# Loop: Getting Started

**URL:** https://forum.fudiabetes.org/t/loop-getting-started/2216
**Category:** Tipwiki
**Tags:** wiki, diy, closed-loop
**Created:** [September 30, 2017, 7:00am UTC](https://forum.fudiabetes.org/t/loop-getting-started/2216 "2017-09-30T07:00:24Z")
**Posts on this page:** 20
**Page:** 1

<div class="post-metadata">

### Author: ![dm61](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/dm61/32/1602_2.png) [@dm61](https://forum.fudiabetes.org/u/dm61)
#### Post date: [September 30, 2017, 7:00am UTC](https://forum.fudiabetes.org/t/loop-getting-started/2216/1 "2017-09-30T07:00:24Z")

</div>

[Loop](https://loopkit.github.io/loopdocs/) is a do-it-yourself closed-loop insulin delivery system, which consists of 4 components:

1. Open-source [Loop app](https://github.com/LoopKit/Loop) running on an iPhone
2. CGM (Dexcom G6, G5, or G4+share, or Medronic Enlite)
3. A compatible insulin pump, including older [Medtronic pumps](https://loopkit.github.io/loopdocs/setup/requirements/mdt-pump/) and (as of April 22, 2019) [Omnipod](https://loopkit.github.io/loopdocs/setup/requirements/omnipod-pump/) (Eros pods, non-DASH)
4. [RileyLink](https://loopkit.github.io/loopdocs/requirements/rileylink/), an open-source hardware device, which serves as a bridge between iPhone’s Bluetooth Low Energy (BLE) radio and pump’s radio.

![image](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/f/fe2d074aa35f50f4167bde3995c663fcda0248f4.jpg)

Assuming all required components are available, assembling the system is pretty simple: download the Loop source code and install the app on the iPhone using Xcode (Apple developer tool) on a Mac computer. Detailed, easy-to-follow directions are available in the [Loop documentation](https://loopkit.github.io/loopdocs/). No programming or any other special technical skills are required.

Here is a brief [YouTube video introduction to Loop](https://www.youtube.com/watch?v=qw_u1lqboCs), and an [Introduction to DIY Looping with Omnipod](https://beyondtype1.org/diy-looping-omnipod-tips/) by Julia Blanchette and Katie DiSimone at BeyondType1.

**User Experience**

During normal operation, there is no need to touch the pump at all. Everything is done through the Loop app on the iPhone or, optionally, through a companion Apple Watch app. In short, the user manually enters meals, and accepts or overrides bolus recommendations, while corrections are performed automatically by the Loop closed-loop algorithm.

The main Loop screen shows graphs of recent CGM data and predicted BG, active insulin (“insulin on board” or IOB), insulin delivery, and active carbs (“carbs on board” or COB). Tapping on the plate icon in the lower left corner brings up a carb entry menu, which allows the user to enter meal time (which could in the future, or in the past), and food type with expected absorption time. The user should be aware of the fact that “carbs” entered should reflect everything one needs to bolus for (including protein and fat effects). Large mixed carb/protein/fat meals often take multiple hours to digest. After a meal is entered, Loop suggests a bolus, which the user may accept or override at will.

![image](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/4/4197d082d07012034a0cf42654391ed7c98a2456.jpg)

Every 5 minutes, Loop updates its BG prediction, calculates a correction dose, and delivers the correction in the form of a temporary basal rate, with the goal of steering the eventual BG towards the user specified correction range, while avoiding lows.

**Is Loop Effective?**

It is estimated that more than 1,000 people, including adults and kids, have been using Loop for extended periods of time now. Informal feedback and reports have generally been very positive, reflecting improved A1c or other bg control metrics with less effort, improved sleep, and improved quality of life.

An [observational study](https://studies.jaeb.org/ndocs/loop/Public/Participant/StudyRecruitment.aspx) conducted by [Jaeb Center for Health Research](https://www.jaeb.org/) is currently in progress.

