# How good is your BG meter?

**URL:** <https://forum.fudiabetes.org/t/how-good-is-your-bg-meter/1716>\
**Category:** C/GM\
**Created:** [August 25, 2017, 2:56pm UTC](https://forum.fudiabetes.org/t/how-good-is-your-bg-meter/1716 "2017-08-25T14:56:33Z")\
**Posts on this page:** 8\
**Page:** 1

<div class="post-metadata">

**Author:** ![Eric](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/eric/32/9209_2.png) [@Eric](https://forum.fudiabetes.org/u/Eric)\
**Post date:** [August 25, 2017, 2:56pm UTC](https://forum.fudiabetes.org/t/how-good-is-your-bg-meter/1716/1 "2017-08-25T14:56:33Z")

</div>

A little information on checking how good your BG meter is in terms of accuracy and precision.

First, here is a quick explanation of the difference between “precise” and “accurate”:  
[precise-or-accurate](https://forum.fudiabetes.org/t/best-bg-meter-precise-or-accurate/371)

* * *

If you are wondering how good your meter is, there are a number of sites that give ratings. So you can start with checking your meter against these sites:

[diatribe - blood-glucose–18-meters](https://diatribe.org/are-blood-glucose-meters-accurate-new-data-18-meters)  
[Diabetes Technology Society Blood Glucose Monitor System](https://www.diabetestechnology.org/surveillance.shtml)  
[integrateddiabetes-blood-glucose-meter-comparisons](http://integrateddiabetes.com/2016-blood-glucose-meter-comparisons/)

* * *

1. Most important (that’s why it’s listed as #1 !), whenever you are checking your blood sugar, and whenever you want to validate how good your meter is, use proper technique - clean and dry fingers! If your fingers have food on them, particularly sweet foods, you can’t blame the meter…

2. Try to check your meter when your blood sugar is not rapidly going up or down, and is in a somewhat normal range, like 80-120 mg/dL (4.4-6.7 mmol/L). Meters are not as accurate for blood sugar that is extremely high or low. (See Table 2: [http://sites.bu.edu/bionicpancreas/files/2016/11/Comparative-Accuracy-Meters-Ekhlaspour-J-Diabetes-Sci-Technol-2016.pdf](http://sites.bu.edu/bionicpancreas/files/2016/11/Comparative-Accuracy-Meters-Ekhlaspour-J-Diabetes-Sci-Technol-2016.pdf) )

3. When you are checking your own meter, a very easy and useful thing to do is do several tests at the same time, even using the same drop of blood if possible. When doing two tests back-to-back, if you see close numbers like 94 and 96 mg/dL, that’s not bad. If you see numbers a bit further apart like 94 and 117 mg/dL, that’s worse. The same drop of blood should always give you close results. If you are not seeing close results, it would be a good idea to start looking for a new meter.

4. Keep in mind, meters are generally within a few percentage points of your BG value. So the higher your BG is, the bigger a difference that percentage will make. As an example, imagine your meter is always with 5% of the correct value:

- If your BG is 80 mg/dL, 5% of 80 is 4 points, so your meter would be between 76 and 84.

- If your BG is 300 mg/dL, 5% of 300 is 15 points, so your meter would be between 285 and 315.

- In the first case (when your BG is 80 mg/dL) you have a 8 point swing, from 76 to 84. But in the second case (when your BG is 300 mg/dL) you have a 30 point swing, from 285 and 315. In this example, that same 5% difference went from 8 points to 30 points.

1. Occasionally you may see a crazy number. This can happen with any meter. If your meter throws out a number that seems wrong, wash and dry and re-check. The key word is “occasionally”. If it happens a lot, you should consider looking into getting a new meter.

2. Using your CGM numbers to validate your meter won’t be useful for determining if your meter is accurate. It’s fun to see that your CGM matches your meter reading. You can be happy and excited about it, but it really isn’t a good test for meter accuracy.

3. Using your A1C result to validate your meter is not extremely useful. The average BG to A1C scale is not exactly the same for everyone. Average BG and A1C do not line-up perfectly. So you can’t look at your A1C and BG average and say your meter is good or bad.

4. Additionally, your meter average is **not** the same as your average BG. Your meter average is only the same as your average BG **for the times you have tested**. Generally more tests are done before and after meals, and before and after going to sleep. Also, people generally will be more likely to do _repeated_ tests when they are high or low, and less likely to repeat a test when their BG is good. So average BG is not useful for validating meter accuracy.

5. It is helpful validate your meter with a lab blood test. This is easy to do. When you are getting your lab tests done during your endocrinologist visit, right before they draw the blood you should check your BG with your meter twice. Write the numbers down! When you get your endocrinologist’s lab blood test results back, they will usually have the blood glucose value as part of the blood test panel they performed. Compare the lab BG value with the results you got from your meter right at the same time as the blood test (which is why you wrote those numbers down!). Those numbers should be close, particularly if you were in normal range.

6. Even if you use your CGM much more than your BG meter, and think the CGM is more useful for you, keep in mind that you are using your BG meter to calibrate your CGM. So it is important to have a meter you can trust with good values.

