Hmm. I can do better than this in my head (and I failed math in high school).

 ![IMG_1255](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/b/bcd9d4cac8660ba05c4a52edca2da71d344b5b8d.jpg)

---

<div class="post-metadata">

**Author:** ![Jen](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/jen/32/4777_2.png) [@Jen](https://forum.fudiabetes.org/u/Jen)\
**Post date:** [January 2, 2018, 1:44am UTC](https://forum.fudiabetes.org/t/pdm-fails-at-math/3051/2 "2018-01-02T01:44:56Z")

</div>

Wow. Can you adjust it manually?

---

<div class="post-metadata">

**Author:** ![Sam](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/sam/32/10595_2.png) [@Sam](https://forum.fudiabetes.org/u/Sam)\
**Post date:** [January 2, 2018, 2:34am UTC](https://forum.fudiabetes.org/t/pdm-fails-at-math/3051/3 "2018-01-02T02:34:27Z")

</div>

> [@Beacher](#):
>
> Hmm. I can do better than this in my head (and I failed math in high school).

Sorry my friend… but you’re still failing;) jk unless I don’t understand what you’re saying… but 5.5 plus 5.55 does equal 11.05

---

<div class="post-metadata">

**Author:** ![TiaG](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/tiag/32/594_2.png) [@TiaG](https://forum.fudiabetes.org/u/TiaG)\
**Post date:** [January 2, 2018, 2:37am UTC](https://forum.fudiabetes.org/t/pdm-fails-at-math/3051/4 "2018-01-02T02:37:54Z")

</div>

I think the error is the 40/60 split… 40% of 11 is 4.4, 60% is 6.6

---

<div class="post-metadata">

**Author:** ![Eric](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/eric/32/9209_2.png) [@Eric](https://forum.fudiabetes.org/u/Eric)\
**Post date:** [January 2, 2018, 2:46am UTC](https://forum.fudiabetes.org/t/pdm-fails-at-math/3051/5 "2018-01-02T02:46:13Z")

</div>

The extended bolus display is better as units than it is for percentage. The delivery will give you exactly what it shows, 5.50 and 5.55 units. But if it’s set for units, the display will always make more sense.

So you will get 5.50 and 5.55. But just switch the display to units and you will be better off.

BTW, the reason yours is having a problem with the math is because it’s a Canadian PDM. 🤣

 ![image](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/8/855fb6bc198badec6113adf401179a5666f46474.jpg)

---

<div class="post-metadata">

**Author:** ![Beacher](https://avatars.discourse-cdn.com/v4/letter/b/a6a055/32.png) [@Beacher](https://forum.fudiabetes.org/u/Beacher)\
**Post date:** [January 2, 2018, 3:09am UTC](https://forum.fudiabetes.org/t/pdm-fails-at-math/3051/7 "2018-01-02T03:09:20Z")

</div>

> [@Eric](#):
>
> But if it’s set for units, the display will always make more sense.

So it does. Thank you, @Eric. I guess I have to do a separate calculation either way and then tell the PDM either units or percentage (though it would be nice if it would _accurately figure it out for me_), but why would the two settings give different numbers? Makes me worry what else the PDM doesn’t calculate accurately …

---

<div class="post-metadata">

**Author:** ![Eric](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/eric/32/9209_2.png) [@Eric](https://forum.fudiabetes.org/u/Eric)\
**Post date:** [January 2, 2018, 3:20am UTC](https://forum.fudiabetes.org/t/pdm-fails-at-math/3051/8 "2018-01-02T03:20:38Z")

</div>

I guess they give you a choice because people do the calcs differently.

There are better ways do do things for everyone. Like for temp basal, % makes more sense for me. The reason is - if you use units, it will stay with that number throughout the length of the temp basal. If your normal basal changes amounts, do you want your temp basal to stay with the normal fluctuations or stay flat? For me, I want my temp basal to follow the same up and down, but just at a higher or lower amount.

So it’s just a matter of trying all the options.

Yes, there are a few quirks I’ve noticed with math. One of them is with temp basal. They round things weirdly because of the 0.05 unit increment. But rest assured, it always delivers what it should. I’ve counted clicks through every possible scenario. I’m comfortable that regardless of any display anomaly, it’s delivering correctly.

---

<div class="post-metadata">

**Author:** ![Beacher](https://avatars.discourse-cdn.com/v4/letter/b/a6a055/32.png) [@Beacher](https://forum.fudiabetes.org/u/Beacher)\
**Post date:** [January 2, 2018, 1:56pm UTC](https://forum.fudiabetes.org/t/pdm-fails-at-math/3051/9 "2018-01-02T13:56:05Z")

</div>

> [@Eric](#):
>
> But rest assured, it always delivers what it should.