**Is Loop Safe?**

Taking insulin in any form is risky, and it is impossible to guarantee that any approach or system is absolutely safe. Loop has been designed and implemented by people who have T1D or who have kids with T1D, and who very well understand the importance of delivering insulin safely. Loop safety features include the following:

- In case of failures, the system defaults back to normal, manual pump operation
- Bolusing requires fingerprint or face recognition identification
- There are limits for maximum allowed bolus, and maximum allowed basal rate
- Loop interface is designed to be transparent to the user, and easy to use
- Open-source code benefits from reviews and feedback by many users

**Is Loop for Me?**

- Loop is an experimental DIY system not approved for therapy. If you are looking for a perfect system backed up by clinical trials, regulatory approvals, and 24/7 technical support, Loop is _not_ for you.
- Currently compatible pumps are older-generation out-of-warranty [Medronic pumps](https://loopkit.github.io/loopdocs/setup/requirements/mdt-pump/), which can be difficult to find. As of April 22, 2019, [Omnipod (Eros, non-DASH) pods](https://loopkit.github.io/loopdocs/setup/requirements/omnipod-pump/) are also supported.
- Beyond basic familiarity with computers, no special skills are required. Nevertheless, people may find DIY concepts and language difficult or intimidating. However, technical support offered by the community is outstanding. As is usually the case, motivation is far more important than any technical skills or aptitude.
- The non-profit [Tidepool](https://tidepool.org/loop/) is working on a version of Loop to be [submitted to FDA](https://tidepool.org/tidepool-delivering-loop/) for approval in future. It is envisioned that the FDA approved Tidepool Loop would be available on the Apple app store and work with in-warranty, compatible insulin pumps. As [reported by diaTribe](https://diatribe.org/tidepool-will-submit-loop-fda-diy-iphone-app-automated-insulin-delivery), Tidepool and Insulet have announced a partnership to support Omnipod DASH in Tidepool Loop.

**Algorithm**

Every 5 minutes, Loop updates its prediction of future BG values over a time period equal to the duration of insulin action. Based on the predicted BG, insulin doses (both boluses and temp basals) are calculated to drive eventual BG into the user specified correction range, while not allowing BG to fall below a user-specified minimum BG level (called Suspend Threshold) at any point in time. BG predictions are based on insulin entries, carb entries and two correction factors (_Glucose Momentum_ and _Retrospective Correction_), which are attempting to mitigate various unmodeled effects.

_Insulin absorption model_

In IOB calculations, Loop algorithms take into account boluses and temp basals, as well as nonlinear absorption curves based on published data. Several choices are available to the user, including legacy curves (based on Walsh’s Pumping Insulin book), as well as scalable exponential curves that are found to better fit actual absorption:

![image](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/c/cbd08e023bf4a63797f639cb97ec20bd22e68927.jpg)

_Dynamic carb absorption model_

The meal entry includes an expected absorption time, which Loop uses only as an initial guess. Based on insulin activity, and observed CGM data, Loop calculates and displays to the user the predicted and the observed carb absorption. Below is an example of a relatively large 90g carbs+protein+fat meal that takes about 5 hours to digest. Based on the observed data, predicted carb absorption is dynamically updated, leading to improved BG predictions and dosing.

![image](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/5/5694dc2ae5971896fd2ba24ad8e01dc9237ab9bc.png)

_Loop playground: contributions to the BG prediction_

All factors contributing to BG predictions are clearly visible to the user on a Loop playground screen. One may turn any of the effects on and off and observe changes in predictions. In addition to the impact of insulin and carbs, the Loop algorithm includes two corrective factors. _Glucose Momentum_ blends in the average slope in BG over past 15 minutes into prediction over next 30 minutes, while _Retrospective Correction_ blends in the observed prediction error over past 30 minutes into the prediction over the next 60 minutes. These corrective features are attempting to mitigate unmodeled effects such as carb entry errors, unmodeled carb absorption, effects of exercise, increased or reduced sensitivity, etc.

