Hey guys I am running extra panel Samsung Note 5. When I restart a sensor it is asking for two blood sugars for calibration but I have a G6 so I’m confused as to why he’s asking is it a setting I can change

---

<div class="post-metadata">

### Author: ![docslotnick](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/docslotnick/32/75_2.png) [@docslotnick](https://forum.fudiabetes.org/u/docslotnick)
#### Post date: [April 18, 2020, 3:26pm UTC](https://forum.fudiabetes.org/t/g6-and-xdrip/9657/2 "2020-04-18T15:26:08Z")

</div>

@BS500 In Settings\>G5/GG6 Debug, check “I am using G6”. In Settings\>Less common settings\> advanced calibration, uncheck “double calibrations”. Also, make sure your calibration code is on the second system status page. You should have entered it when starting the sensor.