# How Long Does a G5 Transmitter REALLY Last? \*\*UPDATE-Transmitter Died\*\* UPDATE-New Transmitter \*\*\*UPDATE-2nd Transmitter 😭

**URL:** <https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307>\
**Category:** C/GM\
**Tags:** dexcom\
**Created:** [June 22, 2017, 4:05pm UTC](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307 "2017-06-22T16:05:04Z")\
**Posts on this page:** 20\
**Page:** 4

<div class="post-metadata">

**Author:** ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)\
**Post date:** [August 4, 2017, 2:16pm UTC](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307/61 "2017-08-04T14:16:27Z")

</div>

@Thomas That is a piece of information I neglected to document. Boxes are long gone. I’ll be sure to note the next ones.

---

<div class="post-metadata">

**Author:** ![daisymae](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/daisymae/32/1199_2.png) [@daisymae](https://forum.fudiabetes.org/u/daisymae)\
**Post date:** [August 4, 2017, 9:47pm UTC](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307/62 "2017-08-04T21:47:59Z")

</div>

if i have a G5 transmitter from 2 years ago that i have never used, is it still functional or does it expire despite the lack of its use?

---

<div class="post-metadata">

**Author:** ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)\
**Post date:** [August 4, 2017, 11:14pm UTC](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307/63 "2017-08-04T23:14:39Z")

</div>

@daisymae I think the batteries do not start discharging until the transmitter has been connected to a sensor. So in theory it should still work.

---

<div class="post-metadata">

**Author:** ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)\
**Post date:** [August 10, 2017, 1:35pm UTC](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307/65 "2017-08-10T13:35:43Z")

</div>

Weekly G5 Transmitter Update: **172 days!**

Well, it appears as if the batteries are getting stronger!

Resistance has **decreased** to 592, and I really have not seen it over 600 this past week. What seems even more amazing is that voltage A and B are slightly increased as well.

It’s looking like all previous guesses on this transmitter’s longevity will be out the window.

Dexcom: Takes a licking and keeps on ticking 😁

 ![08](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/7/7705d647f2c3b858c5a0ced0a1470bbfe2525217.jpg)

---

<div class="post-metadata">

**Author:** ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)\
**Post date:** [August 10, 2017, 1:41pm UTC](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307/66 "2017-08-10T13:41:01Z")

</div>

Almost forgot the now obligatory puppy picture.

From her puppy training graduation about six weeks ago. Given her current state of behavior I think I’m going to request a refund!

![1109](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/7/798bc6ac2ff23f9ebf5dec9e516a242a29f8aab0.jpg)

---

<div class="post-metadata">

**Author:** ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)\
**Post date:** [August 10, 2017, 5:14pm UTC](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307/67 "2017-08-10T17:14:50Z")

</div>

> [@docslotnick](#):
>
> Weekly G5 Transmitter Update: 172 days!
> 
> Well, it appears as if the batteries are getting stronger!

That is absolutely amazing. This is radically changing some of the understanding I had of the G4 and G5.

---

<div class="post-metadata">

**Author:** ![TiaG](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/tiag/32/594_2.png) [@TiaG](https://forum.fudiabetes.org/u/TiaG)\
**Post date:** [August 10, 2017, 8:39pm UTC](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307/68 "2017-08-10T20:39:56Z")

</div>

So do I get the Price is Right dead transmitter then? i think that was my guess…

---

<div class="post-metadata">

**Author:** ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)\
**Post date:** [August 17, 2017, 2:54pm UTC](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307/69 "2017-08-17T14:54:40Z")

</div>

@TiaG Can’t have my dead transmitter yet, it’s still going strong!

**G5 Transmitter Weekly Update**

Still going strong at 179 days! Resistance holding steady at 587, not even close yet to the dreaded 1000 mark.

The only puzzling glitch is that Voltage B has dropped below 300 to 299. Looks like that battery will be the Achilles heel. The question still is when it will fail.

But if it fails this week, well six months isn’t too bad.

 ![37](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/c/ce7e073efd5a8a88666afa65d56c2a6680c0782f.jpg)

---

<div class="post-metadata">

**Author:** ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)\
**Post date:** [August 17, 2017, 6:22pm UTC](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307/70 "2017-08-17T18:22:55Z")

</div>

> [@docslotnick](#):
>
> Still going strong at 179 days!

I am thinking there must be a hidden movement-driven battery charger somewhere in your G5… You have a custom model!

---

<div class="post-metadata">

**Author:** ![Chris](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/chris/32/382_2.png) [@Chris](https://forum.fudiabetes.org/u/Chris)\
**Post date:** [August 17, 2017, 10:35pm UTC](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307/71 "2017-08-17T22:35:33Z")

</div>

Nah, I think that cell B will fall off a cliff at some point, and the then it will be done in quickly.

---

<div class="post-metadata">

**Author:** ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)\
**Post date:** [August 21, 2017, 6:45pm UTC](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307/73 "2017-08-21T18:45:38Z")

</div>

Looks like the end is near. Voltages are low and looks like they’re ready to plummet, and the transmitter to burn and die.

Note how xDrip+ now highlights these values in red. It’s already given me two notifications that the transmitter is almost dead.