![image](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/3/37d3ee23eef99e1f07c2e3b4c42f363bb0f74381.png)

Loop is well integrated with the Apple Health app, including storage and retrieval of BG, carb, and insulin data between Loop and Health. Loop currently relies on the user entered parameters such as default basal rates, carb ratios, and insulin sensitivities. Future versions may include estimation these parameters, automating carb absorption time entries based on meal composition (which could be based on other iOS apps), and other algorithm or usability improvements.

**User blogs, posts and reports**

- [SeeMyCGM](http://seemycgm.com/), blog by Katie DiSimone, T1D parent, and a major contributor in the DIY closed-loop community
- Judy Hoskins’s [Loop story](https://www.healthline.com/diabetesmine/homegrown-diabetes-tool-rileylink)
- [Why We Chose to #Loop #WeAreNotWaiting](http://www.tudiabetes.org/forum/t/why-we-chose-to-loop-wearenotwaiting/63915) post by Lorraine, T1D parent
- [My new diabetes](https://diabetogenic.wordpress.com/2018/01/16/my-new-diabetes/), a blog post by Renza Scibilia
- [Welcome to Looping](https://www.diabetesdaily.com/blog/the-234-diy-artificial-pancreas-welcome-to-looping-549956/) post by Erik Douds, an endurance athlete
- [How I Loop](https://diatribe.org/how-i-loop-two-years-using-iphone-app-automate-my-insulin-delivery): Two Years of Using An iPhone App To Automate My Insulin Delivery, a diaTribe post by Adam Brown.

**Resources**

- [Loop documentation](https://loopkit.github.io/loopdocs/), to get started read [introduction](https://loopkit.github.io/loopdocs/#welcome-to-loop) and [requirements](https://loopkit.github.io/loopdocs/requirements/overview/)
- [Loop Tips](https://kdisimone.github.io/looptips/), a companion to the Loop documentation, provides tips on how Loop is recommending the actions being taken, and how one can improve blood glucose outcomes on Loop.
- [Loop code](https://github.com/LoopKit/Loop): download complete code as a zip file from the Loop github site, and install using Apple [Xcode](https://developer.apple.com/xcode/) app on a Mac following the [instructions](https://loopkit.github.io/loopdocs/build/overview/). You may also take a look at the 16min video [How to build the Loop app](https://www.youtube.com/watch?v=fGvMB98MuX4) by Katie DiSimone.
- Community technical support and forums: [Looped group](https://www.facebook.com/groups/TheLoopedGroup/) on Facebook (need to request to enter), and [Loop Zulip channel](https://loop.zulipchat.com/) (open to view, need to sign up to post)
- [History of Loop](https://medium.com/@loudnate/the-history-of-loop-and-loopkit-59b3caf13805) by the original creator and developer, Nate Racklyeft
- [Building Amazing Diabetes Apps](https://www.youtube.com/watch?v=kxSgi8TojUM) talk by Nate Racklyeft
- Other DIY closed loop systems: [OpenAPS](http://openaps.readthedocs.io/en/latest/index.html), [AndroidAPS](https://github.com/MilosKozak/AndroidAPS/wiki)
- [OpenOmni](http://www.openomni.org/): efforts to enable Omnipod to be used in DYI closed-loop systems are under way. Technical progress can be viewed on the [openomni](https://github.com/openaps/openomni) github site.
- [Nightscout](http://www.nightscout.info/), one of the first #WeAreNotWaiting projects, an open-source system that allows real time access to CGM data via personal website, smartwatch viewers, or apps and widgets available for smartphones. Loop communicates to Nightscout, which integrates all user data: CGM, carbs, Loop actions, etc. allowing real-time web based visualization, sharing and remote monitoring, and a wide range of analysis reports.

* * *

* * *

* * *

_End of wiki ---------- comments start here_

---

<div class="post-metadata">

### Author: ![dm61](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/dm61/32/1602_2.png) [@dm61](https://forum.fudiabetes.org/u/dm61)
#### Post date: [September 30, 2017, 7:02am UTC](https://forum.fudiabetes.org/t/loop-getting-started/2216/2 "2017-09-30T07:02:01Z")

</div>

Here is my attempt at a Loop wiki. I know it’s too long, sorry I am not good at this 🙂

---

<div class="post-metadata">

### Author: ![Thomas](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/thomas/32/1719_2.png) [@Thomas](https://forum.fudiabetes.org/u/Thomas)
#### Post date: [September 30, 2017, 10:33am UTC](https://forum.fudiabetes.org/t/loop-getting-started/2216/3 "2017-09-30T10:33:31Z")

</div>

It looks like a good summary. I would not say it is too long. It was probably quite difficult trying to decide how to boil it down as I am sure you could have easily written pages more.

