# Does the smell of insulin always imply a pump leak?

**URL:** https://forum.fudiabetes.org/t/does-the-smell-of-insulin-always-imply-a-pump-leak/3200
**Category:** Pump
**Tags:** insulin
**Created:** [January 11, 2018, 11:50pm UTC](https://forum.fudiabetes.org/t/does-the-smell-of-insulin-always-imply-a-pump-leak/3200 "2018-01-11T23:50:43Z")
**Posts on this page:** 13
**Page:** 1

<div class="post-metadata">

### Author: ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)
#### Post date: [January 11, 2018, 11:50pm UTC](https://forum.fudiabetes.org/t/does-the-smell-of-insulin-always-imply-a-pump-leak/3200/1 "2018-01-11T23:50:43Z")

</div>

My son has been sick at home for a couple of days. He is injecting large quantities of insulin all day to try to stay in range, and not succeeding.

I can distinctly smell insulin from his Pod. Is this a certain sign of insulin leak? There is no sign of humidity around the Pod, so no visible insulin.

@Beacher has written several times about needing to use a pen to inject more than 5U because the insulin flowed back around his cannula.

Whenever we have a hard time staying in range, we always have this concern about whether the pump is actually working properly 🙂 I am sure most of the time it is unwarranted.

---

<div class="post-metadata">

### Author: ![Beacher](https://avatars.discourse-cdn.com/v4/letter/b/a6a055/32.png) [@Beacher](https://forum.fudiabetes.org/u/Beacher)
#### Post date: [January 12, 2018, 12:11am UTC](https://forum.fudiabetes.org/t/does-the-smell-of-insulin-always-imply-a-pump-leak/3200/2 "2018-01-12T00:11:56Z")

</div>

> [@Michel](#):
>
> I can distinctly smell insulin from his Pod. Is this a certain sign of insulin leak? There is no sign of humidity around the Pod, so no visible insulin.

It’s a strong indicator, for sure. I can’t smell, but whenever my partner smells insulin, even if the adhesive feels dry and the window looks dry, once I remove the pod I’ll find a pool of insulin under it. Tunneling could well be the cause, as your son is bolusing larger amounts right now.

Thanks for reminding me about using the pen! It’s probably a whole lot easier than doing extended boluses for anything over 5U.

---

<div class="post-metadata">

### Author: ![Chris](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/chris/32/382_2.png) [@Chris](https://forum.fudiabetes.org/u/Chris)
#### Post date: [January 12, 2018, 1:01am UTC](https://forum.fudiabetes.org/t/does-the-smell-of-insulin-always-imply-a-pump-leak/3200/3 "2018-01-12T01:01:39Z")

</div>

In our experience, the strong smell of insulin odor indicates the site is not going to work. Best case, it works moderately, worst case, it appears no insulin is getting in the patient. But we never have a strong smell of insulin and a well working site.

---

<div class="post-metadata">

### Author: ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)
#### Post date: [January 12, 2018, 3:24am UTC](https://forum.fudiabetes.org/t/does-the-smell-of-insulin-always-imply-a-pump-leak/3200/4 "2018-01-12T03:24:17Z")

</div>

> [@Chris](#):
>
> we never have a strong smell of insulin and a well working site.

This may also be true of us.

After spending most of the day floating up around 160-180, we finally saw a definite and significant drop into range after our first pen injection of the day. It may be a coincidence, but it may also be a sign.

Tentatively, I am thinking that, from here on, when we seem to have some trouble going down and we have some insulin smell, we’ll make all large corrections with a pen.

---

<div class="post-metadata">

### Author: ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)
#### Post date: [January 12, 2018, 5:41am UTC](https://forum.fudiabetes.org/t/does-the-smell-of-insulin-always-imply-a-pump-leak/3200/5 "2018-01-12T05:41:40Z")

</div>

So the large correction we made with a pen injection worked well. Since that time, we have left a high basal running on the pump. It appears to work satisfactorily.

It looks to me like what we see is similar to @Beacher: big corrections are failing on this site, but regular basal is OK.

We don’t have a good representative sample yet, though.

---

<div class="post-metadata">

### Author: ![TravelingOn](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/travelingon/32/1583_2.png) [@TravelingOn](https://forum.fudiabetes.org/u/TravelingOn)
#### Post date: [January 12, 2018, 1:14pm UTC](https://forum.fudiabetes.org/t/does-the-smell-of-insulin-always-imply-a-pump-leak/3200/6 "2018-01-12T13:14:11Z")

</div>

> [@Michel](#):
>
> I can distinctly smell insulin from his Pod. Is this a certain sign of insulin leak?

For us, it is.

Also, we have had luck splitting the dose (5u then five minutes then more units) or doing an extended bolus.

---

<div class="post-metadata">

### Author: ![Beacher](https://avatars.discourse-cdn.com/v4/letter/b/a6a055/32.png) [@Beacher](https://forum.fudiabetes.org/u/Beacher)
#### Post date: [January 12, 2018, 2:57pm UTC](https://forum.fudiabetes.org/t/does-the-smell-of-insulin-always-imply-a-pump-leak/3200/7 "2018-01-12T14:57:01Z")

</div>

> [@TravelingOn](#):
>
> we have had luck splitting the dose (5u then five minutes then more units) or doing an extended bolus.

