Brand new dexcom sensor failed on startup 3 times now? Know any tricks or just give up on it. Warms up for about 10 minutes then says sensor failed.

ETA I just noticed that this and all my other sensors expired 9/29. Is this a factor? Should they still work?

---

<div class="post-metadata">

### Author: ![Eric](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/eric/32/9209_2.png) [@Eric](https://forum.fudiabetes.org/u/Eric)
#### Post date: [November 6, 2017, 11:56pm UTC](https://forum.fudiabetes.org/t/dexcom-sensor-failed-on-startup/2575/2 "2017-11-06T23:56:01Z")

</div>

I’ve used expired sensors with no problem. How soon is your transmitter set to expire? I’ve seen a similar failure with a transmitter that was about to go bad.

---

<div class="post-metadata">

### Author: ![Sam](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/sam/32/10595_2.png) [@Sam](https://forum.fudiabetes.org/u/Sam)
#### Post date: [November 6, 2017, 11:56pm UTC](https://forum.fudiabetes.org/t/dexcom-sensor-failed-on-startup/2575/3 "2017-11-06T23:56:36Z")

</div>

Put on a new sensor and it did same exact thing. Tried starting it with the receiver instead and the receiver said that the transmitter battery was too weak to pair (had about a week left on transmitter I thought, but too late to check now). When I tried to start it with the phone it warned me that transmitter would expire in one week and this would be its last session

---

<div class="post-metadata">

### Author: ![Sam](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/sam/32/10595_2.png) [@Sam](https://forum.fudiabetes.org/u/Sam)
#### Post date: [November 6, 2017, 11:58pm UTC](https://forum.fudiabetes.org/t/dexcom-sensor-failed-on-startup/2575/4 "2017-11-06T23:58:26Z")

</div>

Doubt dexcom will be interested in replacing expired sensors, guess I better just order more

---

<div class="post-metadata">

### Author: ![Chris](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/chris/32/382_2.png) [@Chris](https://forum.fudiabetes.org/u/Chris)
#### Post date: [November 7, 2017, 12:56am UTC](https://forum.fudiabetes.org/t/dexcom-sensor-failed-on-startup/2575/5 "2017-11-07T00:56:52Z")

</div>

Having only used unexpired ones, I can say I have never seen that behavior before. Bummer.

---

<div class="post-metadata">

### Author: ![Eric](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/eric/32/9209_2.png) [@Eric](https://forum.fudiabetes.org/u/Eric)
#### Post date: [November 7, 2017, 1:04am UTC](https://forum.fudiabetes.org/t/dexcom-sensor-failed-on-startup/2575/6 "2017-11-07T01:04:42Z")

</div>

There’s a way around that expired business. We can txt about it if you want.

---

<div class="post-metadata">

### Author: ![MM2](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mm2/32/3032_2.png) [@MM2](https://forum.fudiabetes.org/u/MM2)
#### Post date: [November 7, 2017, 1:08am UTC](https://forum.fudiabetes.org/t/dexcom-sensor-failed-on-startup/2575/7 "2017-11-07T01:08:22Z")

</div>

I just used a sensor that expired 9/14/17, started it 10/10, and lasted until yesterday (11/5), one of my longest runs.

Hope they all do that well, I have another box with same date.

---

<div class="post-metadata">

### Author: ![Sam](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/sam/32/10595_2.png) [@Sam](https://forum.fudiabetes.org/u/Sam)
#### Post date: [November 7, 2017, 1:17am UTC](https://forum.fudiabetes.org/t/dexcom-sensor-failed-on-startup/2575/8 "2017-11-07T01:17:03Z")

</div>

Yeah Eric’s suspicion was right— it was an “almost dead” transmitter that caused the problem, when I tried to start with the receiver instead of the phone it said that and a new transmitter seems to be working now

---

<div class="post-metadata">

### Author: ![TiaG](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/tiag/32/594_2.png) [@TiaG](https://forum.fudiabetes.org/u/TiaG)
#### Post date: [November 7, 2017, 1:59am UTC](https://forum.fudiabetes.org/t/dexcom-sensor-failed-on-startup/2575/9 "2017-11-07T01:59:44Z")

</div>

We’ve had a few just go bad on startup but usually after the first failure it just tells you to replace. Incidentally we have had a run of bad sensors lately, which go bad after just a few days and are really jittery soon after insertion, and even a few that simply didn’t work from the get-go. Anyone else experiencing this? One possibility I’m worried about is that maybe Samson has developed scar tissue on his arms where his sensors are placed, but I haven’t heard of scar tissue being a big issue with Dexcom sensors.