---

<div class="post-metadata">

### Author: ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)
#### Post date: [September 30, 2017, 1:50pm UTC](https://forum.fudiabetes.org/t/loop-getting-started/2216/4 "2017-09-30T13:50:33Z")

</div>

> [@dm61](#):
>
> Here is my attempt at a Loop wiki.

This is OUTSTANDING!

Would an existing RileyLink be the appropriate hardware to work with an Omnipod when the pod comm is fully cracked, or will the Pod’s RF require different hardware?

Would it be possible to have a section that discusses how to start looping? Or possibly just including the build ([Mac-Xcode Overview - LoopDocs](https://loopkit.github.io/loopdocs/build/overview/)) and use ([https://loopkit.github.io/loopdocs/use/close-loop/](https://loopkit.github.io/loopdocs/use/close-loop/))? I have read the two Build and Use tabs several times, but I still don’t get the feeling that I am quite sure how to attack it ☹ It may well be me though!

A couple of links I have kept track of that you may, or may not, decide to include:

- list of different system options: [Choices for DIY Loops | Diabettech - Diabetes and Technology](http://www.diabettech.com/looping-a-guide/)
- a year with Loop: [A year of closed loop | Diabettech - Diabetes and Technology](http://www.diabettech.com/openaps/a-year-of-looping-how-closed-loop-systems-changed-my-life-with-t1d)
- history of Loop: [The History of Loop and LoopKit. Reflecting on the past in celebration… | by Nate Racklyeft | Medium](https://medium.com/@loudnate/the-history-of-loop-and-loopkit-59b3caf13805)
- DIYPS…org (not the same I know): [https://diyps.org/](https://diyps.org/)
- Judy Hoskins’s Loop story: ["Artificial Pancreas" aka Automated Insulin Delivery: What You Should](https://www.healthline.com/diabetesmine/homegrown-diabetes-tool-rileylink)

I am so pleased you wrote this! This is really WONDERFUL!

---

<div class="post-metadata">

### Author: ![Thomas](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/thomas/32/1719_2.png) [@Thomas](https://forum.fudiabetes.org/u/Thomas)
#### Post date: [September 30, 2017, 2:09pm UTC](https://forum.fudiabetes.org/t/loop-getting-started/2216/5 "2017-09-30T14:09:06Z")

</div>

> [@Michel](#):
>
> Would an existing RileyLink be the appropriate hardware to work with an Omnipod when the pod comm is fully cracked, or will the Pod’s RF require different hardware?

916 MHz (Medtronic) vs 433 MHz (Omnipod)  
(I think?)

---

<div class="post-metadata">

### Author: ![ClaudnDaye](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/claudndaye/32/5_2.png) [@ClaudnDaye](https://forum.fudiabetes.org/u/ClaudnDaye)
#### Post date: [September 30, 2017, 2:54pm UTC](https://forum.fudiabetes.org/t/loop-getting-started/2216/6 "2017-09-30T14:54:56Z")

</div>

Awesome write-up! Thanks for taking the time to put this together!

---

<div class="post-metadata">

### Author: ![CatLady](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/catlady/32/36_2.png) [@CatLady](https://forum.fudiabetes.org/u/CatLady)
#### Post date: [September 30, 2017, 3:28pm UTC](https://forum.fudiabetes.org/t/loop-getting-started/2216/7 "2017-09-30T15:28:02Z")

</div>

Awesome!! 👏🏻

---

<div class="post-metadata">

### Author: ![dm61](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/dm61/32/1602_2.png) [@dm61](https://forum.fudiabetes.org/u/dm61)
#### Post date: [September 30, 2017, 5:08pm UTC](https://forum.fudiabetes.org/t/loop-getting-started/2216/8 "2017-09-30T17:08:43Z")

</div>

> [@Michel](#):
>
> Would an existing RileyLink be the appropriate hardware to work with an Omnipod when the pod comm is fully cracked, or will the Pod’s RF require different hardware?