Do you notice any difference in outcome between these tactics? I’ve considered doing boluses in 5U increments, but worry that it’ll still pool too much insulin in one place, leading to the same problem.

> [@Michel](#):
>
> Since that time, we have left a high basal running on the pump. It appears to work satisfactorily.

I’ve found the normal basal also still works satisfactorily when boluses are backing up. But because I can’t be sure how much is getting through, or how much of a higher basal would get through, I just find it easier to change the pod. Insulet will replace them in these circumstances.

---

<div class="post-metadata">

### Author: ![Eric](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/eric/32/9209_2.png) [@Eric](https://forum.fudiabetes.org/u/Eric)
#### Post date: [January 12, 2018, 6:32pm UTC](https://forum.fudiabetes.org/t/does-the-smell-of-insulin-always-imply-a-pump-leak/3200/8 "2018-01-12T18:32:55Z")

</div>

> [@Michel](#):
>
> Tentatively, I am thinking that, from here on, when we seem to have some trouble going down and we have some insulin smell, we’ll make all large corrections with a pen.

You can call those in for a pod replacement, and just replace the pod.

---

<div class="post-metadata">

### Author: ![Aaron](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/aaron/32/1454_2.png) [@Aaron](https://forum.fudiabetes.org/u/Aaron)
#### Post date: [January 12, 2018, 7:22pm UTC](https://forum.fudiabetes.org/t/does-the-smell-of-insulin-always-imply-a-pump-leak/3200/9 "2018-01-12T19:22:41Z")

</div>

> [@Eric](#):
>
> You can call those in for a pod replacement, and just replace the pod.

Yes - but some of us are stubborn and want to 100% confirmation of the problem before pulling the trigger and get some pleasure out of seeing how long they can make the pod last before pulling it (I am speaking for myself here.)

> [@Michel](#):
>
> I can distinctly smell insulin from his Pod.

For me - if I smell insulin or wipe my finger on the POD and have a faint smell of insulin it usually means something is leaking. I am often surprised when I pull the pod off and find a pool of blood on the underside which is usually a sign for me that the canula got bumped and there is a small amount of leakage.

With that said, if you overfill the pod you can get some insulin leaking out during priming that will smell, but is not a problem. I find that the smell can linger for a bit but eventually fades away.

---

<div class="post-metadata">

### Author: ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)
#### Post date: [January 12, 2018, 7:48pm UTC](https://forum.fudiabetes.org/t/does-the-smell-of-insulin-always-imply-a-pump-leak/3200/10 "2018-01-12T19:48:59Z")

</div>

> [@Eric](#):
>
> You can call those in for a pod replacement, and just replace the pod.

Preserving the boy’s sleep is a big criterion for us. Somehow, pod replacement issues most often show up late evenings and nights…

A secondary factor is that we always lose a lot of insulin: I don’t really like to reuse the insulin from the old pod. Although we may come to it if this becomes common.

---

<div class="post-metadata">

### Author: ![Aaron](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/aaron/32/1454_2.png) [@Aaron](https://forum.fudiabetes.org/u/Aaron)
#### Post date: [January 12, 2018, 7:58pm UTC](https://forum.fudiabetes.org/t/does-the-smell-of-insulin-always-imply-a-pump-leak/3200/11 "2018-01-12T19:58:14Z")

</div>

> [@Michel](#):
>
> A secondary factor is that we always lose a lot of insulin

Why are you concerned about loosing insulin?

I ask the question because I have this thing about wasting things and need to justify expenditures in my head. When I use an insulin pen, I have no problems priming a unit or two for each shot. That means if I give 6-8 shots a day for meals and corrections, I am loosing say 8 units. So over the 3 days of a pod life I am loosing about 24 units. If I prime 2 units, I would be loosing 48 units.

For some reason when I see 24 or 48 units left in a pod after a three days, I see this as a huge waste so I need to put it in the pen perspective to make myself feel better.

I do not pay for insulin (my insurance does) so wasting makes no economic difference to me. I try to use as little as possible to save my insurance company money, because I am of the belief that trying to be frugal with the insurance companies’ money means that rates may not go up 😆

---

<div class="post-metadata">

### Author: ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)
#### Post date: [January 13, 2018, 11:20pm UTC](https://forum.fudiabetes.org/t/does-the-smell-of-insulin-always-imply-a-pump-leak/3200/12 "2018-01-13T23:20:55Z")

</div>

10 posts were split to a new topic: [Omnipod: reuse insulin from old pod?](https://forum.fudiabetes.org/t/omnipod-reuse-insulin-from-old-pod/3218)

---

<div class="post-metadata">

### Author: ![TravelingOn](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/travelingon/32/1583_2.png) [@TravelingOn](https://forum.fudiabetes.org/u/TravelingOn)
#### Post date: [January 15, 2018, 1:30am UTC](https://forum.fudiabetes.org/t/does-the-smell-of-insulin-always-imply-a-pump-leak/3200/13 "2018-01-15T01:30:06Z")

</div>

For a laugh, right after I responded to this, apparently EH had a site failure (okay, not funny, but the timing! Made me chuckle). Which of course we didn’t catch - it didn’t smell like insulin. He just drifted high and stayed there (we thought it was the British beer he’d been drinking). It was time later that night to pull the pod and I realized afterwards “doh! Site failed!” He had a lump where the cannula was inserted and it was quite red. Going to pull this one midway day 3 and see if that helps.