---

<div class="post-metadata">

### Author: ![Chris](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/chris/32/382_2.png) [@Chris](https://forum.fudiabetes.org/u/Chris)
#### Post date: [November 7, 2017, 2:02am UTC](https://forum.fudiabetes.org/t/dexcom-sensor-failed-on-startup/2575/10 "2017-11-07T02:02:24Z")

</div>

@TiaG, we have not had any bad sensors. It is worth pointing out that we have a bit more real estate in a 14 year old than you do.

With that said, we use the back of the arms exclusively, and have had no issues.

---

<div class="post-metadata">

### Author: ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)
#### Post date: [November 7, 2017, 5:01am UTC](https://forum.fudiabetes.org/t/dexcom-sensor-failed-on-startup/2575/11 "2017-11-07T05:01:46Z")

</div>

> [@MM2](#):
>
> started it 10/10, and lasted until yesterday (11/5),

Nice run!

---

<div class="post-metadata">

### Author: ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)
#### Post date: [November 7, 2017, 5:02am UTC](https://forum.fudiabetes.org/t/dexcom-sensor-failed-on-startup/2575/12 "2017-11-07T05:02:59Z")

</div>

> [@TiaG](#):
>
> Anyone else experiencing this?

@TiaG, we have had exactly the same problem for the past 6 weeks. I am not quite sure what to make of it. I have been having my son drink more to see if it is a hydration issue.

---

<div class="post-metadata">

### Author: ![Millz](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/millz/32/12204_2.png) [@Millz](https://forum.fudiabetes.org/u/Millz)
#### Post date: [November 7, 2017, 5:32am UTC](https://forum.fudiabetes.org/t/dexcom-sensor-failed-on-startup/2575/13 "2017-11-07T05:32:20Z")

</div>

@Michel @TiaG how long have you each had the sensors that are giving you more difficulty? Hopefully there not a quality control issue?

---

<div class="post-metadata">

### Author: ![Michel](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/michel/32/387_2.png) [@Michel](https://forum.fudiabetes.org/u/Michel)
#### Post date: [November 7, 2017, 8:10am UTC](https://forum.fudiabetes.org/t/dexcom-sensor-failed-on-startup/2575/14 "2017-11-07T08:10:19Z")

</div>

For us 5-6 weeks. But these are fairly old sensors, although still current.

[Edit] sorry, misunderstood your question yesterday: they are about a year old.

---

<div class="post-metadata">

### Author: ![Thomas](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/thomas/32/1719_2.png) [@Thomas](https://forum.fudiabetes.org/u/Thomas)
#### Post date: [November 7, 2017, 12:19pm UTC](https://forum.fudiabetes.org/t/dexcom-sensor-failed-on-startup/2575/15 "2017-11-07T12:19:23Z")

</div>

We continue to run 11~13 days per sensor quite consistently.

Currently we are using sensors that still have 6 months to go before they hit expiration.

---

<div class="post-metadata">

### Author: ![MM2](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mm2/32/3032_2.png) [@MM2](https://forum.fudiabetes.org/u/MM2)
#### Post date: [November 7, 2017, 1:41pm UTC](https://forum.fudiabetes.org/t/dexcom-sensor-failed-on-startup/2575/16 "2017-11-07T13:41:01Z")

</div>

I hope it continues with current sensor from same Lot.  
Typically I average around 14 days.

---

<div class="post-metadata">

### Author: ![TiaG](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/tiag/32/594_2.png) [@TiaG](https://forum.fudiabetes.org/u/TiaG)
#### Post date: [November 7, 2017, 4:55pm UTC](https://forum.fudiabetes.org/t/dexcom-sensor-failed-on-startup/2575/17 "2017-11-07T16:55:06Z")

</div>

IT’s been about a month or so.

**ETA:** Actually 6 weeks is probably more accurate. We first started having trouble when I was still pregnant.