# OneTouch UltraMini vs Contour Next (One?): Lab Draw Challenge 2019

**URL:** https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691
**Category:** Type 1 & LADA
**Tags:** glucose-meter
**Created:** [February 19, 2019, 4:07pm UTC](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691 "2019-02-19T16:07:39Z")
**Posts on this page:** 20
**Page:** 1

<div class="post-metadata">

### Author: ![Nickyghaleb](https://avatars.discourse-cdn.com/v4/letter/n/5f8ce5/32.png) [@Nickyghaleb](https://forum.fudiabetes.org/u/Nickyghaleb)
#### Post date: [February 19, 2019, 4:07pm UTC](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691/1 "2019-02-19T16:07:39Z")

</div>

Someone’s going down here, and I can’t wait to see who… not that I’m not pretty sure already…

The meter values:  
Test 1:  
OneTouch: 114  
Contour: 70

Test 2:  
OneTouch: 107  
Contour: 75

Test 3:  
OneTouch: 114  
Contour: 70

Extras (sensor values):  
G6: 65  
Guardian: 71 (go ahead, Guardian!)  
Libre: 74

Is it bad if I drop a little _One of these things is not like the others_ allusion?? Funny or too controversial?? I’m trying to be funny…

I’ll update when I get my lab result for anyone who’s interested…

Oh, and I’m going home to do a CVS home a1c just to see how that lines up. Too much excitement for a Tuesday morning…

---

<div class="post-metadata">

### Author: ![Chris](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/chris/32/382_2.png) [@Chris](https://forum.fudiabetes.org/u/Chris)
#### Post date: [February 19, 2019, 4:50pm UTC](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691/2 "2019-02-19T16:50:31Z")

</div>

Do you really wear all three sensors at the same time? What kind of cyborg are you? btw, my guess is that your blood draw glucose will come back at 73

---

<div class="post-metadata">

### Author: ![TiaG](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/tiag/32/594_2.png) [@TiaG](https://forum.fudiabetes.org/u/TiaG)
#### Post date: [February 19, 2019, 5:09pm UTC](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691/3 "2019-02-19T17:09:11Z")

</div>

The OneTouch always gave me super crappy readings. I’d trust the contour much more.

---

<div class="post-metadata">

### Author: ![Nickyghaleb](https://avatars.discourse-cdn.com/v4/letter/n/5f8ce5/32.png) [@Nickyghaleb](https://forum.fudiabetes.org/u/Nickyghaleb)
#### Post date: [February 19, 2019, 5:37pm UTC](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691/4 "2019-02-19T17:37:36Z")

</div>

> [@Chris](#):
>
> Do you really wear all three sensors at the same time?

I’ve been known to roll like that, yes. 😁 Actually, I put back in the Guardian only to check out the updated transmitter (word on the street is it’s a dream machine), and I put the Libre on for fun.

> [@Chris](#):
>
> btw, my guess is that your blood draw glucose will come back at 73

1. I think that might be a little shy… so 68-69. My Contour is usually a hair high.

I’m gonna be mad if it comes back 73… 🤔

---

<div class="post-metadata">

### Author: ![Nickyghaleb](https://avatars.discourse-cdn.com/v4/letter/n/5f8ce5/32.png) [@Nickyghaleb](https://forum.fudiabetes.org/u/Nickyghaleb)
#### Post date: [February 19, 2019, 5:39pm UTC](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691/5 "2019-02-19T17:39:55Z")

</div>

> [@TiaG](#):
>
> The OneTouch always gave me super crappy readings. I’d trust the contour much more.

I know there are people who get good results with it, but I’m definitely not one. Even with that being true, three scores like that… that’s insane.

---

<div class="post-metadata">

### Author: ![Chris](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/chris/32/382_2.png) [@Chris](https://forum.fudiabetes.org/u/Chris)
#### Post date: [February 19, 2019, 5:57pm UTC](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691/6 "2019-02-19T17:57:06Z")

</div>

> [@Nickyghaleb](#):
>
> (word on the street is it’s a dream machine)

Is it?

---

<div class="post-metadata">

### Author: ![mike\_g](https://avatars.discourse-cdn.com/v4/letter/m/34f0e0/32.png) [@mike\_g](https://forum.fudiabetes.org/u/mike_g)
#### Post date: [February 19, 2019, 7:48pm UTC](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691/7 "2019-02-19T19:48:21Z")

</div>

> [@Nickyghaleb](#):
>
> OneTouch: 114

OneTouch Verio?

---

<div class="post-metadata">

### Author: ![Nickyghaleb](https://avatars.discourse-cdn.com/v4/letter/n/5f8ce5/32.png) [@Nickyghaleb](https://forum.fudiabetes.org/u/Nickyghaleb)
#### Post date: [February 19, 2019, 8:14pm UTC](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691/8 "2019-02-19T20:14:54Z")

</div>

OneTouch UltraMini

---

<div class="post-metadata">

### Author: ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)
#### Post date: [February 19, 2019, 10:20pm UTC](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691/9 "2019-02-19T22:20:23Z")

</div>

Oh, @Nickyghaleb, these tests show you have become so totally FUD!

