# Pain when inserting dexcom

**URL:** <https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095>\
**Category:** C/GM\
**Tags:** dexcom\
**Created:** [November 25, 2021, 4:23pm UTC](https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095 "2021-11-25T16:23:41Z")\
**Posts on this page:** 20\
**Page:** 1

<div class="post-metadata">

**Author:** ![Mariethm](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mariethm/32/11739_2.png) [@Mariethm](https://forum.fudiabetes.org/u/Mariethm)\
**Post date:** [November 25, 2021, 4:23pm UTC](https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095/1 "2021-11-25T16:23:41Z")

</div>

For the last couple of months, I’ve been experiencing lasting pain with the cgm. Often a lot of bleeding as well. The first few times, I thought it was my placement, but I have tried everywhere! Sometimes it only hurts for the first few hours, sometimes only when flexing muscle near it. (e.g., if its on my calf or thigh, walking hurt.) And sometimes I end up taking it off after a day of extreme discomfort. The insertion makes me wince as well, when it hasn’t since the first few times. Just wondering if anyone has experienced anything like this…I’m thinking of just quitting it after I use up my stockpile, because its beyond a small annoyance.

---

<div class="post-metadata">

**Author:** ![Chris](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/chris/32/382_2.png) [@Chris](https://forum.fudiabetes.org/u/Chris)\
**Post date:** [November 26, 2021, 3:24pm UTC](https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095/2 "2021-11-26T15:24:25Z")

</div>

My son has experienced this a couple of times out of 180 or so insertions, but certainly no pattern, and not many times in a row. That would be awful. If you don’t find a solution, maybe the Libre would work for you without pain. If you add an external transmitter you can get something similar to a Dex.

---

<div class="post-metadata">

**Author:** ![TomH](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/tomh/32/9063_2.png) [@TomH](https://forum.fudiabetes.org/u/TomH)\
**Post date:** [November 26, 2021, 3:41pm UTC](https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095/3 "2021-11-26T15:41:37Z")

</div>

I’ve had two of these experiences over the last year. One in my left inner arm that was at least a continuous ache/at times painful from the outset but endured for the 10 days. The other was on my right inner thigh, bloody on insertion and so much pain/irritation I took it out on day 3. Have used both locations again with no problems noted. @Chris has the experience factor to follow if your’s continues to be a problem…

---

<div class="post-metadata">

**Author:** ![TiaG](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/tiag/32/594_2.png) [@TiaG](https://forum.fudiabetes.org/u/TiaG)\
**Post date:** [November 26, 2021, 5:08pm UTC](https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095/4 "2021-11-26T17:08:01Z")

</div>

My son _always_ has pain with Dexcom, and almost always significant bleeding.

What we have found helps a bit:

1. Rubbing in lidocaine cream an hour before insertion, then slathering a generous amount on top, and covering with saran wrap for an hour, till the area is completely white meaning it’s numbed.
2. “Shallow” insertion. For that, you place the sticker on the skin, then pull up on the inserter with one hand while keeping another finger on the sticker underneath to keep it adhered. So instead of the plastic inserter lying flush with the skin and sticker, there’s an air gap the shape of a small triangle between the inserter and the sticker. When you press the button, the electrode goes in at a more shallow angle, and so is more likely to stay in interstitial tissue and not bleed as much. I’m not sure if this is the case with you, but my son is a little skinny and he’s small, so I am guessing the pain has to do with the electrode hitting muscle rather than chub. Making it go in at a shallower angle seems to help a bit with this.

Even with these steps it usually hurts my son for a day maybe two, and then it subsides. We have considered switching to Libre as a result, but because he’s on the Tslim with Control-IQ, and because Libre doesn’t provide real-time remote monitoring, we’re toughing it out. I am praying the G7 is an easier insertion with less pain because it’s so awful for him.

You can try other locations on the body as well, but I suspect that you already have.

Good luck!

---

<div class="post-metadata">

**Author:** ![Mariethm](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mariethm/32/11739_2.png) [@Mariethm](https://forum.fudiabetes.org/u/Mariethm)\
**Post date:** [November 26, 2021, 6:30pm UTC](https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095/5 "2021-11-26T18:30:56Z")

</div>

I’ll try the shallow insertion next session. Unfortunately I have more chub than I should, so I doubt that’s a factor.  
Thank you for the suggestion!