As far as the total dose goes, yes. But as far as the dose delivered in the first segment, no, regardless whether it’s set for units or percentage. I don’t particularly care whether it’s 30% or 35% or 40% or whatever (unless I’m following a formula that I know works for something like pizza). For several reasons,\* I don’t want to deliver more than 5U up front, so to be on the safe side I aim for 4.5U when choosing my percentage. But using the units setting this morning, when I said deliver 4.50 units first and the rest over an hour, the “Delivering bolus” screen said it was giving me 5.10 units up front. A bit of a nuisance.

I only spotted this problem a few nights ago, when a 30% up-front dose was significantly higher than the 70% extended, as though the PDM had got the two reversed. Wish I’d had the camera then.

\*To prevent tunneling, and also, since I’ve been doing this, I haven’t had any “third day problem” highs.

---

<div class="post-metadata">

**Author:** ![Lisa](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/lisa/32/2701_2.png) [@Lisa](https://forum.fudiabetes.org/u/Lisa)\
**Post date:** [January 2, 2018, 3:13pm UTC](https://forum.fudiabetes.org/t/pdm-fails-at-math/3051/10 "2018-01-02T15:13:27Z")

</div>

@Beacher have you tried calling Omnipod about this? I use extended boluses often and I’ve never noticed this. I just looked through my history and all the calcs seem to be correct.

---

<div class="post-metadata">

**Author:** ![Eric](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/eric/32/9209_2.png) [@Eric](https://forum.fudiabetes.org/u/Eric)\
**Post date:** [January 2, 2018, 8:17pm UTC](https://forum.fudiabetes.org/t/pdm-fails-at-math/3051/11 "2018-01-02T20:17:40Z")

</div>

> [@Beacher](#):
>
> I only spotted this problem a few nights ago, when a 30% up-front dose was significantly higher than the 70% extended, as though the PDM had got the two reversed. Wish I’d had the camera then.

I am still curious about this. I would like to try and repro this and see if I can get the same kind of wacky numbers.

I can setup my PDM to use % for extended bolus instead of units, and also change my numbers to be like yours and do the same thing and see if it looks like yours.

Do you have carb number, BG number, and IC number you want to give me to try?

---

<div class="post-metadata">

**Author:** ![Beacher](https://avatars.discourse-cdn.com/v4/letter/b/a6a055/32.png) [@Beacher](https://forum.fudiabetes.org/u/Beacher)\
**Post date:** [January 3, 2018, 7:30pm UTC](https://forum.fudiabetes.org/t/pdm-fails-at-math/3051/12 "2018-01-03T19:30:41Z")

</div>

Thanks, Eric. I think this is the first one that I noticed was reversed (I’ve re-entered a whole bunch of meals, but this one looks familiar):  
BG 9.8, carb 47, IC 1:7  
Suggested bolus was 8.95, to which I added 1U. I wanted 4.5 up front, so 4.5/9.95 is 45%. I entered that and the PDM said 45% = 5.75 and 55% = 4.20. Changing the duration of the extended portion doesn’t change the numbers.

The numbers that resulted in what I took the above photo of: BG 9.3, carb 65, IC 1:7  
Suggested bolus 11.05 (11.35 minus 0.30 IOB).  
When I put those numbers in today, I get 40% = 5.65 and 60% is 5.40. The only thing different is the IOB, which would change the total dose, not the percent breakdown.

Here’s one more, from this morning: BG 9.8, carb 50, IC 1:5  
Suggested bolus 12.35, to which I add 0.5  
4.5U is 35%, but the PDM ends up saying 35% = 6.00 and 65% is 6.85

Anyway, I’ll be interested to see what your PDM says. And just to note, I’m not normally in the 9’s before a meal. Maybe this is the PDM’s punishment for being high.

---

<div class="post-metadata">

**Author:** ![Eric](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/eric/32/9209_2.png) [@Eric](https://forum.fudiabetes.org/u/Eric)\
**Post date:** [January 3, 2018, 7:48pm UTC](https://forum.fudiabetes.org/t/pdm-fails-at-math/3051/13 "2018-01-03T19:48:31Z")

</div>

What about target BG? I just put in something bogus to try this, not sure if that makes a difference…

And also correction factor? Maybe that messes with it some?

---

<div class="post-metadata">

**Author:** ![Beacher](https://avatars.discourse-cdn.com/v4/letter/b/a6a055/32.png) [@Beacher](https://forum.fudiabetes.org/u/Beacher)\
**Post date:** [January 3, 2018, 8:05pm UTC](https://forum.fudiabetes.org/t/pdm-fails-at-math/3051/14 "2018-01-03T20:05:28Z")

</div>

Target 5.5. CorrF 1.9 in the first example, 1.8 in the second and third.

