# Adapter to connect luer-lock infusion set to t:slim cartridge

**URL:** <https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223>\
**Category:** Pump\
**Created:** [January 29, 2020, 5:28pm UTC](https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223 "2020-01-29T17:28:25Z")\
**Posts on this page:** 20\
**Page:** 2

<div class="post-metadata">

**Author:** ![Margoriebest](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/margoriebest/32/10077_2.png) [@Margoriebest](https://forum.fudiabetes.org/u/Margoriebest)\
**Post date:** [April 30, 2022, 4:12am UTC](https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223/21 "2022-04-30T04:12:40Z")

</div>

Thanks to everyone for your suggestions. I am really hoping that I can use the Orbitsoft sets I hope will arrive Monday. I will play around with the connection and try the boiling water idea. If that fails I could mail some of the Luer sets to any of you who might be able to come up with a solution???  
I thought my old Medtronic was a 522 but it Is actually a 554. I had forgotten I swapped my 522 with someone who wanted the 522 for looping. From what I have read, the 544 won’t work for looping.  
Medtronic still have Quicksets with Luer ends, and also with Proprietary ends for the Paradigm pumps. I don’t know, but suspect, that they will eventually phase them out as I think the Luer lock ones were for the orignal Minimed 507 series of pumps. When they changed to the Paradigm pumps they changed their sets to proprietary.  
Maybe we should all start a class petition to the pump companies to listen to us!  
It is ridiculous that I AM desperate enough to even consider the cost of a new pump when I love everything else about the Tandem, and have only had it for a few months.  
Anyway “onwards and upwards” I just want to thank you all on this group for your incredible support. It is great to talk to people who actually “get it” 😀

---

<div class="post-metadata">

**Author:** ![Margoriebest](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/margoriebest/32/10077_2.png) [@Margoriebest](https://forum.fudiabetes.org/u/Margoriebest)\
**Post date:** [April 30, 2022, 4:16am UTC](https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223/22 "2022-04-30T04:16:03Z")

</div>

Inhave to agree I can’t see the boiling water working but I’ll post as soon as I have given it a try.

---

<div class="post-metadata">

**Author:** ![Margoriebest](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/margoriebest/32/10077_2.png) [@Margoriebest](https://forum.fudiabetes.org/u/Margoriebest)\
**Post date:** [April 30, 2022, 4:39am UTC](https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223/23 "2022-04-30T04:39:32Z")

</div>

I must admit all this is like a foreign language to me! If I did as you suggest what would they need me to provide? Would they be able to design an adaptor from having a Tlock and a Luer lock, or would I need to obtain the parts as outlined in the link to the solution on tudiabetes?

---

<div class="post-metadata">

**Author:** ![Chris](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/chris/32/382_2.png) [@Chris](https://forum.fudiabetes.org/u/Chris)\
**Post date:** [April 30, 2022, 5:07am UTC](https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223/24 "2022-04-30T05:07:35Z")

</div>

No worries about the foreign language thing. Those darn engineers have a code of their own. I sort of understand enough to translate. No, you would only need to provide the parts to connect to, the CAD person would be able to create the adapter. You have people near you that would 3D print for a very modest sum, so once you had the file, you could get it printed locally, post it on Thingaverse and link to it here, so everyone could benefit. Also, the link that @MM2 provided would allow the CAD person to have a nice starting place to begin their design.

---

<div class="post-metadata">

**Author:** ![Margoriebest](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/margoriebest/32/10077_2.png) [@Margoriebest](https://forum.fudiabetes.org/u/Margoriebest)\
**Post date:** [April 30, 2022, 5:33am UTC](https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223/25 "2022-04-30T05:33:37Z")

</div>

Thanks for that info. So first I will try heating the Luer- lock sets and try brute force  
The 3d printing Sounds like a better (and cheaper) option than buying a new pump, even if I have to hire or buy a 3-D printer 😂

---

<div class="post-metadata">

**Author:** ![Eric](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/eric/32/9209_2.png) [@Eric](https://forum.fudiabetes.org/u/Eric)\
**Post date:** [April 30, 2022, 4:45pm UTC](https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223/26 "2022-04-30T16:45:45Z")

</div>

> [@Chris](#):
>
> Maybe @Eric has a luer lock set and of course I have plenty of the Tlock cartridges.