 ![Screenshot_2017-08-21-13-35-02](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/6/6356b79fcae1e5b5c4a314da4e7d63bed60d90d5.png)

---

<div class="post-metadata">

**Author:** ![Chris](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/chris/32/382_2.png) [@Chris](https://forum.fudiabetes.org/u/Chris)\
**Post date:** [August 21, 2017, 7:19pm UTC](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307/74 "2017-08-21T19:19:20Z")

</div>

Still pretty amazing that you have used it for almost a half a year!

---

<div class="post-metadata">

**Author:** ![Millz](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/millz/32/12204_2.png) [@Millz](https://forum.fudiabetes.org/u/Millz)\
**Post date:** [August 21, 2017, 7:23pm UTC](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307/75 "2017-08-21T19:23:34Z")

</div>

Awe 🙃

---

<div class="post-metadata">

**Author:** ![Osoblu](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/osoblu/32/2258_2.png) [@Osoblu](https://forum.fudiabetes.org/u/Osoblu)\
**Post date:** [August 21, 2017, 9:50pm UTC](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307/76 "2017-08-21T21:50:22Z")

</div>

Hi. I just joined this incredible group. I started my dexcom 5 on tax day, have been using xdrip+ and have already extended the battery life by 1 month. Could you tell me how you get that status screen? Mine is much less informative.  
I’m not sure if this screen shot uploaded correctly.

 ![Screenshot_20170821-173454](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/1/174e3bbaf2bc39c47226d860b17c7e9be4d122b8.png)

---

<div class="post-metadata">

**Author:** ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)\
**Post date:** [August 21, 2017, 9:56pm UTC](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307/77 "2017-08-21T21:56:20Z")

</div>

Hi @Osoblu Welcome to the best DOC on the interwebs. Glad you joined us!

You can get other status pages simply by swiping left. The next one, the G5 status page, is the one with all of the juicy information.

---

<div class="post-metadata">

**Author:** ![ClaudnDaye](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/claudndaye/32/5_2.png) [@ClaudnDaye](https://forum.fudiabetes.org/u/ClaudnDaye)\
**Post date:** [August 21, 2017, 10:00pm UTC](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307/78 "2017-08-21T22:00:32Z")

</div>

I don’t remember, can you view the data on-line via Clarity with xDrip+ also? How do you see the trending data with xDrip? (past 7, 14, 30, 90 days data as you see online with Clarity/glooka)

---

<div class="post-metadata">

**Author:** ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)\
**Post date:** [August 21, 2017, 10:13pm UTC](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307/79 "2017-08-21T22:13:43Z")

</div>

@ClaudnDaye Harold, it’s all there. Just but the “hamburger” menu in the upper right, go to statistics, and you get all kinds of info for today, yesterday, 7 days, 30 days, 90 days. Here’s a sample.

Also, you can get a daily glucose track with daily stats by going to history.

 ![Screenshot_2017-08-21-17-07-48](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/8/879707aaf1d870da36257869ae3d5ec9f983e089.png)  
 ![Screenshot_2017-08-21-17-08-11](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/2/2b96945f87354c5e2e8779fcd3e000f6734ae505.png)

---

<div class="post-metadata">

**Author:** ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)\
**Post date:** [August 22, 2017, 3:27pm UTC](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307/80 "2017-08-22T15:27:21Z")

</div>

Split this topic on August 22, 2017, 3:27pm UTC

3 posts were split to a new topic: [xDrip+ on iPhones?](https://forum.fudiabetes.org/t/xdrip-on-iphones/1673)

---

<div class="post-metadata">

**Author:** ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)\
**Post date:** [August 24, 2017, 3:04pm UTC](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307/81 "2017-08-24T15:04:16Z")

</div>

Well, almost dead. It’s been 186 days.

Been getting notifications from xDrip+ that the transmitter is almost dead for about a week now. Resistance is still pretty good (under 700), but the batterys’ voltages are just about ready to fall off of the cliff.  
(Note that they are highlighted in red by the app)

 ![36](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/1/1a2cd378d2c17446879499f09f3951f78c5219f6.jpg)

These are the notifications I’ve been getting

 ![46](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/c/cf01c7b0a15a56974e85426df0a2c1eaf5d2f428.jpg)

On my last transmitter I only got the notification twice before the transmitter died. So this one doesn’t seem like it’s going to go quietly without a fight.

---

<div class="post-metadata">

**Author:** ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)\
**Post date:** [August 29, 2017, 4:16pm UTC](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307/82 "2017-08-29T16:16:57Z")

</div>

Well, the transmitter that wouldn’t die, died last night. Because I use the watch to connect to the transmitter the readings just stopped without any notification on the phone.

I woke up because my Bg was low at about 6am and noticed I had not gotten a reading for about five hours.

Unfortunately, I botched the screenshot (Bg was 45 so I’m not surprised), but Voltage A was 256 and Voltage B was about 230. Resistance was, remarkably only 557.

I think it lasted for a week and a half after the voltage dropped below 300 because the resistance stayed fairly low. @Chris, is this a valid assumption?

But the transmitter lasted for **191** days!  
That is nearly twice as long as the Dexcom controlled 105 day limit. We’ll see how long the new one that I started this morning will last.

[Previous page](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307.md?page=3)

[Next page](https://forum.fudiabetes.org/t/how-long-does-a-g5-transmitter-really-last-update-transmitter-died-update-new-transmitter-update-2nd-transmitter/1307.md?page=5)