# Poor response to insulin from pump

**URL:** <https://forum.fudiabetes.org/t/poor-response-to-insulin-from-pump/13093>\
**Category:** Pump\
**Tags:** tandem-x2\
**Created:** [August 30, 2022, 3:45pm UTC](https://forum.fudiabetes.org/t/poor-response-to-insulin-from-pump/13093 "2022-08-30T15:45:19Z")\
**Posts on this page:** 13\
**Page:** 1

<div class="post-metadata">

**Author:** ![CarlosLuis](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/carlosluis/32/11389_2.png) [@CarlosLuis](https://forum.fudiabetes.org/u/CarlosLuis)\
**Post date:** [August 30, 2022, 3:45pm UTC](https://forum.fudiabetes.org/t/poor-response-to-insulin-from-pump/13093/1 "2022-08-30T15:45:19Z")

</div>

I change my infusion sites every 3 some times 4 days. This has worked well in the 8 months of using a pump. Yesterday was day 2. Everything was fine until a couple of hours after lunch. BG was rising and not responding well to corrections.

Beginning of day 3 - Night time wasn’t too bad, but BG continued to rise before and after eating breakfast, I did 45 minutes of exercise before breakfast which did bring it down, but after eating, rising too high again. I did a syringe correction and it began to drop. I replaced the set and cartridge 5 hours early. Now I am getting a good response.

No sign of leakage or tunneling. There is no scaring at that site. Any ideas?

---

<div class="post-metadata">

**Author:** ![bostrav59](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/bostrav59/32/7281_2.png) [@bostrav59](https://forum.fudiabetes.org/u/bostrav59)\
**Post date:** [August 30, 2022, 4:21pm UTC](https://forum.fudiabetes.org/t/poor-response-to-insulin-from-pump/13093/2 "2022-08-30T16:21:45Z")

</div>

what kind of pump do you have? @CarlosLuis you are a very experienced diabetic so you probably have more ideas and hypotheses than me. Something wasn’t working, and I would suspect the pump - but if it is fine now … hmmm…

---

<div class="post-metadata">

**Author:** ![CarlosLuis](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/carlosluis/32/11389_2.png) [@CarlosLuis](https://forum.fudiabetes.org/u/CarlosLuis)\
**Post date:** [August 30, 2022, 4:33pm UTC](https://forum.fudiabetes.org/t/poor-response-to-insulin-from-pump/13093/3 "2022-08-30T16:33:43Z")

</div>

> [@bostrav59](#):
>
> what kind of pump do you have?

Tandem with C-IQ

---

<div class="post-metadata">

**Author:** ![bostrav59](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/bostrav59/32/7281_2.png) [@bostrav59](https://forum.fudiabetes.org/u/bostrav59)\
**Post date:** [August 30, 2022, 5:09pm UTC](https://forum.fudiabetes.org/t/poor-response-to-insulin-from-pump/13093/4 "2022-08-30T17:09:29Z")

</div>

Ya know - I’m stumped. Was there any change in your diet? Did you notice any insulin leaking? Did the pump’s measure of insulin available look ok? I’m sure you have checked all those things.

One thing I’ve learned – if you don’t fix it, it will happen again. But the corollary to this is, if you haven’t fixed it, when it happens again you’ll be ready for it and will be able to narrow down the problem.

e

---

<div class="post-metadata">

**Author:** ![bkh](https://avatars.discourse-cdn.com/v4/letter/b/da6949/32.png) [@bkh](https://forum.fudiabetes.org/u/bkh)\
**Post date:** [August 30, 2022, 5:16pm UTC](https://forum.fudiabetes.org/t/poor-response-to-insulin-from-pump/13093/5 "2022-08-30T17:16:02Z")

</div>

> [@CarlosLuis](#):
>
> No sign of leakage or tunneling. There is no scaring at that site. Any ideas?

I sometimes experience impaired absorption even though I can’t feel any tissue changes. So I’d say you found a location for an infusion site to keep an eye on, and maybe avoid.

Anyway, you did the “when in doubt change it out” and got the good result.

> [@CarlosLuis](#):
>
> I change my infusion sites every 3 some times 4 days.

Three days is just a “generally safe” amount. Some people routinely get impaired absorption after a day or two. Others can go a week before experiencing the tissue damage. Personally, I wouldn’t go past 3 days because ruined tissue would cause me more grief than the couple minutes to change the infusion site every 3 days. According to the scuttlebut, once the tissue is damaged it takes months with no infusion in that area to recover, and lots of times it never really recovers to the way it was before the damage.

---

<div class="post-metadata">

**Author:** ![CarlosLuis](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/carlosluis/32/11389_2.png) [@CarlosLuis](https://forum.fudiabetes.org/u/CarlosLuis)\
**Post date:** [August 30, 2022, 5:42pm UTC](https://forum.fudiabetes.org/t/poor-response-to-insulin-from-pump/13093/6 "2022-08-30T17:42:53Z")

</div>

> [@bkh](#):
>
> I sometimes experience impaired absorption even though I can’t feel any tissue changes

The spot is about a inch away from a small cluster of broken veins. That may be the problem.

