# How does a Dexcom Continuous Glucose Monitor work?

**URL:** <https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292>\
**Category:** Tipwiki\
**Tags:** wiki, cgm, dexcom\
**Created:** [October 8, 2017, 2:46pm UTC](https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292 "2017-10-08T14:46:38Z")\
**Posts on this page:** 20\
**Page:** 3

<div class="post-metadata">

**Author:** ![Thomas](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/thomas/32/1719_2.png) [@Thomas](https://forum.fudiabetes.org/u/Thomas)\
**Post date:** [October 10, 2017, 3:51pm UTC](https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292/44 "2017-10-10T15:51:09Z")

</div>

I don’t even know how the calibrate process works. @dm61 do you have an understanding of how the xdrip+ calibration works?

One possibility as I see it (thinking on the G5) is that the calibration data points are stored within the Transmitter memory. (We already know it has a decent sized memory to be able to hold 3 hr of backfill so likely it can hold a number of other data elements.) The transmitter perhaps holds the six most recent calibration data points (BG and time) as entered. The transmitter might send an uncalibrated (raw?) data signal every 5 minutes to the Receiving device. As well, queuing/messaging technology could be used to send the calibration data points and 3 hr of backfill only if needed. The 505 algorithm running on either the Dexcom Receiver or the Dexcom G5 Mobile App or the Tandem t:slim X2 might then use a linear least squares regression method to “on the fly” apply the calibration data points to the raw data so as to produce a final BG reading.

Or it could be doing something entirely different.

---

<div class="post-metadata">

**Author:** ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)\
**Post date:** [October 10, 2017, 4:13pm UTC](https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292/45 "2017-10-10T16:13:17Z")

</div>

> [@Thomas](#):
>
> So, Dexcom lists the original G4 with a height of 13mm which would be 8mm for the transmitter and 5mm additional with the sensor housing.

@Thomas, I added your pic to the transmitter section.

---

<div class="post-metadata">

**Author:** ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)\
**Post date:** [October 10, 2017, 4:14pm UTC](https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292/46 "2017-10-10T16:14:16Z")

</div>

OK, I scrubbed the content, added a bit more about battery replacement, and added to the resources. I am going to wikify it 🙂

---

<div class="post-metadata">

**Author:** ![ClaudnDaye](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/claudndaye/32/5_2.png) [@ClaudnDaye](https://forum.fudiabetes.org/u/ClaudnDaye)\
**Post date:** [October 10, 2017, 4:38pm UTC](https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292/47 "2017-10-10T16:38:15Z")

</div>

There is a guy who did this a lot with G4’s and made videos about them on Youtube. I asked him if he’d be posting a G5 video and he indicated this will be coming in the very near future so hopefully something will be published by him soon - if it is, I’ll link it.

---

<div class="post-metadata">

**Author:** ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)\
**Post date:** [October 10, 2017, 4:42pm UTC](https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292/48 "2017-10-10T16:42:38Z")

</div>

> [@ClaudnDaye](#):
>
> I asked him if he’d be posting a G5 video and he indicated this will be coming in the very near future

That is awesome!

---

<div class="post-metadata">

**Author:** ![ClaudnDaye](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/claudndaye/32/5_2.png) [@ClaudnDaye](https://forum.fudiabetes.org/u/ClaudnDaye)\
**Post date:** [October 10, 2017, 4:43pm UTC](https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292/49 "2017-10-10T16:43:40Z")

</div>

His videos are very thorough and awesome videography also, so I really hope he posts them - he doesn’t leave out a single step and each one is fully explained.

---

<div class="post-metadata">

**Author:** ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)\
**Post date:** [October 10, 2017, 4:44pm UTC](https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292/50 "2017-10-10T16:44:24Z")

</div>

> [@ClaudnDaye](#):
>
> he doesn’t leave out a single step and each one is fully explained.

Should we add a link to a G4 video of his?

---

<div class="post-metadata">

**Author:** ![ClaudnDaye](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/claudndaye/32/5_2.png) [@ClaudnDaye](https://forum.fudiabetes.org/u/ClaudnDaye)\
**Post date:** [October 10, 2017, 4:45pm UTC](https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292/51 "2017-10-10T16:45:21Z")

</div>

[Dexcom G4 Battery Change-out (by Vince P)](https://www.youtube.com/watch?v=TLkyev3lWWw&lc=z22duxxz2kaqzfznfacdp434loikkrsxhbft5kgsahdw03c010c.1507423982710055)

---

<div class="post-metadata">

**Author:** ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)\
**Post date:** [October 10, 2017, 4:46pm UTC](https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292/52 "2017-10-10T16:46:09Z")

</div>

Funny! We already have his link 🙂 Great news, can’t wait to see his G5 video.