---

<div class="post-metadata">

**Author:** ![CatLady](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/catlady/32/36_2.png) [@CatLady](https://forum.fudiabetes.org/u/CatLady)\
**Post date:** [November 26, 2021, 6:42pm UTC](https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095/6 "2021-11-26T18:42:45Z")

</div>

You might try an ice pack on the site before applying the inserter to reduce bleeding and pain.

---

<div class="post-metadata">

**Author:** ![Mariethm](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mariethm/32/11739_2.png) [@Mariethm](https://forum.fudiabetes.org/u/Mariethm)\
**Post date:** [December 1, 2021, 12:19am UTC](https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095/7 "2021-12-01T00:19:36Z")

</div>

I tried the shallow insertion this time. Doesn’t hurt at all! Hopefully it continues to be painless…  
🙂

---

<div class="post-metadata">

**Author:** ![TiaG](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/tiag/32/594_2.png) [@TiaG](https://forum.fudiabetes.org/u/TiaG)\
**Post date:** [December 1, 2021, 2:57am UTC](https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095/8 "2021-12-01T02:57:27Z")

</div>

Wow, so glad to hear this! Hope it’s a trend and not a fluke 🙂

---

<div class="post-metadata">

**Author:** ![Hammer](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/hammer/32/7809_2.png) [@Hammer](https://forum.fudiabetes.org/u/Hammer)\
**Post date:** [December 2, 2021, 1:00am UTC](https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095/9 "2021-12-02T01:00:09Z")

</div>

The shallow method works great for me. And I have plenty of ‘chub’ where I insert.

---

<div class="post-metadata">

**Author:** ![irrational\_John](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/irrational_john/32/9006_2.png) [@irrational\_John](https://forum.fudiabetes.org/u/irrational_John)\
**Post date:** [December 2, 2021, 5:20pm UTC](https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095/11 "2021-12-02T17:20:15Z")

</div>

> [@PP777](#):
>
> maybe just not press so hard

That approach seems most reasonable to me.

I am not sure why someone would want to press the applicator “hard”. Since your skin, and thus the sensor probe, is just going to rebound/withdraw when the pressure is released, I don’t see what benefit would derive from pushing harder than necessary to hold the applicator in place when inserting. 😕

---

<div class="post-metadata">

**Author:** ![irrational\_John](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/irrational_john/32/9006_2.png) [@irrational\_John](https://forum.fudiabetes.org/u/irrational_John)\
**Post date:** [December 2, 2021, 9:47pm UTC](https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095/13 "2021-12-02T21:47:13Z")

</div>

> [@PP777](#):
>
> IDK why I do it this way either!

I press down relatively hard _after_ I have inserted my Tandem Autosoft 90 infusion sets, but my intent is completely different in that context. I press down on my infusion sets to attempt to make sure the adhesive is holding before I pull out the insertion needle by removing the inserter, I am always afraid the infusion set will pull off when I remove the insertion needle from Autosoft 90s.

Do I “like” the Autosoft 90 sets? Nope. Not really. But what ya gonna do? That seems to be the available set closest to what I was used to from my decade before switching to Tandem. 🙄

While I am pressing hard though I have also “pinched up” the fat at the insertion site. I do this to try to avoid pushing the insertion needle too far in and possibly getting close to or hitting muscle. I never have that concern with the Dexcom applicator. I don’t worry about adhesive with the G6 sensor until after I have inserted. (After inserting I run a finger around the G6 adhesive ring at least 3 times in each direction to try to make sure it is adhering well.)

Overall I am happy with the Dexcom G6 applicator. The biggest problem I have with it is the horrifying amount of plastic waste using it generates. I hope that is at least somewhat reduced when the G6 is replaced by G7. A reusable applicator would be even better, but somehow throwing things away seems to be what we do with everything more and more.

---

<div class="post-metadata">

**Author:** ![Mariethm](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mariethm/32/11739_2.png) [@Mariethm](https://forum.fudiabetes.org/u/Mariethm)\
**Post date:** [December 2, 2021, 10:25pm UTC](https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095/15 "2021-12-02T22:25:48Z")

</div>

Why would you need to put the applicator in a Sharps container? It has no needles, its just an insertion mechanism. Honestly, I’m curious…

---

<div class="post-metadata">

**Author:** ![Chris](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/chris/32/382_2.png) [@Chris](https://forum.fudiabetes.org/u/Chris)\
**Post date:** [December 2, 2021, 11:04pm UTC](https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095/16 "2021-12-02T23:04:00Z")

</div>

Well it does have a needle, it is just inside the mechanism. With that being said, my son doesn’t put it in the sharps container, he just chucks it since the needle is safely inside and not exposed.