---

<div class="post-metadata">

**Author:** ![ClaudnDaye](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/claudndaye/32/5_2.png) [@ClaudnDaye](https://forum.fudiabetes.org/u/ClaudnDaye)\
**Post date:** [January 4, 2018, 4:09am UTC](https://forum.fudiabetes.org/t/pdm-fails-at-math/3051/15 "2018-01-04T04:09:38Z")

</div>

5 posts were split to a new topic: [Max Topic Replies Recommendation](https://forum.fudiabetes.org/t/max-topic-replies-recommendation/3084)

---

<div class="post-metadata">

**Author:** ![Eric](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/eric/32/9209_2.png) [@Eric](https://forum.fudiabetes.org/u/Eric)\
**Post date:** [January 3, 2018, 8:45pm UTC](https://forum.fudiabetes.org/t/pdm-fails-at-math/3051/16 "2018-01-03T20:45:53Z")

</div>

@Beacher

**Here is your answer.**

Correction bolus can’t be extended. Only meal bolus can be extended.

So it matters on the BG number you enter because that part (the correction) will not be extended, so the percentage doesn’t include that.

Here are some examples I did **with different BG numbers** just to demonstrate the concept. I kept carbs the same.

![image](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/2/2ed4e0b2073c22b6c9e1d1efecb43e5a259ccee2.jpg)

![image](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/7/7827b54b0d7e5ede0e566da55791bba9e09e2973.jpg)

![image](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/c/cd78069365ee9411f49f58ee00a05a3e46f12ce3.jpg)

* * *

A simple example. Suppose you have a BG that requires a 1.0 unit correction. And you are eating carbs that need 1.0 units. That dose total would be 2.0 units.

If you do 50% now and 50% extended, you might think 50% of 2.0 units. But that’s not right. Since 1.0 units is correction, that is not included.  
So it would be:  
2.0 (total) - 1.0 (correction) = 1.0 which is the only part that is extended at 50%.

Make sense?  
😃

* * *

EDIT  
I changed it back to units and this screen makes the math more obvious:

![image](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/8/8543117b35a53e98c86477209793d586e88d7db4.jpg)

---

<div class="post-metadata">

**Author:** ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)\
**Post date:** [January 4, 2018, 4:07am UTC](https://forum.fudiabetes.org/t/pdm-fails-at-math/3051/21 "2018-01-04T04:07:30Z")

</div>

> [@Eric](#):
>
> Correction bolus can’t be extended. Only meal bolus can be extended.
> 
> So it matters on the BG number you enter because that part (the correction) will not be extended, so the percentage doesn’t include that.

Interesting. We never use a BG correction calculated by the PDM, so this may be why we have never seen this behavior.

---

<div class="post-metadata">

**Author:** ![ClaudnDaye](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/claudndaye/32/5_2.png) [@ClaudnDaye](https://forum.fudiabetes.org/u/ClaudnDaye)\
**Post date:** [January 4, 2018, 4:11am UTC](https://forum.fudiabetes.org/t/pdm-fails-at-math/3051/22 "2018-01-04T04:11:12Z")

</div>

A post was merged into an existing topic: [Max Topic Replies Recommendation](https://forum.fudiabetes.org/t/max-topic-replies-recommendation/3084)

---

<div class="post-metadata">

**Author:** ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)\
**Post date:** [January 4, 2018, 5:47am UTC](https://forum.fudiabetes.org/t/pdm-fails-at-math/3051/23 "2018-01-04T05:47:24Z")

</div>

11 posts were split to a new topic: [Omnipod: auto-calc by PDM or manual bolus calculations?](https://forum.fudiabetes.org/t/omnipod-auto-calc-by-pdm-or-manual-bolus-calculations/3088)

---

<div class="post-metadata">

**Author:** ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)\
**Post date:** [January 4, 2018, 5:02am UTC](https://forum.fudiabetes.org/t/pdm-fails-at-math/3051/29 "2018-01-04T05:02:28Z")

</div>

3 posts were merged into an existing topic: [Do you use preset bolus numbers in your Omnipod PDM?](https://forum.fudiabetes.org/t/do-you-use-preset-bolus-numbers-in-your-omnipod-pdm/3085)

---

<div class="post-metadata">

**Author:** ![Beacher](https://avatars.discourse-cdn.com/v4/letter/b/a6a055/32.png) [@Beacher](https://forum.fudiabetes.org/u/Beacher)\
**Post date:** [January 4, 2018, 2:24pm UTC](https://forum.fudiabetes.org/t/pdm-fails-at-math/3051/30 "2018-01-04T14:24:33Z")

</div>

Definitely makes sense! And it explains why I’ve only noticed the disparities when I’m running higher, not when I’m starting out near target – because then there’s no (or not significant) correction involved.

Even in the units setting, you have to account for the correction, because if you say deliver 4.50U, the final screen might have changed that to 5.40, more than I want to deliver up front. It might be easier to stick with the percentage setting and just keep going back to lower the percentage until I’m near my 4.50 limit.

Thanks again, @Eric, for making it clear!