Pete Schwamb has successfully sent commands to a Pod using the existing RileyLink hardware. The sub-GHz chip on RileyLink is capable of operating in both 433 MHz and 916 MHz bands. For improved reception, it might be good to modify the RF matching network some, but it looks like the same hardware could conveniently work for all pumps.

---

<div class="post-metadata">

### Author: ![dm61](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/dm61/32/1602_2.png) [@dm61](https://forum.fudiabetes.org/u/dm61)
#### Post date: [September 30, 2017, 5:11pm UTC](https://forum.fudiabetes.org/t/loop-getting-started/2216/9 "2017-09-30T17:11:11Z")

</div>

> [@Michel](#):
>
> Would it be possible to have a section that discusses how to start looping? Or possibly just including the build ([Mac-Xcode Overview - LoopDocs](https://loopkit.github.io/loopdocs/build/overview/)) and use ([https://loopkit.github.io/loopdocs/use/close-loop/](https://loopkit.github.io/loopdocs/use/close-loop/))? I have read the two Build and Use tabs several times, but I still don’t get the feeling that I am quite sure how to attack it ☹ It may well be me though!

I have a short statement in the top part on how to get going, and I’ve also included a bit more in the resources. It really is not difficult at all. It’s like trying to describe how to learn to drive a car - no amount of words or books can help - you just need to sit down and give it a try.

---

<div class="post-metadata">

### Author: ![dm61](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/dm61/32/1602_2.png) [@dm61](https://forum.fudiabetes.org/u/dm61)
#### Post date: [October 4, 2017, 11:18pm UTC](https://forum.fudiabetes.org/t/loop-getting-started/2216/10 "2017-10-04T23:18:30Z")

</div>

Any further comments, questions, suggestions? If not, @Michel, feel free to convert it into a published wiki post if you wish, or let me know how to do that.

---

<div class="post-metadata">

### Author: ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)
#### Post date: [October 4, 2017, 11:21pm UTC](https://forum.fudiabetes.org/t/loop-getting-started/2216/11 "2017-10-04T23:21:52Z")

</div>

Dragan, thank you so much!

I’ll give it a day or two for comments and convert it. Give me the green light whenever you feel it is good to go! I will also write down what I do for that for documentation.

---

<div class="post-metadata">

### Author: ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)
#### Post date: [October 10, 2017, 7:03am UTC](https://forum.fudiabetes.org/t/loop-getting-started/2216/12 "2017-10-10T07:03:24Z")

</div>

@dm61, is this lovely post ready to convert?

---

<div class="post-metadata">

### Author: ![dm61](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/dm61/32/1602_2.png) [@dm61](https://forum.fudiabetes.org/u/dm61)
#### Post date: [October 10, 2017, 2:23pm UTC](https://forum.fudiabetes.org/t/loop-getting-started/2216/13 "2017-10-10T14:23:30Z")

</div>

Sure. Any further edits or corrections could be made live.

---

<div class="post-metadata">

### Author: ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)
#### Post date: [October 10, 2017, 4:00pm UTC](https://forum.fudiabetes.org/t/loop-getting-started/2216/14 "2017-10-10T16:00:23Z")

</div>

Done! Thanks so much, this is a GREAT wiki!

---

<div class="post-metadata">

### Author: ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)
#### Post date: [October 13, 2017, 3:00am UTC](https://forum.fudiabetes.org/t/loop-getting-started/2216/15 "2017-10-13T03:00:03Z")

</div>

@dm61, I have been scouring the Loop wiki. What I understand from it and its links (I am sure all obvious to you!):

- we need to run Loop on an iPhone
- to integrate all the data needed by Loop, we need to run Nightscout in the background, in the cloud.
- we need to use the Dexcom apps to interface to the G4/G5 CGMs
- we need an Apple computer to compile the app
- and, naturally, we do not have the option to run xDrip/xDrip+ within Loop.

