelly Pro phone to act as a collector over a bigger phone which would have a bigger battery and last longer.  
The smaller the better (but with decent battery)! 🙂  
I already have a backup up battery charger that she carries in her school bag should we ever get in a situation were we could get caught (for e.g forget to charge, power outage etc.). I also have one in my bag.

---

<div class="post-metadata">

**Author:** ![Thomas](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/thomas/32/1719_2.png) [@Thomas](https://forum.fudiabetes.org/u/Thomas)\
**Post date:** [November 18, 2017, 9:59pm UTC](https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642/32 "2017-11-18T21:59:32Z")

</div>

I tend to get obsessive with batteries. I try to keep my OCD focused into areas where I can’t do any real harm.  
ha ha ha

---

<div class="post-metadata">

**Author:** ![Laura](https://avatars.discourse-cdn.com/v4/letter/l/a4c791/32.png) [@Laura](https://forum.fudiabetes.org/u/Laura)\
**Post date:** [November 18, 2017, 10:12pm UTC](https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642/33 "2017-11-18T22:12:36Z")

</div>

Any input is appreciated 😄 Every suggestion/idea here is a good one and helps

---

<div class="post-metadata">

**Author:** ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)\
**Post date:** [November 21, 2017, 2:53am UTC](https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642/34 "2017-11-21T02:53:03Z")

</div>

The lesson I am drawing is that, thanks to @docslotnick and to @Laura, we have a good idea of the Jelly Pro’s battery capacity when set up as an xDrip+ collector only. This is a pretty definite result!

---

<div class="post-metadata">

**Author:** ![Qwertypoi](https://avatars.discourse-cdn.com/v4/letter/q/ecae2f/32.png) [@Qwertypoi](https://forum.fudiabetes.org/u/Qwertypoi)\
**Post date:** [September 4, 2018, 8:29am UTC](https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642/35 "2018-09-04T08:29:21Z")

</div>

Can anyone pls summarise what exact settings is nedded for the _JELLY PRO_ xdrip+. I’m lost pls help.

Thanks

---

<div class="post-metadata">

**Author:** ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)\
**Post date:** [September 4, 2018, 6:25pm UTC](https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642/36 "2018-09-04T18:25:55Z")

</div>

@Qwertypoi Are you trying to set up a G5 or G6? There’s a little difference.

---

<div class="post-metadata">

**Author:** ![Qwertypoi](https://avatars.discourse-cdn.com/v4/letter/q/ecae2f/32.png) [@Qwertypoi](https://forum.fudiabetes.org/u/Qwertypoi)\
**Post date:** [September 4, 2018, 9:59pm UTC](https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642/37 "2018-09-04T21:59:14Z")

</div>

Thanks for you reply. G5

---

<div class="post-metadata">

**Author:** ![Qwertypoi](https://avatars.discourse-cdn.com/v4/letter/q/ecae2f/32.png) [@Qwertypoi](https://forum.fudiabetes.org/u/Qwertypoi)\
**Post date:** [September 4, 2018, 10:04pm UTC](https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642/38 "2018-09-04T22:04:10Z")

</div>

![Screenshot_20180904-230240](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/b/b864a4e150c3b342578ed85f4f0884ea10be2b26.jpeg)

Only the above 3 options are checked.

---

<div class="post-metadata">

**Author:** ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)\
**Post date:** [September 4, 2018, 11:00pm UTC](https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642/39 "2018-09-04T23:00:39Z")

</div>

@Qwertypoi. Ok. Let’s start from the beginning.

In xDrip+ long press the the xDrip+ icon on the home screen. That will open an options box. One of the options is “data source wizard” . Select that and now you should see “data source wizard” on your home screen of xDrip+.

Now run the data source wizard and choose G5/G6\>G5. This will apply the most basic settings for the G5 to work. Make sure you copy down your transmitter ID#, you’ll need it.

Insert the sensor and transmitter. Tap the menu and select “start sensor”. And wait for the startup.

On the phone settings make sure you turn on Bluetooth, WiFi and Location. Exclude xDrip+ from any power saving routine on the phone.

You should get the prompt to calibrate in about 2:30 hours.

---

<div class="post-metadata">

**Author:** ![Qwertypoi](https://avatars.discourse-cdn.com/v4/letter/q/ecae2f/32.png) [@Qwertypoi](https://forum.fudiabetes.org/u/Qwertypoi)\
**Post date:** [September 4, 2018, 11:59pm UTC](https://forum.fudiabetes.org/t/problem-with-jelly-pro-dexcom-g5-and-xdrip/2642/40 "2018-09-04T23:59:54Z")

</div>

That’s exactly what I’ve done. I was told the blututh of the jelly pro is not the best, anyone?