---

<div class="post-metadata">

**Author:** ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)\
**Post date:** [October 10, 2017, 5:53pm UTC](https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292/53 "2017-10-10T17:53:47Z")

</div>

> [@dm61](#):
>
> that’s why I thought the calibration algorithm was on the receiver. At least, I’m pretty sure it’s on the receiver for the G4 system.

I am still puzzled by the calibration questions 🙂 It seems to me it is very likely to be on the G4 receiver — although the G4 could pass the updated code along to the transmitter in an update process.

> [@dm61](#):
>
> The fact that you can calibrate G5 on either the receiver or the iPhone seamlessly implies that the G5 calibration might be performed on the transmitter

it seems like the likely thing, since, when you calibrate on any device, the results are shortly reflected on others. but…

> [@dm61](#):
>
> there are other possibilities

I agree too.

> [@dm61](#):
>
> I also understand that Xdrip+ has its own G5 calibration algorithm, which would imply that calibration is not done on the transmitter.

This is new info to me – yes, I also agree that it implies it happens outside the transmitter.

> [@dm61](#):
>
> the wiki may stay silent about where calibration is performed.

Best way to go.

---

<div class="post-metadata">

**Author:** ![dm61](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/dm61/32/1602_2.png) [@dm61](https://forum.fudiabetes.org/u/dm61)\
**Post date:** [October 11, 2017, 4:57pm UTC](https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292/54 "2017-10-11T16:57:22Z")

</div>

> [@Thomas](#):
>
> do you have an understanding of how the xdrip+ calibration works?

Not really. I’m not using Xdrip+ so I’ve not looked into the code or any other materials or discussions around Xdrip+. If you are into reading Java, here is how [calibration is implemented.](https://github.com/jamorham/xDrip-plus/blob/master/app/src/main/java/com/eveningoutpost/dexdrip/Models/Calibration.java) At a glance, it appears the calculations are done based on raw data coming from the transmitter, and values from the meter, as one would expect. For people interested in Xdrip+ technical details or questions, I’d recommend the [Xdrip-plus gitter channel](https://gitter.im/jamorham/xDrip-plus).

---

<div class="post-metadata">

**Author:** ![Sam](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/sam/32/10595_2.png) [@Sam](https://forum.fudiabetes.org/u/Sam)\
**Post date:** [October 13, 2017, 3:27am UTC](https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292/55 "2017-10-13T03:27:36Z")

</div>

> [@dm61](#):
>
> maybe the receiver and the iPhone app sync independently (?). I also understand that Xdrip+ has its own G5 calibr

Should be an easy experiment to determine if the syncing is occurring on the receiver or not with the G5. Run it with the phone while the receiver is off. Then then turn off phone and the receiver on and see if it picks up the correct data-- if so it must have came from the transmitter

---

<div class="post-metadata">

**Author:** ![dm61](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/dm61/32/1602_2.png) [@dm61](https://forum.fudiabetes.org/u/dm61)\
**Post date:** [October 13, 2017, 4:14am UTC](https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292/56 "2017-10-13T04:14:50Z")

</div>

> [@Sam](#):
>
> Should be an easy experiment

Good point - agreed. I do not have G5, so I am unable to perform the experiment, but I’d be interested to know of the outcome of the experiment.

After looking around some more, I think the most likely scenario is that G5 calibration algorithm still resides on the receiver and on the app, but the calibration parameters are communicated back to the transmitter, which is how all receiving devices are ultimately synced regardless of their on/off state at the time of calibration.

---

<div class="post-metadata">

**Author:** ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)\
**Post date:** [October 13, 2017, 9:41am UTC](https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292/57 "2017-10-13T09:41:04Z")

</div>

> [@Sam](#):
>
> then turn off phone and the receiver on and see if it picks up the correct data-- if so it must have came from the transmitter

That wouldn’t prove it, since, as @dm61 writes, the parameters could have been sent to the transmitter.