---

<div class="post-metadata">

**Author:** ![CatLady](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/catlady/32/36_2.png) [@CatLady](https://forum.fudiabetes.org/u/CatLady)\
**Post date:** [August 30, 2022, 6:53pm UTC](https://forum.fudiabetes.org/t/poor-response-to-insulin-from-pump/13093/7 "2022-08-30T18:53:17Z")

</div>

> [@CarlosLuis](#):
>
> No sign of leakage or tunneling. There is no scaring at that site.

I had a similar problem with my Pod yesterday. BG kept going up despite corrections and exercise (mopping the kitchen floor usually kicks a high in the butt). So I did an IM correction, changed my pump, and did a meal bolus with a pen. No obvious signs of anything wrong, just weird, sigh…

---

<div class="post-metadata">

**Author:** ![Hammer](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/hammer/32/7809_2.png) [@Hammer](https://forum.fudiabetes.org/u/Hammer)\
**Post date:** [August 30, 2022, 8:14pm UTC](https://forum.fudiabetes.org/t/poor-response-to-insulin-from-pump/13093/8 "2022-08-30T20:14:55Z")

</div>

What insulin are you using?  
I started having that issue with Fiasp in my tandem X2 pump.  
I ended up switching to Lyumjev and haven’t had the issue since then.  
I didn’t have the issue on normal Novolog either, but I really despise using the slower insulin types now.

I am not sure, but I swear something changes with the tandem cartridge and/or infusion sets, and they react more to Fiasp.  
Well, I guess it could have been a change in the Fiasp as well.  
Either way, Fiasp only seems to work about 2 ish days for my in my tandem pump.  
Lyumjev, I can get 5 days without issue if I really need to. (Actually, I reused a cartridge not that long ago, and it lasted around 9 days total! Not what I wanted, but prescriptions kept getting messed up.)

If you have this happen more often, I would start suspecting the equipment rather than then site. If it’s random, then the site could easily cause this as well.

---

<div class="post-metadata">

**Author:** ![CarlosLuis](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/carlosluis/32/11389_2.png) [@CarlosLuis](https://forum.fudiabetes.org/u/CarlosLuis)\
**Post date:** [August 30, 2022, 8:37pm UTC](https://forum.fudiabetes.org/t/poor-response-to-insulin-from-pump/13093/9 "2022-08-30T20:37:03Z")

</div>

> [@Hammer](#):
>
> What insulin are you using?

Humalog, my doc originally tried to prescribe Novolog but stinking insurance wouldn’t cover it. Insurance bean counters know more medical science than doctors, doncha know?

---

<div class="post-metadata">

**Author:** ![CarlosLuis](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/carlosluis/32/11389_2.png) [@CarlosLuis](https://forum.fudiabetes.org/u/CarlosLuis)\
**Post date:** [August 31, 2022, 1:46am UTC](https://forum.fudiabetes.org/t/poor-response-to-insulin-from-pump/13093/10 "2022-08-31T01:46:20Z")

</div>

Update that place where the set was is quite swollen an painful to the touch. There was no pain until an hour after removal.

---

<div class="post-metadata">

**Author:** ![Eric](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/eric/32/9209_2.png) [@Eric](https://forum.fudiabetes.org/u/Eric)\
**Post date:** [August 31, 2022, 3:41am UTC](https://forum.fudiabetes.org/t/poor-response-to-insulin-from-pump/13093/11 "2022-08-31T03:41:25Z")

</div>

> [@CarlosLuis](#):
>
> Humalog, my doc originally tried to prescribe Novolog but stinking insurance wouldn’t cover it.

I have a lot of medical study references that show NovoLog lasts longer in a pump than Humalog.

I can find the post and share it, if you want to try an appeal.

---

<div class="post-metadata">

**Author:** ![MM2](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mm2/32/3032_2.png) [@MM2](https://forum.fudiabetes.org/u/MM2)\
**Post date:** [August 31, 2022, 4:40am UTC](https://forum.fudiabetes.org/t/poor-response-to-insulin-from-pump/13093/12 "2022-08-31T04:40:47Z")

</div>

> [@CarlosLuis](#):
>
> my doc originally tried to prescribe Novolog but stinking insurance wouldn’t cover it.

My insurance also doesn’t cover Novolog, but its my preferred insulin. I use a savings card and get 3 vials for $99. (without using my insurance, cash pay).

> **[MyInsulinRx™ Program Card Registration | Novocare®](https://www.novocare.com/diabetes/help-with-costs/help-with-insulin-costs/myinsulinrx.html)**
>
> Enroll in the MyInsulinRx insulin savings program to pay $35 for a monthly supply of any combination of Novo Nordisk insulin products, if eligible.

---

<div class="post-metadata">

**Author:** ![Hammer](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/hammer/32/7809_2.png) [@Hammer](https://forum.fudiabetes.org/u/Hammer)\
**Post date:** [August 31, 2022, 6:25pm UTC](https://forum.fudiabetes.org/t/poor-response-to-insulin-from-pump/13093/13 "2022-08-31T18:25:56Z")

</div>

I used to use the NovoCare to get my Fiasp for that same $99 cost. Their program actually works well, and it easy to use.