Just a few points on checking how good your meter is. Please feel free to ask questions.

---

<div class="post-metadata">

**Author:** ![Thomas](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/thomas/32/1719_2.png) [@Thomas](https://forum.fudiabetes.org/u/Thomas)\
**Post date:** [August 25, 2017, 3:43pm UTC](https://forum.fudiabetes.org/t/how-good-is-your-bg-meter/1716/2 "2017-08-25T15:43:07Z")

</div>

@Eric - Awesome post.  
🙂

I am not sure if you have a queue of testing lined up. But it would be interesting to know how the first drop of blood (from a good and proper fingerstick) compares to the second drop of blood.

Eh. I guess this is already done and not so interesting. lol

[http://www.mayoclinic.org/diseases-conditions/diabetes/expert-blog/testing-blood-sugar/bgp-20056579](http://www.mayoclinic.org/diseases-conditions/diabetes/expert-blog/testing-blood-sugar/bgp-20056579)

The conclusion of this study, which included 123 participants, is: “The first drop of blood can be used for self-monitored glucose testing, but only after washing hands. If washing hands is not possible and they are not visibly soiled or exposed to a sugar-containing product, it is acceptable to use the second drop of blood after wiping away the first drop.” The study also found that external pressure (or squeezing) of the finger can lead to unreliable readings, as well.

You can find the article at: [http://care.diabetesjournals.org/content/early/2011/02/01/dc10-1694](http://care.diabetesjournals.org/content/early/2011/02/01/dc10-1694)

---

<div class="post-metadata">

**Author:** ![Eric](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/eric/32/9209_2.png) [@Eric](https://forum.fudiabetes.org/u/Eric)\
**Post date:** [August 25, 2017, 3:49pm UTC](https://forum.fudiabetes.org/t/how-good-is-your-bg-meter/1716/3 "2017-08-25T15:49:47Z")

</div>

Thanks for the additional info!

> [@Thomas](#):
>
> The study also found that external pressure (or squeezing) of the finger can lead to unreliable readings, as well.

I have heard that also. Squeezing puts more of the non-blood (interstitial) fluids on the strip. The problem is, at some point it becomes difficult to get a drop of blood out without squeezing!

---

<div class="post-metadata">

**Author:** ![ClaudnDaye](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/claudndaye/32/5_2.png) [@ClaudnDaye](https://forum.fudiabetes.org/u/ClaudnDaye)\
**Post date:** [August 25, 2017, 3:51pm UTC](https://forum.fudiabetes.org/t/how-good-is-your-bg-meter/1716/4 "2017-08-25T15:51:35Z")

</div>

We nearly always have to squeeze (not hard, but enough) because the blood doesn’t just pop out like I wish it would every time.

---

<div class="post-metadata">

**Author:** ![Thomas](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/thomas/32/1719_2.png) [@Thomas](https://forum.fudiabetes.org/u/Thomas)\
**Post date:** [August 25, 2017, 3:59pm UTC](https://forum.fudiabetes.org/t/how-good-is-your-bg-meter/1716/5 "2017-08-25T15:59:39Z")

</div>

> [@Eric](#):
>
> at some point it becomes difficult to get a drop of blood out without squeezing!

I have a cat that is very quickly able to generate blood from a finger with no squeezing.

---

<div class="post-metadata">

**Author:** ![Eric](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/eric/32/9209_2.png) [@Eric](https://forum.fudiabetes.org/u/Eric)\
**Post date:** [August 25, 2017, 4:18pm UTC](https://forum.fudiabetes.org/t/how-good-is-your-bg-meter/1716/6 "2017-08-25T16:18:27Z")

</div>

> [@Thomas](#):
>
> I have a cat that is very quickly able to generate blood from a finger with no squeezing.

Oooh! Market that idea!

 ![image](https://canada1.discourse-cdn.com/flex030/uploads/fudiabetes/original/2X/6/6d6185ac44a7f5d7837fc5a54d2999721b9108ff.jpg)

---

<div class="post-metadata">

**Author:** ![Chris](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/chris/32/382_2.png) [@Chris](https://forum.fudiabetes.org/u/Chris)\
**Post date:** [August 25, 2017, 4:43pm UTC](https://forum.fudiabetes.org/t/how-good-is-your-bg-meter/1716/7 "2017-08-25T16:43:10Z")

</div>

The best part of the cat, is you can lance more than one finger at a time!

---

<div class="post-metadata">

**Author:** ![Beacher](https://avatars.discourse-cdn.com/v4/letter/b/a6a055/32.png) [@Beacher](https://forum.fudiabetes.org/u/Beacher)\
**Post date:** [August 25, 2017, 4:44pm UTC](https://forum.fudiabetes.org/t/how-good-is-your-bg-meter/1716/8 "2017-08-25T16:44:58Z")

</div>

> [@Chris](#):
>
> The best part of the cat, is you can lance more than one finger at a time!

Plus experiment with many, many new sites!