---

<div class="post-metadata">

**Author:** ![irrational\_John](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/irrational_john/32/9006_2.png) [@irrational\_John](https://forum.fudiabetes.org/u/irrational_John)\
**Post date:** [December 2, 2021, 11:14pm UTC](https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095/17 "2021-12-02T23:14:50Z")

</div>

> [@Mariethm](#):
>
> Why would you need to put the applicator in a Sharps container? It has no needles, its just an insertion mechanism.

Actually, the G6 applicator **is** a “Sharps container”. The applicator does come with an insertion needle. As part of the insertion, the needle is automagically retracted back into the body of the applicator. You then have the option of throwing the entire applicator into the trash where is likely to only hurt the environment.

Another, more tedious option is to disassemble the applicator and remove the needle and putting that into a sharps container. Depends on how horrendous an environmental insult you consider the applicator to be, I guess.

Below is the video I post for reference when I raise the possibility of disassembling the G6 applicator prior to disposing of its parts.

[![](https://img.youtube.com/vi/M0Q6jLpgqX4/maxresdefault.jpg "Disassembly of Dexcom G6 inserter / applicator") ](https://www.youtube.com/watch?v=M0Q6jLpgqX4)

---

<div class="post-metadata">

**Author:** ![Mariethm](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mariethm/32/11739_2.png) [@Mariethm](https://forum.fudiabetes.org/u/Mariethm)\
**Post date:** [December 2, 2021, 11:28pm UTC](https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095/18 "2021-12-02T23:28:20Z")

</div>

@Chris @irrational_John  
Thank you for the correction. When I was shown how to use it, I was told by the Dexcom guy that it did not need to be put in a Sharps container.  
But, honestly, I’m not going to start breaking the whole thing down in order to put one tiny piece in a different container, and the rest in the trash…

---

<div class="post-metadata">

**Author:** ![Chris](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/chris/32/382_2.png) [@Chris](https://forum.fudiabetes.org/u/Chris)\
**Post date:** [December 3, 2021, 3:48am UTC](https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095/19 "2021-12-03T03:48:24Z")

</div>

> [@Mariethm](#):
>
> I was told by the Dexcom guy that it did not need to be put in a Sharps container.

And I think your Dexcom guy is correct.

---

<div class="post-metadata">

**Author:** ![TiaG](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/tiag/32/594_2.png) [@TiaG](https://forum.fudiabetes.org/u/TiaG)\
**Post date:** [December 3, 2021, 4:41am UTC](https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095/20 "2021-12-03T04:41:48Z")

</div>

The whole point of the sharps container is to protect the trash collector so they don’t accidentally poke themselves on biohazard. It makes it less safe, IMO, to take apart the housing to expose the needle, and then put it in a sharps container that could easily be opened…  
So I’m with you, just chuck in in the trash.

---

<div class="post-metadata">

**Author:** ![irrational\_John](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/irrational_john/32/9006_2.png) [@irrational\_John](https://forum.fudiabetes.org/u/irrational_John)\
**Post date:** [December 3, 2021, 5:44am UTC](https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095/21 "2021-12-03T05:44:46Z")

</div>

> [@TiaG](#):
>
> It makes it less safe, IMO, to take apart the housing to expose the needle, and then put it in a sharps container that could easily be opened…

Only if you throw your sharps container away in the trash. For what it’s worth, I think technically you are **not** supposed to do that. It’s supposed to be returned to someplace that can dispose of it as medical waste.

---

<div class="post-metadata">

**Author:** ![Chris](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/chris/32/382_2.png) [@Chris](https://forum.fudiabetes.org/u/Chris)\
**Post date:** [December 3, 2021, 5:53am UTC](https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095/22 "2021-12-03T05:53:11Z")

</div>

Most residential trash haulers will say that you can throw your sharps in a properly labelled container in the trash. The biohazard disposal is for things that are really contaminated. You can certainly dispose of your sharps using a biohazard disposal pathway, but I don’t think it is required in most municipalities and adds cost.

---

<div class="post-metadata">

**Author:** ![ClaudnDaye](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/claudndaye/32/5_2.png) [@ClaudnDaye](https://forum.fudiabetes.org/u/ClaudnDaye)\
**Post date:** [December 3, 2021, 3:25pm UTC](https://forum.fudiabetes.org/t/pain-when-inserting-dexcom/12095/23 "2021-12-03T15:25:52Z")

</div>

Each state has a process for throwing sharps away in the regular trash can. We have a process we follow in VA and it’s fully correct to use it.