Yes, I do have a bunch of stuff like that. But I don’t have any Tandem parts.

---

<div class="post-metadata">

**Author:** ![elver](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/elver/32/8772_2.png) [@elver](https://forum.fudiabetes.org/u/elver)\
**Post date:** [April 30, 2022, 5:38pm UTC](https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223/27 "2022-04-30T17:38:18Z")

</div>

@Eric what Tandem parts do you require? I might have what you need (says the copious Tandem supply hoarder)

---

<div class="post-metadata">

**Author:** ![Eric](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/eric/32/9209_2.png) [@Eric](https://forum.fudiabetes.org/u/Eric)\
**Post date:** [April 30, 2022, 5:55pm UTC](https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223/28 "2022-04-30T17:55:46Z")

</div>

> [@elver](#):
>
> @Eric what Tandem parts do you require?

I don’t know! I have never used a tubed pump, so I don’t really know how it all works.

I have the “other” side, the part you stick into your skin. The end of it has a luer adapter. It’s the non-Tandem part.

But you want to connect that to your Tandem, right? And they have some proprietary adapter that means you need to buy it all from them?

So we want to connect luer to tandem.

How about a non-toxic adhesive that bonds the 2 parts together? Silicone is non-toxic. And epoxy is non-toxic after it sets. It becomes an inert plastic.

---

<div class="post-metadata">

**Author:** ![elver](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/elver/32/8772_2.png) [@elver](https://forum.fudiabetes.org/u/elver)\
**Post date:** [April 30, 2022, 6:00pm UTC](https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223/29 "2022-04-30T18:00:15Z")

</div>

@Eric I can send you both sides of the connector if wanted. The reservoir side and the cannula side. How many would you like for your experiment?

---

<div class="post-metadata">

**Author:** ![Eric](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/eric/32/9209_2.png) [@Eric](https://forum.fudiabetes.org/u/Eric)\
**Post date:** [April 30, 2022, 7:26pm UTC](https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223/30 "2022-04-30T19:26:27Z")

</div>

One is fine. Just want to take a look and see what all is involved.

---

<div class="post-metadata">

**Author:** ![MM2](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mm2/32/3032_2.png) [@MM2](https://forum.fudiabetes.org/u/MM2)\
**Post date:** [April 30, 2022, 8:29pm UTC](https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223/31 "2022-04-30T20:29:24Z")

</div>

> [@Eric](#):
>
> But you want to connect that to your Tandem, right? And they have some proprietary adapter that means you need to buy it all from them?

I think Margoriebest is non-US, and has luer-lock end on her Quick-set (from Med-T). This needs to convert to the Tandem T-Lock cartridge connector.

I have US Medtronic Quick sets, with connector that attaches to Medtronic reservoir (not luer-lock). So that would be different. (3D printer solution offered in the TUD link).

If you check post 10 from Payton, there is example of luer-lock to T-Lock. (from my prior TUD link).

---

<div class="post-metadata">

**Author:** ![Eric](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/eric/32/9209_2.png) [@Eric](https://forum.fudiabetes.org/u/Eric)\
**Post date:** [April 30, 2022, 8:47pm UTC](https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223/32 "2022-04-30T20:47:39Z")

</div>

Thanks @MM2. I looked at the TuD post. That looks like it’s not too hard. Why not just do that?

---

<div class="post-metadata">

**Author:** ![MM2](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mm2/32/3032_2.png) [@MM2](https://forum.fudiabetes.org/u/MM2)\
**Post date:** [May 1, 2022, 2:28am UTC](https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223/33 "2022-05-01T02:28:44Z")

</div>

> [@Eric](#):
>
> Why not just do that?

I want to, but waiting for relative to help out with 3D printing! He uses them at work, and even printed parts to make a bigger better 3D printer!!

But he is very busy with work and kids, and I’m ok with what I have until he has time.

I will be visiting later this summer, maybe then!