---

<div class="post-metadata">

### Author: ![dm61](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/dm61/32/1602_2.png) [@dm61](https://forum.fudiabetes.org/u/dm61)
#### Post date: [October 13, 2017, 4:06am UTC](https://forum.fudiabetes.org/t/loop-getting-started/2216/16 "2017-10-13T04:06:38Z")

</div>

> [@Michel](#):
>
> we need to run Loop on an iPhone

Correct. There is no Android version of Loop.

> [@Michel](#):
>
> to integrate all the data needed by Loop, we need to run Nightscout in the background, in the cloud.

No, Loop does not require Nightscout at all, and it does not pull any data from Nightscout. In fact, Loop operates just fine offline - does not require internet access for normal operation. Loop is capable of sending its data to Nightscout, which is very convenient for parents who follow kids using Nightscout. But, Nightscout is optional for Loop, not required.

> [@Michel](#):
>
> we need to use the Dexcom apps to interface to the G4/G5 CGMs

Correct (but see more detailed answer below to your last question)

> [@Michel](#):
>
> we need an Apple computer to compile the app

Correct.

> [@Michel](#):
>
> and, naturally, we do not have the option to run xDrip/xDrip+ within Loop.

Not entirely correct. It is possible to use [iosxdripreader](https://github.com/JohanDegraeve/iosxdripreader) with Loop, but that option requires internet access (there is some talk about potentially integrating iosxdripreader features within Loop, but I’d not wait for that). Also, iosxdripreader is in earlier work-in-progress stages compared to Xdrip+ for Android. People who are using iosxdripreader with Loop are those who are trying to extend G5 transmitter life beyond Dex imposed exp date, or those who are using G5 transmitters with DIY-replaced batteries. In such situations, some people are even using two phones, Android with Xdrip+, and iPhone with Loop.

---

<div class="post-metadata">

### Author: ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)
#### Post date: [October 13, 2017, 9:44am UTC](https://forum.fudiabetes.org/t/loop-getting-started/2216/17 "2017-10-13T09:44:46Z")

</div>

That is great info, clarifies a lot.

---

<div class="post-metadata">

### Author: ![Thomas](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/thomas/32/1719_2.png) [@Thomas](https://forum.fudiabetes.org/u/Thomas)
#### Post date: [October 13, 2017, 11:42am UTC](https://forum.fudiabetes.org/t/loop-getting-started/2216/18 "2017-10-13T11:42:06Z")

</div>

Would it make sense to have a paragraph on Loop vs xdrip? Very high level. One paragraph only. Include this in both the loop and the xdrip wiki. The paragraph in the loop wiki has a pointer to the xdrip wiki and vice versa.

??

---

<div class="post-metadata">

### Author: ![Mariana](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mariana/32/4679_2.png) [@Mariana](https://forum.fudiabetes.org/u/Mariana)
#### Post date: [November 7, 2018, 7:32pm UTC](https://forum.fudiabetes.org/t/loop-getting-started/2216/19 "2018-11-07T19:32:00Z")

</div>

this is just amazing! I just got my Riley Link yesterday and am buuilding as I read, thank you for making it easier for all of us 😃

---

<div class="post-metadata">

### Author: ![christopherasnider](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/christopherasnider/32/5330_2.png) [@christopherasnider](https://forum.fudiabetes.org/u/christopherasnider)
#### Post date: [December 5, 2018, 5:00pm UTC](https://forum.fudiabetes.org/t/loop-getting-started/2216/20 "2018-12-05T17:00:07Z")

</div>

> [@dm61](#):
>
> No formal studies have been done to evaluate performance of Loop or any other DIY closed-loop system.

The Jaeb Center for Health Research will be conducting an observational study of current Loopers next year. The Looped Facebook group will be the primary place for updates on recruitment.

[Next page](https://forum.fudiabetes.org/t/loop-getting-started/2216.md?page=2)