I don’t have any experience with Dexcom myself, but I suspect your transmitter is close to failure. The resistance is well above 1000. According to @docslotnick, transmitters die when the resistance is around 1000. He reported his experiences with the life span of Dexcom transmitters in this topic: [How Long Does a G5 Transmitter REALLY Last? \*\*UPDATE-Transmitter Died\*\* UPDATE-New Transmitter \*\*\*UPDATE-2nd Transmitter 😭](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307/19)

Edit: Re-reading the aforementioned topic, apparently some transmitters keep functioning quite well despite a resistance exceeding 1000. Battery voltages dropping below 300 are another indication that the transmitter may fail shortly.

---