---

<div class="post-metadata">

**Author:** ![elver](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/elver/32/8772_2.png) [@elver](https://forum.fudiabetes.org/u/elver)\
**Post date:** [May 1, 2022, 5:05am UTC](https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223/34 "2022-05-01T05:05:16Z")

</div>

@Eric same address the Steamers went to? You sure one is enough? I could easily spare 10 if needed/wanted

---

<div class="post-metadata">

**Author:** ![Eric](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/eric/32/9209_2.png) [@Eric](https://forum.fudiabetes.org/u/Eric)\
**Post date:** [May 1, 2022, 2:56pm UTC](https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223/35 "2022-05-01T14:56:31Z")

</div>

How about 2?

Not sure if I can come up with anything better than what the TuD post showed. That looks like a solid answer. 🤷🏻‍♂️

---

<div class="post-metadata">

**Author:** ![Chris](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/chris/32/382_2.png) [@Chris](https://forum.fudiabetes.org/u/Chris)\
**Post date:** [May 1, 2022, 3:51pm UTC](https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223/36 "2022-05-01T15:51:20Z")

</div>

I am not sure the other end of the adapter they built is Leur. If it is, then no modification is needed, and all you need to do is contact someone who is offering 3D printed parts locally, there are many hobbyists that will print for a small amount. You don’t have to go to a company offering these services, although those exist too. I just wasn’t sure the adapter they built is Tlock to Leur. It looked like it was Tlock to something else, but I don’t have a lot of experience with Leur locks.

---

<div class="post-metadata">

**Author:** ![bkh](https://avatars.discourse-cdn.com/v4/letter/b/da6949/32.png) [@bkh](https://forum.fudiabetes.org/u/bkh)\
**Post date:** [May 1, 2022, 4:10pm UTC](https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223/37 "2022-05-01T16:10:13Z")

</div>

> [@Chris](#):
>
> I am not sure the other end of the adapter they built is Leur.

If you mean the picture in post #45 that’s t:lock to medtronic paradigm, not t:lock to luer. Medtronic sells quicksets with a choice of luer or paradigm connectors.  
[https://aws1.discourse-cdn.com/tudiabetes/original/3X/2/e/2ed79e9ff3af54bb29155ac5e945b653abaafc49.jpeg](https://aws1.discourse-cdn.com/tudiabetes/original/3X/2/e/2ed79e9ff3af54bb29155ac5e945b653abaafc49.jpeg)

---

<div class="post-metadata">

**Author:** ![Chris](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/chris/32/382_2.png) [@Chris](https://forum.fudiabetes.org/u/Chris)\
**Post date:** [May 1, 2022, 4:21pm UTC](https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223/38 "2022-05-01T16:21:16Z")

</div>

Right, so someone needs to convert the Paradigm side to Leur in the design. Then post that one as well.

---

<div class="post-metadata">

**Author:** ![MM2](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mm2/32/3032_2.png) [@MM2](https://forum.fudiabetes.org/u/MM2)\
**Post date:** [May 1, 2022, 4:54pm UTC](https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223/39 "2022-05-01T16:54:32Z")

</div>

There are 2 different solutions posted in TUD link.  
One for luer-lock, other for paradigm reservoir connection.  
Payton was converting animas luer lock sets.

Luer lock in post 10 by Payton.  
[T lock to Luer lock adapter? - #10 by Paytone - Type 1 and LADA - TuDiabetes Forum](https://forum.tudiabetes.org/t/t-lock-to-luer-lock-adapter/81336/10)?

Non luer Quickset/Medtronic reservoir to tandem T:Lock in Post 45 by Luca.  
[T lock to Luer lock adapter? - #45 by Luca - Type 1 and LADA - TuDiabetes Forum](https://forum.tudiabetes.org/t/t-lock-to-luer-lock-adapter/81336/45)?

---

<div class="post-metadata">

**Author:** ![MM2](https://yyz2.discourse-cdn.com/flex030/user_avatar/forum.fudiabetes.org/mm2/32/3032_2.png) [@MM2](https://forum.fudiabetes.org/u/MM2)\
**Post date:** [May 1, 2022, 5:23pm UTC](https://forum.fudiabetes.org/t/adapter-to-connect-luer-lock-infusion-set-to-t-slim-cartridge/9223/40 "2022-05-01T17:23:55Z")

</div>

> [@bkh](#):
>
> Medtronic sells quicksets with a choice of luer or paradigm connectors.

In my case, i have existing stock of quick sets that are not luer, so need the other solution.