---

<div class="post-metadata">

### Author: ![mike\_g](https://avatars.discourse-cdn.com/v4/letter/m/34f0e0/32.png) [@mike\_g](https://forum.fudiabetes.org/u/mike_g)
#### Post date: [February 20, 2019, 2:44am UTC](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691/10 "2019-02-20T02:44:26Z")

</div>

> [@Nickyghaleb](#):
>
> OneTouch UltraMini

Darn… Was hoping it was a verio, so would add more support to my verio joke…"seems as though "verio"is a good name, since it seems to “verio” all over the place…

---

<div class="post-metadata">

### Author: ![Aaron](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/aaron/32/1454_2.png) [@Aaron](https://forum.fudiabetes.org/u/Aaron)
#### Post date: [February 20, 2019, 9:42pm UTC](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691/11 "2019-02-20T21:42:53Z")

</div>

> [@mike\_g](#):
>
> "seems as though "verio"is a good name, since it seems to “verio” all over the place…

😆

@Nickyghaleb - the suspense is killing me…

---

<div class="post-metadata">

### Author: ![Trying](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/trying/32/4078_2.png) [@Trying](https://forum.fudiabetes.org/u/Trying)
#### Post date: [February 20, 2019, 10:19pm UTC](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691/12 "2019-02-20T22:19:41Z")

</div>

I’m actually surprised all but the OneTouch are so close! Thanks for this!

---

<div class="post-metadata">

### Author: ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)
#### Post date: [February 20, 2019, 10:36pm UTC](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691/13 "2019-02-20T22:36:32Z")

</div>

> [@Aaron](#):
>
> the suspense is killing me…

@Nickyghaleb, when is your blood test result due?

---

<div class="post-metadata">

### Author: ![T1john](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/t1john/32/5546_2.png) [@T1john](https://forum.fudiabetes.org/u/T1john)
#### Post date: [February 21, 2019, 12:09am UTC](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691/14 "2019-02-21T00:09:19Z")

</div>

Second, or third? That responce.

---

<div class="post-metadata">

### Author: ![MM2](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mm2/32/3032_2.png) [@MM2](https://forum.fudiabetes.org/u/MM2)
#### Post date: [February 21, 2019, 2:21am UTC](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691/15 "2019-02-21T02:21:39Z")

</div>

> [@Trying](#):
>
> I’m actually surprised all but the OneTouch are so close! Thanks for this!

[Precise or accurate](https://forum.fudiabetes.org/t/best-bg-meter-precise-or-accurate/371)

---

<div class="post-metadata">

### Author: ![Nickyghaleb](https://avatars.discourse-cdn.com/v4/letter/n/5f8ce5/32.png) [@Nickyghaleb](https://forum.fudiabetes.org/u/Nickyghaleb)
#### Post date: [February 21, 2019, 2:41am UTC](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691/16 "2019-02-21T02:41:19Z")

</div>

So… Of course this is how this would go… but it turns out my endo didn’t “add glucose levels” to the lab request— only a1c. I’ve never had that happen before where I’ve had an a1c done and not received a blood glucose value from the lab draw. So I’ve got nuthin, @Aaron. Sorry. 😕

The only thing I do have is that it turns out that my CVS Health At Home A1c Test Kit was .4 off from my lab a1c. Surprised me… I’ve never had one be more than .1 off before. However, I also usually use the A1c Now brand— so I think I’ll order some more of that brand from now on.

So I guess it’s _mission failed. We’ll get em next time._ 🙄 Sorry again.

---

<div class="post-metadata">

### Author: ![Chris](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/chris/32/382_2.png) [@Chris](https://forum.fudiabetes.org/u/Chris)
#### Post date: [February 21, 2019, 2:51am UTC](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691/17 "2019-02-21T02:51:52Z")

</div>

Man, I was hoping to be right for once. Oh well, will have to wait until next year.

---

<div class="post-metadata">

### Author: ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)
#### Post date: [February 24, 2019, 6:38pm UTC](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691/18 "2019-02-24T18:38:43Z")

</div>

33 posts were split to a new topic: [How often do you see your endocrinologist?](https://forum.fudiabetes.org/t/how-often-do-you-see-your-endocrinologist/6781)

---

<div class="post-metadata">

### Author: ![Eric](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/eric/32/9209_2.png) [@Eric](https://forum.fudiabetes.org/u/Eric)
#### Post date: [February 21, 2019, 4:24am UTC](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691/24 "2019-02-21T04:24:00Z")

</div>

> [@Nickyghaleb](#):
>
> So… Of course this is how this would go… but it turns out my endo didn’t “add glucose levels” to the lab request— only a1c. I’ve never had that happen before where I’ve had an a1c done and not received a blood glucose value from the lab draw. So I’ve got nuthin, @Aaron. Sorry. 😕

You were 68 at the time of the draw. Just in case you were wondering.

---

<div class="post-metadata">

### Author: ![Eric](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/eric/32/9209_2.png) [@Eric](https://forum.fudiabetes.org/u/Eric)
#### Post date: [February 21, 2019, 4:32am UTC](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691/25 "2019-02-21T04:32:21Z")

</div>

> [@T1john](#):
>
> accu-check was way off by over like 40 points.  
> I believe it was whole blood 89  
> Next 86  
> Next link 81  
> Accu-check 140

And the winner of the _Most Inappropriately Named Product of the Year_ award goes to…

[Next page](https://forum.fudiabetes.org/t/onetouch-ultramini-vs-contour-next-one-lab-draw-challenge-2019/6691.md?page=2)