---

<div class="post-metadata">

**Author:** ![Sam](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/sam/32/10595_2.png) [@Sam](https://forum.fudiabetes.org/u/Sam)\
**Post date:** [October 13, 2017, 1:19pm UTC](https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292/58 "2017-10-13T13:19:57Z")

</div>

> [@Michel](#):
>
> the parameters could have been sent to the transmitter

That would be what it would determine… whether the transmitter is a 2- way device or just a “transmitter”

ETA just tried it and even with my phone completely off the transmitter fired up, required no calibration and reflected same level as phone had before I powered it down so I agree that the calibration must be somehow acknowledged by the transmitter itself

---

<div class="post-metadata">

**Author:** ![Beacher](https://avatars.discourse-cdn.com/v4/letter/b/a6a055/32.png) [@Beacher](https://forum.fudiabetes.org/u/Beacher)\
**Post date:** [October 13, 2017, 3:04pm UTC](https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292/59 "2017-10-13T15:04:21Z")

</div>

Small point, but in your Oct 9 Resources list, the link for Dexcom Clarity is for the US site, which last time I checked used mg/dL only. People in countries using mmol/L (such as 🇨🇦) will want to go to [https://clarity.dexcom.eu/](https://clarity.dexcom.eu/) (which allows use of either unit of measurement).

---

<div class="post-metadata">

**Author:** ![Thomas](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/thomas/32/1719_2.png) [@Thomas](https://forum.fudiabetes.org/u/Thomas)\
**Post date:** [October 13, 2017, 3:38pm UTC](https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292/60 "2017-10-13T15:38:09Z")

</div>

> [@Sam](#):
>
> That would be what it would determine… whether the transmitter is a 2- way device or just a “transmitter”

I would bet the farm that the G5 Transmitter is a 2-way transmission device.

---

<div class="post-metadata">

**Author:** ![bkh](https://avatars.discourse-cdn.com/v4/letter/b/da6949/32.png) [@bkh](https://forum.fudiabetes.org/u/bkh)\
**Post date:** [October 15, 2017, 10:04pm UTC](https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292/61 "2017-10-15T22:04:13Z")

</div>

With respect to the question of whether the arithmetic that converts raw sensor data to a BG value happens in the receiver or the transmitter, the answer is that it was in the receiver for the G4, but in the transmitter starting with the G5. The G5 “smart transmitter” was explained (at a high level) by the CEO in a couple quarterly conference calls a while ago (about a year?). The G5 transmitter does it so that it can send consistent BG values to simple display devices (like watches) without requiring that they implement all the logic, and to make it easy for multiple devices to interact with the G5 transmitter without complex protocols to keep track of which receiver/pump/phone is in charge: the transmitter is always in charge.

---

<div class="post-metadata">

**Author:** ![Thomas](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/thomas/32/1719_2.png) [@Thomas](https://forum.fudiabetes.org/u/Thomas)\
**Post date:** [October 15, 2017, 10:13pm UTC](https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292/62 "2017-10-15T22:13:02Z")

</div>

> [@bkh](#):
>
> The G5 “smart transmitter” was explained (at a high level) by the CEO in a couple quarterly conference calls a while ago (about a year?).

@bkh - Would you be able to provide a link? I just searched the 2016 and 2017 quarterly earnings calls and couldn’t find it so I must be missing it. I would be interested to read what he said. Thanks.

---

<div class="post-metadata">

**Author:** ![Thomas](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/thomas/32/1719_2.png) [@Thomas](https://forum.fudiabetes.org/u/Thomas)\
**Post date:** [October 15, 2017, 10:29pm UTC](https://forum.fudiabetes.org/t/how-does-a-dexcom-continuous-glucose-monitor-work/2292/63 "2017-10-15T22:29:41Z")

</div>

I think I found the reference to the Dexcom smart transmitter - going back further - in the 2015 Q1 Earnings Call.

However, I don’t see how this makes sense in terms of the xdrip functionality also working? Maybe @docslotnick has some thoughts on how the xdrip could fit into this?

April 29, 2015  
Kevin Sayer, CEO, Dexcom

> Let’s start with Gen 5. As I stated in my opening remarks, the G5 mobile system was filed with the FDA in February of this year.

> The key component of this system is a smart transmitter. The algorithm currently running on the receiver will instead reside in the transmitter. The smart transmitter will directly send a glucose value via Bluetooth low energy to up to two devices. Patients will no longer be required to carry a receiver for connectivity, excuse me – as they do now. The signal will go straight to the smartphone.

> The G5 mobile platform will have a significant impact on DexCom as well. The G5 mobile system utilizes our FDA-approved G4 PLATINUM sensor with the improved 505 algorithm so there will be no disruption in sensor operations for this launch.

> Currently, we believe that the FDA will require each new patient to purchase a DexCom Receiver, which will more than likely be used as a backup.